Just as the title says, I am wanting to swap from our current setup to the TP20.

This morning I removed the SP25 scanning setup and placed on the TP20. After I did this, I was unable to get the CMM to move. Spoke to Renishaw and they said I need blanks to go on the end of the multiwire on the probe head? I'm not sure what I am needing to do.

Equipment used currently - PH10QM/SP25M/SP25-2 = PHC10-2 Probe Head Controller. PCDMIS 2020R1

Any input would be appreciated.
Parents
  • Do you have to restart pcdmis to do this? OR change any settings? I did exactly what you described and the CMM shuts down immediately. Machine start via controller, hit play and then shuts back off.
  • the only thing I have to do is run a program (since I don't have a rack system). Program starts with ANY probe (of the type currently loaded), in your case the scanner probe. Then do a loadprobe command of the TP20 probe (doesn't matter which one, just needs to be a TP20 build). THEN it prompts you to hit the e-stop, you change the probe to the TP20, reset the e-stop, good to go:

    LOADPROBE/000-SP25-3X21
                TIP/T1A0B0, SHANKIJK=0, 0, 1, ANGLE=0
                FORMAT/TEXT,OPTIONS, ,HEADINGS,SYMBOLS, ;NOM,TOL,MEAS,DEV,OUTTOL, , 
                PROBECOM/ON
                LOADPROBE/000-TP20
                COMMENT/OPER,NO,FULL SCREEN=NO,AUTO-CONTINUE=NO,
                HIT E-STOP BUTTON ON JOGBOX
                CHANGE TO TP20 PROBE BODY
                RELEASE E-STOP BUTTON
                PRESS MACHINE START BUTTON
                TIP/T1A0B0, SHANKIJK=0, 0, 1, ANGLE=0
                COMMENT/OPER,NO,FULL SCREEN=NO,AUTO-CONTINUE=NO,
                make sure probe is enabled
    


    Reverse it to go from TP20 to scanner
Reply
  • the only thing I have to do is run a program (since I don't have a rack system). Program starts with ANY probe (of the type currently loaded), in your case the scanner probe. Then do a loadprobe command of the TP20 probe (doesn't matter which one, just needs to be a TP20 build). THEN it prompts you to hit the e-stop, you change the probe to the TP20, reset the e-stop, good to go:

    LOADPROBE/000-SP25-3X21
                TIP/T1A0B0, SHANKIJK=0, 0, 1, ANGLE=0
                FORMAT/TEXT,OPTIONS, ,HEADINGS,SYMBOLS, ;NOM,TOL,MEAS,DEV,OUTTOL, , 
                PROBECOM/ON
                LOADPROBE/000-TP20
                COMMENT/OPER,NO,FULL SCREEN=NO,AUTO-CONTINUE=NO,
                HIT E-STOP BUTTON ON JOGBOX
                CHANGE TO TP20 PROBE BODY
                RELEASE E-STOP BUTTON
                PRESS MACHINE START BUTTON
                TIP/T1A0B0, SHANKIJK=0, 0, 1, ANGLE=0
                COMMENT/OPER,NO,FULL SCREEN=NO,AUTO-CONTINUE=NO,
                make sure probe is enabled
    


    Reverse it to go from TP20 to scanner