Hello all, could you please verify that I am calibrating my tips properly! I recently calibrated my machine and my results seem offset in one axis. (Calibration sphere comes off the table after every calibration)
I start by selecting A0B0 only ( Say Yes, Sphere has moved) Calibrate it by itself
Then I Select all tips (Including A0B0) and say No, Sphere has not moved)
The second time I calibrate my tip set should I be including A0B0? Also, Do you guys have a separate probe file called something like "Master probe" with just A0B0? or do you calibrate it within the tip set.
I haven't been on a machine in many years and I'm trying to remember the correct way to do this.
The answer is that it doesn't matter unless the sphere has actually moved. But I find that I get the best results by not remeasuring the master probe. So my order of operations is to measure the the master tip and answer yes to has it moved, then I measure all other tips.
So you have "Master Probe" with just A0B0? then you go into your tip set and measure all saying (No not moved) right ? And you include A0B0 in your tip set calibration? ( I think you have to ) Thanks for your response Peter.
So you have "Master Probe" with just A0B0? then you go into your tip set and measure all saying (No not moved) right ? And you include A0B0 in your tip set calibration? ( I think you have to ) Thanks for your response Peter.