I am scanning a hexalobe and it is scanning good. The problem is in the report. It doesn't show where the scan is scanning and reporting the part way out. Why don't they match up closer? Here are some pics.
When you click on "Connect" it will say "Machine not Responding" just hit ok or close or whatever it is. Then click on Find once it's connected type in DRAWSCANASPOINTS, no spaces, then you'll see the value it will say "0" change it to a "1" hit save and exit, count to 10 seconds, then open PC-DMIS back up, then your scans should show as points. You might have to right click on Settings Editor and Run As Administrator
When you click on "Connect" it will say "Machine not Responding" just hit ok or close or whatever it is. Then click on Find once it's connected type in DRAWSCANASPOINTS, no spaces, then you'll see the value it will say "0" change it to a "1" hit save and exit, count to 10 seconds, then open PC-DMIS back up, then your scans should show as points. You might have to right click on Settings Editor and Run As Administrator