qualifying multiple probes with multiple spheres in operator mode

i need to locate two spheres with my first probe, qualify the first probe on the big sphere (sn24), and qualify the second probe on the smaller sphere (sn25)  .

how do i do this in operator mode.

right now the first probe locates and qualifies on big sphere, but the second sphere is never located

Attached Files
Parents
  • Marked sets allows you to create sections of your program to be ran according to how they are defined in marked sets.
    Parameter sets in probe calibration allow you to define a set of probe calibrations that you can use the drop down or auto-calibrate statements in your program to run a set of predefined calibrations.
    Then you can add auto-calibrate statements to your programs and define them using marked sets that will allow your operator to select between running the program, or the probe calibration through the marked sets window
  • got it. but how do i modify the program so that operators have that functionality from their "calibrate tips" prompt?

    that is, i need the large probe to locate both spheres, qualify the large tip on the large sphere, then qualify the small probe on the small sphere.

    see attached screenshot below