Different time measurements on the same machines

Hello

In my lab. We have two Hexagon Global Advantage 15.26.14 machines . We have problem with d
ifferent time measurements on this machines. We do the calibration of stylus and on first machine measurment takes 3.3h and on second takes 5h. Where is the problem? We have the same programs with the same parameters in program : hits, prehit/retract, move speed, touch speed. Parameters in the windows F9 and F10 are the same. Please show me the way to solutions this problem .

Greetings Ryan91
Parents
  • One more thing to check after all the other suggestions - an edge case sort of thing. Make sure both CMMs are communicating over ethernet. Maybe one of the machines is using ethernet connection while the other is using RS-232 (serial) connection. That makes a big difference when calibrating an analog probe.
Reply
  • One more thing to check after all the other suggestions - an edge case sort of thing. Make sure both CMMs are communicating over ethernet. Maybe one of the machines is using ethernet connection while the other is using RS-232 (serial) connection. That makes a big difference when calibrating an analog probe.