My probe head has been alarming out here and there when I put too many angles in it.... Just during calibration
I have a large prog with appx 20 tip angles... after 2 or 3 qualifications it will stop for a min and maybe pickup again or alarm out all together...
Anyone had this problem?
Thanks
Parrish Dial More information required perhaps to enable assistance, probe type, wrist type, alarm code, 20 tip angles is not a lot to be fair, there is also a current thread where someone else is having problems, is it similar ?
HP-S-X1 scanning head... sorry lol
The fix was a shorter probe. Think i just had too many angles and the probe length was a little long ie 60mm..
Stuck a 1x20 on and it cal up great... go figure