Hello all, could you please verify that I am calibrating my tips properly! I recently calibrated my machine and my results seem offset in one axis. (Calibration sphere comes off the table after every calibration)
I start by selecting A0B0 only ( Say Yes, Sphere has moved) Calibrate it by itself
Then I Select all tips (Including A0B0) and say No, Sphere has not moved)
The second time I calibrate my tip set should I be including A0B0? Also, Do you guys have a separate probe file called something like "Master probe" with just A0B0? or do you calibrate it within the tip set.
I haven't been on a machine in many years and I'm trying to remember the correct way to do this.
The master probe shouldn't be used for measurement but you have to do what you have to do sometimes. If you have more angles on your master probe, you need to select all of you tip angles including A0B0 and answer yes the sphere is moved. It will only us A0B0 to locate the sphere and then calibrate the other angles accordingly. No need to do it again answering No.
The master probe shouldn't be used for measurement but you have to do what you have to do sometimes. If you have more angles on your master probe, you need to select all of you tip angles including A0B0 and answer yes the sphere is moved. It will only us A0B0 to locate the sphere and then calibrate the other angles accordingly. No need to do it again answering No.