My calibration sphere is fixed at a point. But whenever I qualify a new probe and select no change in movement of sphere, the value of x y and z of Sphere always is different from nominal
Are you sure you don't mean the xyz of the tip you have just measured ? You have nominals for tip and actuals as long as you StdDev is within 0.005mm you are ok.