Is there a standard setup for calibrating tips on a qual sphere? Number of hits? Prehit/retract? Move speed? Touch speed? Levels? Number of shank hits? start/end angles? etc...
Or is it all about preference? Min of 5 on 2 levels?
Also is there a way to add avoidance in a certain axis during calibration? I have a 10mm star probe and when measuring T5 at a90b180 the Tp20 is hitting the 1 inch shpere while trying to come around the back side of the sphere. So if I could drop it in the Z before it makes that move, I could get to the last hit without it crashing.
I believe that 2 levels are not enough...
The speed of calibration must be the same than measuring speed.
If you have to calibrate a star probe at A90B...,the sphere stem must be at an angle (not 0,0,1) or you have to use 2 spheres.
I believe that 2 levels are not enough...
The speed of calibration must be the same than measuring speed.
If you have to calibrate a star probe at A90B...,the sphere stem must be at an angle (not 0,0,1) or you have to use 2 spheres.