Hello all, could you please verify that I am calibrating my tips properly! I recently calibrated my machine and my results seem offset in one axis. (Calibration sphere comes off the table after every calibration)
I start by selecting A0B0 only ( Say Yes, Sphere has moved) Calibrate it by itself
Then I Select all tips (Including A0B0) and say No, Sphere has not moved)
The second time I calibrate my tip set should I be including A0B0? Also, Do you guys have a separate probe file called something like "Master probe" with just A0B0? or do you calibrate it within the tip set.
I haven't been on a machine in many years and I'm trying to remember the correct way to do this.
For "Master" probe calibration clarification directly from Hexagon, log in to the support portal, and search for "Relating Multiple Probes". It is an OLD 6-page PDF file that goes back to wilcox days. It goes over the details of how and why you need to maintain a master probe with simplicity & great clarity.
As for your PH9 probe angles failing: High standard deviation is indicative of false-trigger hits during probe calibration.
--When you have false hits at any angle other than T1A0B0, it typically means one of two things.
1: Either your probe build is too long or too heavy for a TP20 sensor (a 20mm extension with a 2x20 stylus should be completely fine at any angle), which leads to next thing,
2: Your TP20 probe module is likely worn out.
--Try the same probe build with a known good TP20 probe module. Renishaw TP20 modules are designed & documented to only last about 2,000,000 hits, which in my world doesn't even last 6 months (hence why we went to TP200's for all our machines).