So, I have an old Renishaw ACR1 changer on one of my machines and a spare MCR20 that I would like to incorporate into the machine. I have worked with multiple changers of the same type, but never 2 dissimilar types. Is this possible considering they are 2 different types of changers?

My thought was setting up a slot on the ACR1 with just the TP20 body and using that to pick up the TP20 modules in the MCR20. Would this body without a module pick up and return modules to the MCR20 before returning to the ACR1 to pick up other probe assemblies in the ACR1?

Parents
  • That is exactly how they work.  I have this combination, and much more.  The biggest issue is properly defining the contents of each slot.  In general (let me phrase it this way) if any piece of a probe assembly is on the slot, then the probe name is added to the slot.  Some slots will have multiple probe names.  For example, you might have two TP20 bodies, and each can use any one of several probe modules.

Reply
  • That is exactly how they work.  I have this combination, and much more.  The biggest issue is properly defining the contents of each slot.  In general (let me phrase it this way) if any piece of a probe assembly is on the slot, then the probe name is added to the slot.  Some slots will have multiple probe names.  For example, you might have two TP20 bodies, and each can use any one of several probe modules.