After a few weeks when I open CMM programs I receive and error message saying "Tips not calibrated frequently," or something like that. Is there a setting where you can change that interval? Like, lets say I want the popup to come up one month after calibration.
If you go into probe utilities, then click on Setup... you can change the number of hours/days that it asks you for calibration. It's just saying you haven't qualified the probes to the sphere in "X" amount of time.
I think if you set it up as -1, it will never ask you.
Perfect, thanks so much. Haha, I know, Matthew. The machines barely get used, though. I'm pushing them to be used more *shrugs* we have 4 machines and the amount of runtime is probably less than having one machine...