hexagon logo

HR-R3 CALIBRATION AND CONFIGURATION, NEW CMM

Hi team,

I'm new to the group and using PC-DMIS.

I have experience with other CMM softwares, but this a new CMM Global S was installed at my company.

It has a travel rack attached to the right side, a scanner HP-L.10.10, and an HR-R3.

And HP-S-X1H sensor.

The CMM was calibrated, but not the racks.

After reviewing some pages and documents, I finally calibrated the travel rack and stylus.

However, the issue now is with the HR-R3.

I have multiple questions regarding the calibration of this rack.

Unfortunately, I don't have response of Hexagon at this moment after 2 weeks , for support and calibration for HR-R3 rack.

By this reason I'm trying to ask to the community.

This is the normal stylus configuration

And it is the travel rack calibrated

Trying to calibrate the HR-R3

1)I cant find the HR-R3 on the options for the rack, should be select the TESASTAR-R(HR-R) ?

I believe like this is the correct option because is very similar to the pictures on some documents

2)The 2nd tab COMM, the PC have 3 COMMs , how I can know the correct port and settings?

 

3)I need use both racks on some parts , use the stylus for example the stylus defined 2 3x70 on travel rack port 2, then switch to scanner HP-L and measure other things in the same part.

My dude is how configure the ports on 2nd rack HR-R3 for use the stylus 2 3x70 dock it on port 1, and then take the HP-L on port 4

should be declare port 1  with the 5 stylus of travel rack?

only 1 stylus?

I really appreciate any help about this issue

Regards

Parents
  • That is frustrating that if it is a new install, that the rack was not calibrated for you or atleast you were trained on how to calibrate it moving forward.

    I do not have experience with the travelling rack, but to work out the COM port, you could unplug the head/rack controller as you are looking at the device manager to see which COM port disappears.

    Once you get the COM port, you could then try the Tesastar HR-R rack?

    I would be interested to know once you get it sorted.

Reply
  • That is frustrating that if it is a new install, that the rack was not calibrated for you or atleast you were trained on how to calibrate it moving forward.

    I do not have experience with the travelling rack, but to work out the COM port, you could unplug the head/rack controller as you are looking at the device manager to see which COM port disappears.

    Once you get the COM port, you could then try the Tesastar HR-R rack?

    I would be interested to know once you get it sorted.

Children