Hello all, could you please verify that I am calibrating my tips properly! I recently calibrated my machine and my results seem offset in one axis. (Calibration sphere comes off the table after every calibration)
I start by selecting A0B0 only ( Say Yes, Sphere has moved) Calibrate it by itself
Then I Select all tips (Including A0B0) and say No, Sphere has not moved)
The second time I calibrate my tip set should I be including A0B0? Also, Do you guys have a separate probe file called something like "Master probe" with just A0B0? or do you calibrate it within the tip set.
I haven't been on a machine in many years and I'm trying to remember the correct way to do this.
Well I disassembled everything, cleaned it, put it back together. Remade my probe file. I keep randomly getting a PHC9 datum error, could this be causing this issue ? Is it possible the PH9 head is causing issues because its starting to fail ?
Yes Matthew this PH9 wants to retire. I did tighten the set screw on the side to get rid of the PHC9 error. Thank you for your suggestion about the aluminum extensions I will run this by management.