hexagon logo

Different time measurements on the same machines

Hello

In my lab. We have two Hexagon Global Advantage 15.26.14 machines . We have problem with d
ifferent time measurements on this machines. We do the calibration of stylus and on first machine measurment takes 3.3h and on second takes 5h. Where is the problem? We have the same programs with the same parameters in program : hits, prehit/retract, move speed, touch speed. Parameters in the windows F9 and F10 are the same. Please show me the way to solutions this problem .

Greetings Ryan91
Parents Reply Children
No Data