I've recently installed the HR-MP 6 port probe changer on one of our CMMs. It was calibrated and successfully changed probes without issue. However, when I close the lids, instead of pushing the covers over when picking up a probe, it moves over slightly above the lid then crashes on top. I've tried playing with the (before and after pick-up distances) and (edit the probe data) to try and "force" the CMM to slide the lids over before picking up a probe. I would assume the probe changer should operate with the lids closed but for now I'm leaving them open. Is there anything I need to change to make this work with the lids closed?
So I tried taking a hit on the top of the rear support tab of the port, which is lower than the divider the door closes against, and it's still not working
So I tried taking a hit on the top of the rear support tab of the port, which is lower than the divider the door closes against, and it's still not working