Fast question. I calibrate a tip with a "A" config in terms of prehit retract speed etc, I recalibrate the same tip changing this config to B. The values for both configs are stored or B replaces A? In a program that requires both configs, could I use them both with calibrating A and B before the program start? Thanks in advance!
If you have a rotating probe head, you may be able to calibrate two separate angles that have the same probe vector. Example: A0B0/A0B180 or A90B0/A-90B180. Then you can use different calibration settings for each.