In my company in TIGO SF machines we use TP20 modules in various configurations. Normally they use low force modules when a 0.5 tip is mounted and a 6way module for all the rest. The TP20 machines as compared to TIGO machines with X1C head are way...way less repeatable/accurate.
I read in the specs that 6way are generally less precise than standard force modules and also more expensive. I would like to replace 2 probes specifically 1 with 6way that is in star formation (length of probes out of specs with 40 mm vertically and 1 tip at 30 mm horizontally) and a low force module that is in star formation with 0.5mm tips left right and a pointing (x+) 2x40 mm tip. I know both probes are out of specs but replacing them with standard force modules does not seem to improve the results by a lot (at least some preliminary testing).
So i would like more experienced user to let me know if using low force modules in star configs is a good idea (out of specs according to renishaw), how can I mount a 0.5mm in a star probe if a low force module is needed and finally what are your opinions of 6way probes for none special applications needing measures in Z-
IMHO touchspeed is just as relevant in this discussion, as the modules themselves.
Any probes under 1mm should have a significantly reduced touchspeed implemented (and calibrated whilst set to to the slower touchspeed), any time it is used.
IMHO touchspeed is just as relevant in this discussion, as the modules themselves.
Any probes under 1mm should have a significantly reduced touchspeed implemented (and calibrated whilst set to to the slower touchspeed), any time it is used.