Probe not scanning during calibration

I just got back from PCDMIS 202.  We were told to disable TRAX calibration if we wanted the probe to also scan while doing calibration.  Otherwise, it would just do the standard 24 points (or whatever value we select).  I disabled TRAX and the probe is still not scanning when calibrating.  Any idea what I'm doing wrong?

Parents
  • using Trax is dependent on your CMM and controller. if you have a Global chrome with an analog probe then you turn it off if you have a global blue or green with an analog probe you turn it on there are other conditions that matter as well. Reach out to Hex tech support and get the manual for your probe sensor. it will tell you what settings you need.

Reply
  • using Trax is dependent on your CMM and controller. if you have a Global chrome with an analog probe then you turn it off if you have a global blue or green with an analog probe you turn it on there are other conditions that matter as well. Reach out to Hex tech support and get the manual for your probe sensor. it will tell you what settings you need.