Good afternoon all,

"The save process can not find the start of the Measurement Routine."

We have had this error pop a few times randomly on some of our machines. I haven't a clue what it means. Will not allow the machine to run unless you do a full restart/rehome then its perfectly fine. Happens at the "open program" screen before any program has been opened. We are running 2019r2 on all our machines.

Thanks!

Attached Files
Parents
  • Are you:
    - loading the programs off of a network drive?
    - using an antivirus/antimalware program?

    Does this dialog pop up after you have selected a program to open or when the open dialog is shown and you are about to select a program?


    Program is called from a local drive and has been run in the past. Dialog pops up when trying to save a program that is open. The program itself is having a different issue - there is an autocal in the middle of the program that it is skipping everytime. Could the two problems be related?
  • So the program should do calibration at start, then runs a few features, then does another calibration, runs the rest... Anything to check or deal with middle calibration result ?