RUNOUT - what's the correct way

Hello,

I'm faceing a problem with measuring RUNOUT on a 540mm tube.

Let me explain:

The tube consist of two datums - A and B. Datum A is at the front of the tube, Datum B is slightly off center on the opposite site. Datums defined out of circles, because measured cylinders are short in distance.

1. We need to measure what's the runout on a circle on the far side of the tube. RUNOUT DIA42 [A - B]. The problem is that with turning the tube on a fixture, we get up to 0.04 variation.
2. We need to measure whats the runout on a circle which is datum A looking at datum B - RUNOUT DIA_A . The problem is the same but with much bigger variation.

I know that form has a big impact on runout - form of datums are not the same.

What I've tried is and I don't know if this is correct:
I've constructed a cylinder out of two circles which are opposite and on the far side of the tube and did the alignment: level and origin on that cylinder- alignment is now in the middle of the tube.

When doing this, the RUNOUT on datum is better, we get the same variation - up to 0.04 and not up to 0.1. The datum [A-B] is measured the same as previous alignment.

Is there any other way?


Now I need to convince our technology, that we measure correct.


BR



Parents
  • Depending up your probing system 200mm may be far too long. If it's a TP20 or Tesa MP type probe then this is the case.

    You might get better results using more hits - these probes will give a tri-lobed (3 pointed) form, which gets bigger the longer the stylus. Using only 4 hits this will likely be skewed one way, assuming you're rotating the probe to get in each end this could have a significant impact.

    You also need to make sure the probe calibrations relate correctly to each other (i.e. they are calibrated at the same time on the same sphere location ideally).

    However, using this type probing for this (touch trigger probe with too long an extension) you are always going to be compromised in terms of accuracy and repeatability.


    If it's an X1-S-H, X3, X5 or an SP25 with the appropriate module then you should be okay, but you should still use more than 4 hits.

    For a 40mm diameter I'd probably use 13 to 17 hits.

    The same as above goes for probe calibration however.


    Probably more relevant is if neither A or B is big enough to measure as a cylinder (which would provide a stable datum to level to), then the drawing likely doesn't reflect how the part assembles and functions.

    That said you should be able to generate a stable A-B datum from two circles to evaluate the dimeter at the far end of the tube.

    You will never get good results check A to B if B isn't long enough (relative to the distance between A and B).

  • When checking runout the point density needs to be as high as practical to capture the form error, we use a minimum of 24 when using trigger probes and 1mm density with scanning heads.