I seem to be running into a problem with my 2way and 4way checks inside my programs. All of a sudden 2 fixtures in a row are showing an offset on my 4way and 2way 0.5mm in the same direction.
For Example I pick up the tooling balls come back check 2 way and 4way and both fixtures show they are out 0.5 mm in same direction (Machine axis). Does this make any sense ? could it be a bent probe that could be causing this issue ? I'm not sure whats going on here.
We only use the one Probe Setup TP20,100,20,20x2mm. I made sure everything is tight in the probe assembly and just finished calibrating my tips again. Time to see if this changed anything. After Calibrating my tip angles I checked back on the Master Sphere by calling it 0,0,0 and measuring different tips along different axis. Something isn't adding up here
Hi, my conclusion is that my calibration seems good and that there is somehow an offset in the program. I was able to Calibrate with a master, and verify with a bunch of other angles I was good. So it has to be the program so now i'm working through that.
Hi, my conclusion is that my calibration seems good and that there is somehow an offset in the program. I was able to Calibrate with a master, and verify with a bunch of other angles I was good. So it has to be the program so now i'm working through that.