Morning all
We have a global S machine on the floor, this morning I found the 9 hits calibration changed to scan mode, Which option should I check/uncheck to fix it ?
Thank you
Your Products have been synced, click here to refresh
Morning all
We have a global S machine on the floor, this morning I found the 9 hits calibration changed to scan mode, Which option should I check/uncheck to fix it ?
Thank you
Is the checkbox under the measure window that mentions "Calibrate ScanRDV" checked by chance?
Someone checked the user defined calibration order and unchecked use trax calibration. I posted a note beside the desktop mentioned that .
Would you share your probe configuration as screenshot? so that we can help you in more detail.
This is the right setting.
This is the right setting.
Can you also share "Measure" as screenshot please?
This happened yesterday, one of QA unchecked "use trax calibration". After I did check the box, it went back to normal. it fixed yesterday. Thank you .