Let me preface that I'm very new to programming, as well as, operating the CMM. So something that might seem like a obvious solution to most of you is like finding a needle in a hay stack for me, haha. Thanks!

While writing a program for a part, I have run into an issue with the probes movement. You can see, in the image above, that there is a line moving out of frame to the cylinder highlighted. I have been taught to prevent unintentional contact leading to a system crash to add a move point as seen (or use clearance planes/cube). The move point line should move all the way to the probe (indicated by the arrow), but instead stops at the question mark then deviates downwards and along the x-axis the then back towards the cylinder. The path then collides with the structure (indicated by the yellow X) and we cant be having that. My question is why? Why is it just randomly pivoting like this?

  • Makes me wonder if there are any freaks out there that use Dmis mode.

  • Hahaha, guilty as charged! I generally use summary mode.

  •              ,SET WARNING LIMIT = FALSE,MINIMUM = 10,MAXIMUM = 40
                  ,PART SENSOR NUM=DEFAULT,X SCALE= 20,Y SCALE= 20,Z SCALE= 20,PART TEMP=20
                MOVESPEED/ 250
                TOUCHSPEED/ 2
                PREHIT/0.2
                CHECK/0.3,1
                RETRACT/0.2
    READPOINT  =FEAT/POINT,CARTESIAN
                THEO/<1.1491,0.1249,1.155>,<0,0,1>
                ACTL/<1.1491,0.1249,1.155>,<0,0,1>
                READPOINT/
                MODE/DCC
    A1         =ALIGNMENT/START,RECALL:STARTUP,LIST=YES
                  ALIGNMENT/TRANS,XAXIS,READPOINT
                  ALIGNMENT/TRANS_OFFSET,XAXIS,-1.1491
                  ALIGNMENT/TRANS,YAXIS,READPOINT
                  ALIGNMENT/TRANS_OFFSET,YAXIS,-0.1249
                  ALIGNMENT/TRANS,ZAXIS,READPOINT
                  ALIGNMENT/TRANS_OFFSET,ZAXIS,-1.155
                ALIGNMENT/END
                MOVE/POINT,NORMAL,<0.2166,-2.0445,2>
                TIP/T1A-10B0, SHANKIJK=0, -0.1736, 0.9848, ANGLE=0
                WORKPLANE/YMINUS
    PLNC VECTOR POINTS=GROUP/SHOWALLPARAMS=YES
    PNT1         =FEAT/CONTACT/VECTOR POINT/DEFAULT,POLAR
                  THEO/<0.9864,77.3168,0>,<0,-1,0>
                  ACTL/<0.9864,77.3168,0>,<0,-1,0>
                  TARG/<0.9864,77.3168,0>,<0,-1,0>
                  SNAP=NO
                  SHOW FEATURE PARAMETERS=NO
                  SHOW CONTACT PARAMETERS=YES
                    AVOIDANCE MOVE=BOTH,DISTANCE BEFORE=0.1,DISTANCE AFTER=0.1
                  SHOW HITS=NO
    PNT2         =FEAT/CONTACT/VECTOR POINT/DEFAULT,POLAR
                  THEO/<0.3091,45.4103,0>,<0,-1,0>
                  ACTL/<0.3091,45.4103,0>,<0,-1,0>
                  TARG/<0.3091,45.4103,0>,<0,-1,0>
                  SNAP=NO
                  SHOW FEATURE PARAMETERS=NO
                  SHOW CONTACT PARAMETERS=YES
                    AVOIDANCE MOVE=BOTH,DISTANCE BEFORE=0.1,DISTANCE AFTER=0.1
                  SHOW HITS=NO
    PNT3         =FEAT/CONTACT/VECTOR POINT/DEFAULT,POLAR
                  THEO/<3.9353,2.8347,0>,<0,-1,0>
                  ACTL/<3.9353,2.8347,0>,<0,-1,0>
                  TARG/<3.9353,2.8347,0>,<0,-1,0>
                  SNAP=NO
                  SHOW FEATURE PARAMETERS=NO
                  SHOW CONTACT PARAMETERS=YES
                    AVOIDANCE MOVE=BOTH,DISTANCE BEFORE=0.1,DISTANCE AFTER=0.1
                  SHOW HITS=NO
    PNT4         =FEAT/CONTACT/VECTOR POINT/DEFAULT,POLAR
                  THEO/<4.0503,14.3909,0>,<0,-1,0>
                  ACTL/<4.0503,14.3909,0>,<0,-1,0>
                  TARG/<4.0503,14.3909,0>,<0,-1,0>
                  SNAP=NO
                  SHOW FEATURE PARAMETERS=NO
                  SHOW CONTACT PARAMETERS=YES
                    AVOIDANCE MOVE=BOTH,DISTANCE BEFORE=0.1,DISTANCE AFTER=0.1
                  SHOW HITS=NO
    PNT5         =FEAT/CONTACT/VECTOR POINT/DEFAULT,POLAR
                  THEO/<8.1976,1.279,0>,<0,-1,0>
                  ACTL/<8.1976,1.279,0>,<0,-1,0>
                  TARG/<8.1976,1.279,0>,<0,-1,0>
                  SNAP=NO
                  SHOW FEATURE PARAMETERS=NO
                  SHOW CONTACT PARAMETERS=YES
                    AVOIDANCE MOVE=BOTH,DISTANCE BEFORE=0.1,DISTANCE AFTER=0.1
                  SHOW HITS=NO
    PNT6         =FEAT/CONTACT/VECTOR POINT/DEFAULT,POLAR
                  THEO/<8.108,7.4369,0>,<0,-1,0>
                  ACTL/<8.108,7.4369,0>,<0,-1,0>
                  TARG/<8.108,7.4369,0>,<0,-1,0>
                  SNAP=NO
                  SHOW FEATURE PARAMETERS=NO
                  SHOW CONTACT PARAMETERS=YES
                    AVOIDANCE MOVE=BOTH,DISTANCE BEFORE=0.1,DISTANCE AFTER=0.1
                  SHOW HITS=NO
    PNT7         =FEAT/CONTACT/VECTOR POINT/DEFAULT,POLAR
                  THEO/<11.7275,5.1391,0>,<0,-1,0>
                  ACTL/<11.7275,5.1391,0>,<0,-1,0>
                  TARG/<11.7275,5.1391,0>,<0,-1,0>
                  SNAP=NO
                  SHOW FEATURE PARAMETERS=NO
                  SHOW CONTACT PARAMETERS=YES
                    AVOIDANCE MOVE=BOTH,DISTANCE BEFORE=0.1,DISTANCE AFTER=0.1
                  SHOW HITS=NO
    PNT8         =FEAT/CONTACT/VECTOR POINT/DEFAULT,POLAR
                  THEO/<12.0839,0.8217,0>,<0,-1,0>
                  ACTL/<12.0839,0.8217,0>,<0,-1,0>
                  TARG/<12.0839,0.8217,0>,<0,-1,0>
                  SNAP=NO
                  SHOW FEATURE PARAMETERS=NO
                  SHOW CONTACT PARAMETERS=YES
                    AVOIDANCE MOVE=BOTH,DISTANCE BEFORE=0.1,DISTANCE AFTER=0.1
                  SHOW HITS=NO
    PNT9         =FEAT/CONTACT/VECTOR POINT/DEFAULT,POLAR
                  THEO/<16.5842,0.7549,0>,<0,-1,0>
                  ACTL/<16.5842,0.7549,0>,<0,-1,0>
                  TARG/<16.5842,0.7549,0>,<0,-1,0>
                  SNAP=NO
                  SHOW FEATURE PARAMETERS=NO
                  SHOW CONTACT PARAMETERS=YES
                    AVOIDANCE MOVE=BOTH,DISTANCE BEFORE=0.1,DISTANCE AFTER=0.1
                  SHOW HITS=NO
    PNT10        =FEAT/CONTACT/VECTOR POINT/DEFAULT,POLAR
                  THEO/<16.5632,3.4367,0>,<0,-1,0>
                  ACTL/<16.5632,3.4367,0>,<0,-1,0>
                  TARG/<16.5632,3.4367,0>,<0,-1,0>
                  SNAP=NO
                  SHOW FEATURE PARAMETERS=NO
                  SHOW CONTACT PARAMETERS=YES
                    AVOIDANCE MOVE=BOTH,DISTANCE BEFORE=0.1,DISTANCE AFTER=0.1
                  SHOW HITS=NO
    PNT11        =FEAT/CONTACT/VECTOR POINT/DEFAULT,POLAR
                  THEO/<22.5614,2.4764,0>,<0,-1,0>
                  ACTL/<22.5614,2.4764,0>,<0,-1,0>
                  TARG/<22.5614,2.4764,0>,<0,-1,0>
                  SNAP=NO
                  SHOW FEATURE PARAMETERS=NO
                  SHOW CONTACT PARAMETERS=YES
                    AVOIDANCE MOVE=BOTH,DISTANCE BEFORE=0.1,DISTANCE AFTER=0.1
                  SHOW HITS=NO
                  MOVE/POINT,NORMAL,<22.5,-0.2296,2>
                  MOVE/POINT,NORMAL,<24,-0.2296,2>
    PNT12        =FEAT/CONTACT/VECTOR POINT/DEFAULT,POLAR
                  THEO/<25.5378,1.2829,0>,<0,-1,0>
                  ACTL/<25.5378,1.2829,0>,<0,-1,0>
                  TARG/<25.5378,1.2829,0>,<0,-1,0>
                  SNAP=NO
                  SHOW FEATURE PARAMETERS=NO
                  SHOW CONTACT PARAMETERS=YES
                    AVOIDANCE MOVE=BOTH,DISTANCE BEFORE=0.1,DISTANCE AFTER=0.1
                  SHOW HITS=NO
    PNT13        =FEAT/CONTACT/VECTOR POINT/DEFAULT,POLAR
                  THEO/<25.8009,0.4669,0>,<0,-1,0>
                  ACTL/<25.8009,0.4669,0>,<0,-1,0>
                  TARG/<25.8009,0.4669,0>,<0,-1,0>
                  SNAP=NO
                  SHOW FEATURE PARAMETERS=NO
                  SHOW CONTACT PARAMETERS=YES
                    AVOIDANCE MOVE=BOTH,DISTANCE BEFORE=0.1,DISTANCE AFTER=0.1
                  SHOW HITS=NO
                ENDGROUP/ID=PLNC VECTOR POINTS
    PLNC       =FEAT/PLANE,CARTESIAN,OUTLINE,NO,LEAST_SQR
                THEO/<11.9441,0,0.6006>,<0,-1,0>
                ACTL/<11.9441,0,0.6006>,<0,-1,0>
                CONSTR/PLANE,BF,PNT1,PNT2,PNT3,PNT4,PNT5,PNT6,PNT7,PNT8,PNT9,PNT10,PNT11,PNT12,PNT13,,
                OUTLIER_REMOVAL/OFF,3
                FILTER/OFF,WAVELENGTH=0
                MOVE/POINT,NORMAL,<25.8,-2,0.3748>
                TIP/T1A105B-180, SHANKIJK=0, -0.9659, -0.2588, ANGLE=0
                MOVE/POINT,NORMAL,<25.8478,-1,3>
                MOVE/POINT,NORMAL,<25.8478,-1,-0.5>
                WORKPLANE/ZPLUS
    PLNA VECTOR POINTS=GROUP/SHOWALLPARAMS=YES
    PNT14        =FEAT/CONTACT/VECTOR POINT/DEFAULT,POLAR
                  THEO/<26.2351,6.1072,0>,<0,0,-1>
                  ACTL/<26.2351,6.1072,0>,<0,0,-1>
                  TARG/<26.2351,6.1072,0>,<0,0,-1>
                  SNAP=NO
                  SHOW FEATURE PARAMETERS=NO
                  SHOW CONTACT PARAMETERS=YES
                    AVOIDANCE MOVE=BOTH,DISTANCE BEFORE=0.1,DISTANCE AFTER=0.1
                  SHOW HITS=NO
    PNT15        =FEAT/CONTACT/VECTOR POINT/DEFAULT,POLAR
                  THEO/<25.8492,0.5987,0>,<0,0,-1>
                  ACTL/<25.8492,0.5987,0>,<0,0,-1>
                  TARG/<25.8492,0.5987,0>,<0,0,-1>
                  SNAP=NO
                  SHOW FEATURE PARAMETERS=NO
                  SHOW CONTACT PARAMETERS=YES
                    AVOIDANCE MOVE=BOTH,DISTANCE BEFORE=0.1,DISTANCE AFTER=0.1
                  SHOW HITS=NO
                  MOVE/POINT,NORMAL,<25.8492,-6.5,-0.5>
                  MOVE/POINT,NORMAL,<19,-6.5,-0.5>
    PNT16        =FEAT/CONTACT/VECTOR POINT/DEFAULT,POLAR
                  THEO/<18.5,0.5312,0>,<0,0,-1>
                  ACTL/<18.5,0.5312,0>,<0,0,-1>
                  TARG/<18.5,0.5312,0>,<0,0,-1>
                  SNAP=NO
                  SHOW FEATURE PARAMETERS=NO
                  SHOW CONTACT PARAMETERS=YES
                    AVOIDANCE MOVE=BOTH,DISTANCE BEFORE=0.1,DISTANCE AFTER=0.1
                  SHOW HITS=NO
    PNT17        =FEAT/CONTACT/VECTOR POINT/DEFAULT,POLAR
                  THEO/<19.3664,17.9033,0>,<0,0,-1>
                  ACTL/<19.3664,17.9033,0>,<0,0,-1>
                  TARG/<19.3664,17.9033,0>,<0,0,-1>
                  SNAP=NO
                  SHOW FEATURE PARAMETERS=NO
                  SHOW CONTACT PARAMETERS=YES
                    AVOIDANCE MOVE=BOTH,DISTANCE BEFORE=0.1,DISTANCE AFTER=0.1
                  SHOW HITS=NO
                  MOVE/POINT,NORMAL,<19,-6.5,-0.5>
                  MOVE/POINT,NORMAL,<9.5,-6.5,-0.5>
    PNT18        =FEAT/CONTACT/VECTOR POINT/DEFAULT,POLAR
                  THEO/<8.9681,1.2077,0>,<0,0,-1>
                  ACTL/<8.9681,1.2077,0>,<0,0,-1>
                  TARG/<8.9681,1.2077,0>,<0,0,-1>
                  SNAP=NO
                  SHOW FEATURE PARAMETERS=NO
                  SHOW CONTACT PARAMETERS=YES
                    AVOIDANCE MOVE=BOTH,DISTANCE BEFORE=0.1,DISTANCE AFTER=0.1
                  SHOW HITS=NO
    PNT19        =FEAT/CONTACT/VECTOR POINT/DEFAULT,POLAR
                  THEO/<10.8788,33.2165,0>,<0,0,-1>
                  ACTL/<10.8788,33.2165,0>,<0,0,-1>
                  TARG/<10.8788,33.2165,0>,<0,0,-1>
                  SNAP=NO
                  SHOW FEATURE PARAMETERS=NO
                  SHOW CONTACT PARAMETERS=YES
                    AVOIDANCE MOVE=BOTH,DISTANCE BEFORE=0.1,DISTANCE AFTER=0.1
                  SHOW HITS=NO
                  MOVE/POINT,NORMAL,<9.5,-6.5,-0.5>
                  MOVE/POINT,NORMAL,<3,-6.5,-0.5>
    PNT20        =FEAT/CONTACT/VECTOR POINT/DEFAULT,POLAR
                  THEO/<2.6084,4.0779,0>,<0,0,-1>
                  ACTL/<2.6084,4.0779,0>,<0,0,-1>
                  TARG/<2.6084,4.0779,0>,<0,0,-1>
                  SNAP=NO
                  SHOW FEATURE PARAMETERS=NO
                  SHOW CONTACT PARAMETERS=YES
                    AVOIDANCE MOVE=BOTH,DISTANCE BEFORE=0.1,DISTANCE AFTER=0.1
                  SHOW HITS=NO
    PNT21        =FEAT/CONTACT/VECTOR POINT/DEFAULT,POLAR
                  THEO/<6.529,65.9993,0>,<0,0,-1>
                  ACTL/<6.529,65.9993,0>,<0,0,-1>
                  TARG/<6.529,65.9993,0>,<0,0,-1>
                  SNAP=NO
                  SHOW FEATURE PARAMETERS=NO
                  SHOW CONTACT PARAMETERS=YES
                    AVOIDANCE MOVE=BOTH,DISTANCE BEFORE=0.1,DISTANCE AFTER=0.1
                  SHOW HITS=NO
                  MOVE/POINT,NORMAL,<3,-1,-0.5>
                  TIP/T1A105B-170, SHANKIJK=0.1677, -0.9513, -0.2588, ANGLE=-34.2656
    PNT46        =FEAT/CONTACT/VECTOR POINT/DEFAULT,POLAR
                  THEO/<7.6754,85.514,0>,<0,0,-1>
                  ACTL/<7.6754,85.514,0>,<0,0,-1>
                  TARG/<7.6754,85.514,0>,<0,0,-1>
                  SNAP=NO
                  SHOW FEATURE PARAMETERS=NO
                  SHOW CONTACT PARAMETERS=YES
                    AVOIDANCE MOVE=BOTH,DISTANCE BEFORE=0.1,DISTANCE AFTER=0.1
                  SHOW HITS=NO
    PNT47        =FEAT/CONTACT/VECTOR POINT/DEFAULT,POLAR
                  THEO/<0.6682,51.4336,0>,<0,0,-1>
                  ACTL/<0.6682,51.4336,0>,<0,0,-1>
                  TARG/<0.6682,51.4336,0>,<0,0,-1>
                  SNAP=NO
                  SHOW FEATURE PARAMETERS=NO
                  SHOW CONTACT PARAMETERS=YES
                    AVOIDANCE MOVE=BOTH,DISTANCE BEFORE=0.1,DISTANCE AFTER=0.1
                  SHOW HITS=NO
                  MOVE/POINT,NORMAL,<1,-6,-0.5>
                  TIP/T1A105B90, SHANKIJK=-0.9659, 0, -0.2588, ANGLE=90
                  MOVE/POINT,NORMAL,<-3,-6.5,-0.5>
                  MOVE/POINT,NORMAL,<-3,11.8,-1>
    PNT23        =FEAT/CONTACT/VECTOR POINT/DEFAULT,POLAR
                  THEO/<12.2404,73.4433,0>,<0,0,-1>
                  ACTL/<12.2404,73.4433,0>,<0,0,-1>
                  TARG/<12.2404,73.4433,0>,<0,0,-1>
                  SNAP=NO
                  SHOW FEATURE PARAMETERS=NO
                  SHOW CONTACT PARAMETERS=YES
                    AVOIDANCE MOVE=BOTH,DISTANCE BEFORE=0.1,DISTANCE AFTER=0.1
                  SHOW HITS=NO
    PNT24        =FEAT/CONTACT/VECTOR POINT/DEFAULT,POLAR
                  THEO/<15.0682,70.268,0>,<0,0,-1>
                  ACTL/<15.0682,70.268,0>,<0,0,-1>
                  TARG/<15.0682,70.268,0>,<0,0,-1>
                  SNAP=NO
                  SHOW FEATURE PARAMETERS=NO
                  SHOW CONTACT PARAMETERS=YES
                    AVOIDANCE MOVE=BOTH,DISTANCE BEFORE=0.1,DISTANCE AFTER=0.1
                  SHOW HITS=NO
    PNT25        =FEAT/CONTACT/VECTOR POINT/DEFAULT,POLAR
                  THEO/<20.4957,80.2999,0>,<0,0,-1>
                  ACTL/<20.4957,80.2999,0>,<0,0,-1>
                  TARG/<20.4957,80.2999,0>,<0,0,-1>
                  SNAP=NO
                  SHOW FEATURE PARAMETERS=NO
                  SHOW CONTACT PARAMETERS=YES
                    AVOIDANCE MOVE=BOTH,DISTANCE BEFORE=0.1,DISTANCE AFTER=0.1
                  SHOW HITS=NO
                  MOVE/POINT,NORMAL,<-2.5,20,-1>
                  MOVE/POINT,NORMAL,<-2.5,20,8>
                  TIP/T1A-105B170, SHANKIJK=0.1677, 0.9513, -0.2588, ANGLE=34.2656
                  MOVE/POINT,NORMAL,<1,26.5,8>
                  MOVE/POINT,NORMAL,<1,26.5,-0.5>
    PNT48        =FEAT/CONTACT/VECTOR POINT/DEFAULT,POLAR
                  THEO/<18.7324,88.5898,0>,<0,0,-1>
                  ACTL/<18.7324,88.5898,0>,<0,0,-1>
                  TARG/<18.7324,88.5898,0>,<0,0,-1>
                  SNAP=NO
                  SHOW FEATURE PARAMETERS=NO
                  SHOW CONTACT PARAMETERS=YES
                    AVOIDANCE MOVE=BOTH,DISTANCE BEFORE=0.1,DISTANCE AFTER=0.1
                  SHOW HITS=NO
    PNT49        =FEAT/CONTACT/VECTOR POINT/DEFAULT,POLAR
                  THEO/<25.8488,88.9401,0>,<0,0,-1>
                  ACTL/<25.8488,88.9401,0>,<0,0,-1>
                  TARG/<25.8488,88.9401,0>,<0,0,-1>
                  SNAP=NO
                  SHOW FEATURE PARAMETERS=NO
                  SHOW CONTACT PARAMETERS=YES
                    AVOIDANCE MOVE=BOTH,DISTANCE BEFORE=0.1,DISTANCE AFTER=0.1
                  SHOW HITS=NO
    PNT22        =FEAT/CONTACT/VECTOR POINT/DEFAULT,POLAR
                  THEO/<26.0729,84.1607,0>,<0,0,-1>
                  ACTL/<26.0729,84.1607,0>,<0,0,-1>
                  TARG/<26.0729,84.1607,0>,<0,0,-1>
                  SNAP=NO
                  SHOW FEATURE PARAMETERS=NO
                  SHOW CONTACT PARAMETERS=YES
                    AVOIDANCE MOVE=BOTH,DISTANCE BEFORE=0.1,DISTANCE AFTER=0.1
                  SHOW HITS=NO
                  MOVE/POINT,NORMAL,<3,29,-0.5>
                  TIP/T1A-105B180, SHANKIJK=0, 0.9659, -0.2588, ANGLE=0
                  MOVE/POINT,NORMAL,<9.2,29,-0.5>
    PNT26        =FEAT/CONTACT/VECTOR POINT/DEFAULT,POLAR
                  THEO/<27.6436,69.7299,0>,<0,0,-1>
                  ACTL/<27.6436,69.7299,0>,<0,0,-1>
                  TARG/<27.6436,69.7299,0>,<0,0,-1>
                  SNAP=NO
                  SHOW FEATURE PARAMETERS=NO
                  SHOW CONTACT PARAMETERS=YES
                    AVOIDANCE MOVE=BOTH,DISTANCE BEFORE=0.1,DISTANCE AFTER=0.1
                  SHOW HITS=NO
    PNT27        =FEAT/CONTACT/VECTOR POINT/DEFAULT,POLAR
                  THEO/<22.4211,64.2187,0>,<0,0,-1>
                  ACTL/<22.4211,64.2187,0>,<0,0,-1>
                  TARG/<22.4211,64.2187,0>,<0,0,-1>
                  SNAP=NO
                  SHOW FEATURE PARAMETERS=NO
                  SHOW CONTACT PARAMETERS=YES
                    AVOIDANCE MOVE=BOTH,DISTANCE BEFORE=0.1,DISTANCE AFTER=0.1
                  SHOW HITS=NO
                  MOVE/POINT,NORMAL,<9.2,29,-0.5>
                  MOVE/POINT,NORMAL,<18.5,29,-0.5>
    PNT28        =FEAT/CONTACT/VECTOR POINT/DEFAULT,POLAR
                  THEO/<32.2604,53.5417,0>,<0,0,-1>
                  ACTL/<32.2604,53.5417,0>,<0,0,-1>
                  TARG/<32.2604,53.5417,0>,<0,0,-1>
                  SNAP=NO
                  SHOW FEATURE PARAMETERS=NO
                  SHOW CONTACT PARAMETERS=YES
                    AVOIDANCE MOVE=BOTH,DISTANCE BEFORE=0.1,DISTANCE AFTER=0.1
                  SHOW HITS=NO
    PNT29        =FEAT/CONTACT/VECTOR POINT/DEFAULT,POLAR
                  THEO/<27.8,46.5471,0>,<0,0,-1>
                  ACTL/<27.8,46.5471,0>,<0,0,-1>
                  TARG/<27.8,46.5471,0>,<0,0,-1>
                  SNAP=NO
                  SHOW FEATURE PARAMETERS=NO
                  SHOW CONTACT PARAMETERS=YES
                    AVOIDANCE MOVE=BOTH,DISTANCE BEFORE=0.1,DISTANCE AFTER=0.1
                  SHOW HITS=NO
                  MOVE/POINT,NORMAL,<18.5,29,-0.5>
                  MOVE/POINT,NORMAL,<26,29,-0.5>
    PNT30        =FEAT/CONTACT/VECTOR POINT/DEFAULT,POLAR
                  THEO/<36.2712,44.7865,0>,<0,0,-1>
                  ACTL/<36.2712,44.7865,0>,<0,0,-1>
                  TARG/<36.2712,44.7865,0>,<0,0,-1>
                  SNAP=NO
                  SHOW FEATURE PARAMETERS=NO
                  SHOW CONTACT PARAMETERS=YES
                    AVOIDANCE MOVE=BOTH,DISTANCE BEFORE=0.1,DISTANCE AFTER=0.1
                  SHOW HITS=NO
    PNT31        =FEAT/CONTACT/VECTOR POINT/DEFAULT,POLAR
                  THEO/<33.3581,39.5128,0>,<0,0,-1>
                  ACTL/<33.3581,39.5128,0>,<0,0,-1>
                  TARG/<33.3581,39.5128,0>,<0,0,-1>
                  SNAP=NO
                  SHOW FEATURE PARAMETERS=NO
                  SHOW CONTACT PARAMETERS=YES
                    AVOIDANCE MOVE=BOTH,DISTANCE BEFORE=0.1,DISTANCE AFTER=0.1
                  SHOW HITS=NO
                  MOVE/POINT,NORMAL,<26,29,-0.5>
                  MOVE/POINT,NORMAL,<26,29,8>
                  TIP/T1A105B-90, SHANKIJK=0.9659, 0, -0.2588, ANGLE=-90
                  MOVE/POINT,NORMAL,<30,22,8>
                  MOVE/POINT,NORMAL,<30,22,-1>
    PNT32        =FEAT/CONTACT/VECTOR POINT/DEFAULT,POLAR
                  THEO/<31.6628,39.5427,0>,<0,0,-1>
                  ACTL/<31.6628,39.5427,0>,<0,0,-1>
                  TARG/<31.6628,39.5427,0>,<0,0,-1>
                  SNAP=NO
                  SHOW FEATURE PARAMETERS=NO
                  SHOW CONTACT PARAMETERS=YES
                    AVOIDANCE MOVE=BOTH,DISTANCE BEFORE=0.1,DISTANCE AFTER=0.1
                  SHOW HITS=NO
    PNT33        =FEAT/CONTACT/VECTOR POINT/DEFAULT,POLAR
                  THEO/<27.7351,27.9311,0>,<0,0,-1>
                  ACTL/<27.7351,27.9311,0>,<0,0,-1>
                  TARG/<27.7351,27.9311,0>,<0,0,-1>
                  SNAP=NO
                  SHOW FEATURE PARAMETERS=NO
                  SHOW CONTACT PARAMETERS=YES
                    AVOIDANCE MOVE=BOTH,DISTANCE BEFORE=0.1,DISTANCE AFTER=0.1
                  SHOW HITS=NO
    PNT34        =FEAT/CONTACT/VECTOR POINT/DEFAULT,POLAR
                  THEO/<25.2601,13.6339,0>,<0,0,-1>
                  ACTL/<25.2601,13.6339,0>,<0,0,-1>
                  TARG/<25.2601,13.6339,0>,<0,0,-1>
                  SNAP=NO
                  SHOW FEATURE PARAMETERS=NO
                  SHOW CONTACT PARAMETERS=YES
                    AVOIDANCE MOVE=BOTH,DISTANCE BEFORE=0.1,DISTANCE AFTER=0.1
                  SHOW HITS=NO
                ENDGROUP/ID=PLNA VECTOR POINTS
    PLNA       =FEAT/PLANE,CARTESIAN,OUTLINE,NO,LEAST_SQR
                THEO/<12.4556,12.9766,0>,<0,0,-1>
                ACTL/<12.4556,12.9766,0>,<0,0,-1>
                CONSTR/PLANE,BF,PNT14,PNT15,PNT16,PNT17,PNT18,PNT19,PNT20,PNT21,PNT23,PNT24,PNT25,PNT22,PNT26,PNT27,PNT28,PNT29,PNT30,PNT31,PNT32,PNT33,PNT34,PNT49,PNT48,PNT46,PNT47,,
                OUTLIER_REMOVAL/OFF,3
                FILTER/OFF,WAVELENGTH=0
                MOVE/POINT,NORMAL,<30,7,-1>
                MOVE/POINT,NORMAL,<30,7,8>
                TIP/T1A10B90, SHANKIJK=-0.1736, 0, 0.9848, ANGLE=90
                MOVE/POINT,NORMAL,<30,7,8>
                MOVE/POINT,NORMAL,<0,7,8>
    PLNB VECTOR POINTS=GROUP/SHOWALLPARAMS=YES
    PNT35        =FEAT/CONTACT/VECTOR POINT/DEFAULT,POLAR
                  THEO/<0.1243,90,0.6971>,<-1,0,0>
                  ACTL/<0.1243,90,0.6971>,<-1,0,0>
                  TARG/<0.1243,90,0.6971>,<-1,0,0>
                  SNAP=NO
                  SHOW FEATURE PARAMETERS=NO
                  SHOW CONTACT PARAMETERS=YES
                    AVOIDANCE MOVE=BOTH,DISTANCE BEFORE=0.1,DISTANCE AFTER=0.1
                  SHOW HITS=NO
    PNT36        =FEAT/CONTACT/VECTOR POINT/DEFAULT,POLAR
                  THEO/<0.1404,90,0.0641>,<-1,0,0>
                  ACTL/<0.1404,90,0.0641>,<-1,0,0>
                  TARG/<0.1404,90,0.0641>,<-1,0,0>
                  SNAP=NO
                  SHOW FEATURE PARAMETERS=NO
                  SHOW CONTACT PARAMETERS=YES
                    AVOIDANCE MOVE=BOTH,DISTANCE BEFORE=0.1,DISTANCE AFTER=0.1
                  SHOW HITS=NO
                  MOVE/POINT,NORMAL,<-0.2295,0.1404,0.7>
                  MOVE/POINT,NORMAL,<-0.2295,2.5,0.7>
    PNT37        =FEAT/CONTACT/VECTOR POINT/DEFAULT,POLAR
                  THEO/<3.8476,90,0.0427>,<-1,0,0>
                  ACTL/<3.8476,90,0.0427>,<-1,0,0>
                  TARG/<3.8476,90,0.0427>,<-1,0,0>
                  SNAP=NO
                  SHOW FEATURE PARAMETERS=NO
                  SHOW CONTACT PARAMETERS=YES
                    AVOIDANCE MOVE=BOTH,DISTANCE BEFORE=0.1,DISTANCE AFTER=0.1
                  SHOW HITS=NO
    PNT38        =FEAT/CONTACT/VECTOR POINT/DEFAULT,POLAR
                  THEO/<7.018,90,0.0324>,<-1,0,0>
                  ACTL/<7.018,90,0.0324>,<-1,0,0>
                  TARG/<7.018,90,0.0324>,<-1,0,0>
                  SNAP=NO
                  SHOW FEATURE PARAMETERS=NO
                  SHOW CONTACT PARAMETERS=YES
                    AVOIDANCE MOVE=BOTH,DISTANCE BEFORE=0.1,DISTANCE AFTER=0.1
                  SHOW HITS=NO
    PNT39        =FEAT/CONTACT/VECTOR POINT/DEFAULT,POLAR
                  THEO/<10.6898,90,0.0332>,<-1,0,0>
                  ACTL/<10.6898,90,0.0332>,<-1,0,0>
                  TARG/<10.6898,90,0.0332>,<-1,0,0>
                  SNAP=NO
                  SHOW FEATURE PARAMETERS=NO
                  SHOW CONTACT PARAMETERS=YES
                    AVOIDANCE MOVE=BOTH,DISTANCE BEFORE=0.1,DISTANCE AFTER=0.1
                  SHOW HITS=NO
                  MOVE/POINT,NORMAL,<-0.2295,10.6898,0.7>
                  MOVE/POINT,NORMAL,<-0.2295,14,0.7>
    PNT40        =FEAT/CONTACT/VECTOR POINT/DEFAULT,POLAR
                  THEO/<14.3484,90,0.0478>,<-1,0,0>
                  ACTL/<14.3484,90,0.0478>,<-1,0,0>
                  TARG/<14.3484,90,0.0478>,<-1,0,0>
                  SNAP=NO
                  SHOW FEATURE PARAMETERS=NO
                  SHOW CONTACT PARAMETERS=YES
                    AVOIDANCE MOVE=BOTH,DISTANCE BEFORE=0.1,DISTANCE AFTER=0.1
                  SHOW HITS=NO
    PNT41        =FEAT/CONTACT/VECTOR POINT/DEFAULT,POLAR
                  THEO/<17.8816,90,0.0405>,<-1,0,0>
                  ACTL/<17.8816,90,0.0405>,<-1,0,0>
                  TARG/<17.8816,90,0.0405>,<-1,0,0>
                  SNAP=NO
                  SHOW FEATURE PARAMETERS=NO
                  SHOW CONTACT PARAMETERS=YES
                    AVOIDANCE MOVE=BOTH,DISTANCE BEFORE=0.1,DISTANCE AFTER=0.1
                  SHOW HITS=NO
    PNT42        =FEAT/CONTACT/VECTOR POINT/DEFAULT,POLAR
                  THEO/<21.1727,90,0.0405>,<-1,0,0>
                  ACTL/<21.1727,90,0.0405>,<-1,0,0>
                  TARG/<21.1727,90,0.0405>,<-1,0,0>
                  SNAP=NO
                  SHOW FEATURE PARAMETERS=NO
                  SHOW CONTACT PARAMETERS=YES
                    AVOIDANCE MOVE=BOTH,DISTANCE BEFORE=0.1,DISTANCE AFTER=0.1
                  SHOW HITS=NO
                  MOVE/POINT,NORMAL,<-0.2295,21.1727,0.3>
                  MOVE/POINT,NORMAL,<-0.2295,24.25,0.3>
    PNT43        =FEAT/CONTACT/VECTOR POINT/DEFAULT,POLAR
                  THEO/<24.2576,90,0.047>,<-1,0,0>
                  ACTL/<24.2576,90,0.047>,<-1,0,0>
                  TARG/<24.2576,90,0.047>,<-1,0,0>
                  SNAP=NO
                  SHOW FEATURE PARAMETERS=NO
                  SHOW CONTACT PARAMETERS=YES
                    AVOIDANCE MOVE=BOTH,DISTANCE BEFORE=0.1,DISTANCE AFTER=0.1
                  SHOW HITS=NO
                  MOVE/POINT,NORMAL,<-0.2295,24.25,0.7>
                  MOVE/POINT,NORMAL,<-0.2295,26,0.8>
    PNT44        =FEAT/CONTACT/VECTOR POINT/DEFAULT,POLAR
                  THEO/<26.0264,90,0.729>,<-1,0,0>
                  ACTL/<26.0264,90,0.729>,<-1,0,0>
                  TARG/<26.0264,90,0.729>,<-1,0,0>
                  SNAP=NO
                  SHOW FEATURE PARAMETERS=NO
                  SHOW CONTACT PARAMETERS=YES
                    AVOIDANCE MOVE=BOTH,DISTANCE BEFORE=0.1,DISTANCE AFTER=0.1
                  SHOW HITS=NO
    PNT45        =FEAT/CONTACT/VECTOR POINT/DEFAULT,POLAR
                  THEO/<26,90,0.1275>,<-1,0,0>
                  ACTL/<26,90,0.1275>,<-1,0,0>
                  TARG/<26,90,0.1275>,<-1,0,0>
                  SNAP=NO
                  SHOW FEATURE PARAMETERS=NO
                  SHOW CONTACT PARAMETERS=YES
                    AVOIDANCE MOVE=BOTH,DISTANCE BEFORE=0.1,DISTANCE AFTER=0.1
                  SHOW HITS=NO
                ENDGROUP/ID=PLNB VECTOR POINTS
    PLNB       =FEAT/PLANE,CARTESIAN,OUTLINE,NO,LEAST_SQR
                THEO/<0,13.7733,0.1729>,<-1,0,0>
                ACTL/<0,13.7733,0.1729>,<-1,0,0>
                CONSTR/PLANE,BF,PNT35,PNT36,PNT37,PNT38,PNT39,PNT40,PNT41,PNT42,PNT43,PNT44,PNT45,,
                OUTLIER_REMOVAL/OFF,3
                FILTER/OFF,WAVELENGTH=0
    A2         =ALIGNMENT/START,RECALL:STARTUP,LIST=YES
                  ALIGNMENT/LEVEL,ZMINUS,PLNA
                  ALIGNMENT/TRANS,ZAXIS,PLNA
                  ALIGNMENT/ROTATE,XMINUS,TO,PLNB,ABOUT,ZMINUS
                  ALIGNMENT/TRANS,XAXIS,PLNB
                  ALIGNMENT/TRANS,YAXIS,PLNC
                ALIGNMENT/END
                MOVE/POINT,NORMAL,<-0.2295,25.7,6>
                TIP/T1A-10B0, SHANKIJK=0, -0.1736, 0.9848, ANGLE=0
    PLN1       =FEAT/CONTACT/PLANE/DEFAULT,CARTESIAN,OUTLINE,LEAST_SQR
                THEO/<0.6897,25.9023,0.7596>,<0,-1,0>
                ACTL/<0.6897,25.9023,0.7596>,<0,-1,0>
                TARG/<0.6897,25.9023,0.7596>,<0,-1,0>
                ANGLE VEC=<1,0,0>,SQUARE
                SHOW FEATURE PARAMETERS=NO
                SHOW CONTACT PARAMETERS=YES
                  NUMHITS=2,NUMROWS=2
                  SPACER=0
                  AVOIDANCE MOVE=NO
                  USE BOUNDARY OFFSET=YES,OFFSET=0.0394
                SHOW HITS=NO
                MOVE/POINT,NORMAL,<0.1564,25.6728,2>
                TIP/T1A10B90, SHANKIJK=-0.1736, 0, 0.9848, ANGLE=90
                MOVE/POINT,NORMAL,<0.1564,22,2>
    PLN2       =FEAT/CONTACT/PLANE/DEFAULT,CARTESIAN,OUTLINE,LEAST_SQR
                THEO/<1.1787,20.4557,0.7326>,<-1,0,0>
                ACTL/<1.1787,20.4557,0.7326>,<-1,0,0>
                TARG/<1.1787,20.4557,0.7326>,<-1,0,0>
                ANGLE VEC=<0,0,-1>,SQUARE
                SHOW FEATURE PARAMETERS=NO
                SHOW CONTACT PARAMETERS=YES
                  NUMHITS=10,NUMROWS=2
                  SPACER=0
                  AVOIDANCE MOVE=NO
                SHOW HITS=NO
                MOVE/POINT,NORMAL,<0.9492,0.4475,2.5>
                TIP/T1A10B0, SHANKIJK=0, 0.1736, 0.9848, ANGLE=0
                MOVE/POINT,NORMAL,<0.9492,0.4475,1.5>
    PLN3       =FEAT/CONTACT/PLANE/DEFAULT,CARTESIAN,OUTLINE,LEAST_SQR
                THEO/<0.8811,0.2392,0.6575>,<0,1,0>
                ACTL/<0.8811,0.2392,0.6575>,<0,1,0>
                TARG/<0.8811,0.2392,0.6575>,<0,1,0>
                ANGLE VEC=<-1,0,0>,SQUARE
                SHOW FEATURE PARAMETERS=NO
                SHOW CONTACT PARAMETERS=YES
                  NUMHITS=2,NUMROWS=2
                  SPACER=0
                  AVOIDANCE MOVE=NO
                  USE BOUNDARY OFFSET=YES,OFFSET=0.0394
                SHOW HITS=NO
                TIP/T1A0B0, SHANKIJK=0, 0, 1, ANGLE=0
    PLN4       =FEAT/CONTACT/PLANE/DEFAULT,CARTESIAN,OUTLINE,LEAST_SQR
                THEO/<0.8629,20.8596,0.08>,<0,0,1>
                ACTL/<0.8629,20.8596,0.08>,<0,0,1>
                TARG/<0.8629,20.8596,0.08>,<0,0,1>
                ANGLE VEC=<1,0,0>,SQUARE
                SHOW FEATURE PARAMETERS=NO
                SHOW CONTACT PARAMETERS=YES
                  NUMHITS=5,NUMROWS=2
                  SPACER=0
                  AVOIDANCE MOVE=NO
                  USE BOUNDARY OFFSET=YES,OFFSET=0.0394
                SHOW HITS=NO
                CLEARP/ZPLUS,2,ZPLUS,2,ON
                MOVE/CLEARPLANE
                TIP/T1A10B0, SHANKIJK=0, 0.1736, 0.9848, ANGLE=0
                CLEARP/ZPLUS,2,ZPLUS,2,OFF
    PLN5       =FEAT/PLANE,CARTESIAN,OUTLINE
                THEO/<13.9839,0.06,0.8422>,<0,1,0>
                ACTL/<13.9839,0.06,0.8422>,<0,1,0>
                MEAS/PLANE,13
                  HIT/BASIC,NORMAL,<1.5561,0.06,1.0734>,<0,1,0>,<1.5561,0.06,1.0734>,USE THEO=YES
                  HIT/BASIC,NORMAL,<1.5567,0.06,0.7791>,<0,1,0>,<1.5567,0.06,0.7791>,USE THEO=YES
                  HIT/BASIC,NORMAL,<4.1699,0.06,1.0714>,<0,1,0>,<4.1699,0.06,1.0714>,USE THEO=YES
                  HIT/BASIC,NORMAL,<6.5737,0.06,0.7859>,<0,1,0>,<6.5737,0.06,0.7859>,USE THEO=YES
                  HIT/BASIC,NORMAL,<9.4337,0.06,1.0559>,<0,1,0>,<9.4337,0.06,1.0559>,USE THEO=YES
                  HIT/BASIC,NORMAL,<12.7182,0.06,0.773>,<0,1,0>,<12.7182,0.06,0.773>,USE THEO=YES
                  MOVE/POINT,NORMAL,<13.5355,0.2881,1.6>
                  MOVE/POINT,NORMAL,<14.113,0.2905,1.6>
                  HIT/BASIC,NORMAL,<14.6912,0.06,1.0806>,<0,1,0>,<14.6912,0.06,1.0806>,USE THEO=YES
                  HIT/BASIC,NORMAL,<16.7152,0.06,0.7917>,<0,1,0>,<16.7152,0.06,0.7917>,USE THEO=YES
                  HIT/BASIC,NORMAL,<19.2592,0.06,1.0571>,<0,1,0>,<19.2592,0.06,1.0571>,USE THEO=YES
                  HIT/BASIC,NORMAL,<21.5622,0.06,0.7867>,<0,1,0>,<21.5622,0.06,0.7867>,USE THEO=YES
                  HIT/BASIC,NORMAL,<23.2716,0.06,0.9999>,<0,1,0>,<23.2716,0.06,0.9999>,USE THEO=YES
                  HIT/BASIC,NORMAL,<25.8983,0.06,0.4209>,<0,1,0>,<25.8983,0.06,0.4209>,USE THEO=YES
                  HIT/BASIC,NORMAL,<24.3851,0.06,0.2735>,<0,1,0>,<24.3851,0.06,0.2735>,USE THEO=YES
                ENDMEAS/
    PLN6       =FEAT/PLANE,CARTESIAN,OUTLINE
                THEO/<12.9814,0.2555,0.3049>,<0,1,0>
                ACTL/<12.9814,0.2555,0.3049>,<0,1,0>
                MEAS/PLANE,10
                  HIT/BASIC,NORMAL,<24.207,0.2555,0.0577>,<0,1,0>,<24.207,0.2555,0.0577>,USE THEO=YES
                  HIT/BASIC,NORMAL,<24.0441,0.2555,0.5387>,<0,1,0>,<24.0441,0.2555,0.5387>,USE THEO=YES
                  HIT/BASIC,NORMAL,<20.6758,0.2555,0.0336>,<0,1,0>,<20.6758,0.2555,0.0336>,USE THEO=YES
                  HIT/BASIC,NORMAL,<17.6988,0.2555,0.5628>,<0,1,0>,<17.6988,0.2555,0.5628>,USE THEO=YES
                  HIT/BASIC,NORMAL,<14.5183,0.2555,0.0537>,<0,1,0>,<14.5183,0.2555,0.0537>,USE THEO=YES
                  MOVE/POINT,NORMAL,<14.2641,2.5,0.0943>
                  MOVE/POINT,NORMAL,<13.4276,2.5,0.2278>
                  HIT/BASIC,NORMAL,<11.1915,0.2555,0.578>,<0,1,0>,<11.1915,0.2555,0.578>,USE THEO=YES
                  HIT/BASIC,NORMAL,<8.5331,0.2555,0.0652>,<0,1,0>,<8.5331,0.2555,0.0652>,USE THEO=YES
                  HIT/BASIC,NORMAL,<5.7747,0.2555,0.5504>,<0,1,0>,<5.7747,0.2555,0.5504>,USE THEO=YES
                  HIT/BASIC,NORMAL,<1.5816,0.2555,0.0471>,<0,1,0>,<1.5816,0.2555,0.0471>,USE THEO=YES
                  HIT/BASIC,NORMAL,<1.5886,0.2555,0.5622>,<0,1,0>,<1.5886,0.2555,0.5622>,USE THEO=YES
                ENDMEAS/
                CLEARP/ZPLUS,2,ZPLUS,2,ON
                MOVE/CLEARPLANE
                TIP/T1A-10B90, SHANKIJK=0.1736, 0, 0.9848, ANGLE=90
    PLN7       =FEAT/PLANE,CARTESIAN,OUTLINE
                THEO/<1.2587,2.996,0.7179>,<1,0,0>
                ACTL/<1.2587,2.996,0.7179>,<1,0,0>
                MEAS/PLANE,5
                  MOVE/CLEARPLANE
                  HIT/BASIC,NORMAL,<1.2587,0.412,0.9745>,<1,0,0>,<1.2587,0.412,0.9745>,USE THEO=YES
                  HIT/BASIC,NORMAL,<1.2587,0.5985,0.0411>,<1,0,0>,<1.2587,0.5985,0.0411>,USE THEO=YES
                  HIT/BASIC,NORMAL,<1.2587,2.9004,1.0094>,<1,0,0>,<1.2587,2.9004,1.0094>,USE THEO=YES
                  HIT/BASIC,NORMAL,<1.2587,5.5169,0.0382>,<1,0,0>,<1.2587,5.5169,0.0382>,USE THEO=YES
                  HIT/BASIC,NORMAL,<1.2587,5.5521,1.5262>,<1,0,0>,<1.2587,5.5521,1.5262>,USE THEO=YES
                ENDMEAS/
                MOVE/CLEARPLANE
                TIP/T1A-10B0, SHANKIJK=0, -0.1736, 0.9848, ANGLE=0
    PLN8       =FEAT/PLANE,CARTESIAN,OUTLINE
                THEO/<13.1703,5.8785,0.8847>,<0,-1,0>
                ACTL/<13.1703,5.8785,0.8847>,<0,-1,0>
                MEAS/PLANE,9
                  MOVE/CLEARPLANE
                  HIT/BASIC,NORMAL,<1.5923,5.8785,1.5918>,<0,-1,0>,<1.5923,5.8785,1.5918>,USE THEO=YES
                  HIT/BASIC,NORMAL,<1.5894,5.8785,0.0599>,<0,-1,0>,<1.5894,5.8785,0.0599>,USE THEO=YES
                  HIT/BASIC,NORMAL,<5.9449,5.8785,1.5151>,<0,-1,0>,<5.9449,5.8785,1.5151>,USE THEO=YES
                  HIT/BASIC,NORMAL,<9.6761,5.8785,0.0892>,<0,-1,0>,<9.6761,5.8785,0.0892>,USE THEO=YES
                  HIT/BASIC,NORMAL,<13.4655,5.8785,1.553>,<0,-1,0>,<13.4655,5.8785,1.553>,USE THEO=YES
                  HIT/BASIC,NORMAL,<17.4277,5.8785,0.1281>,<0,-1,0>,<17.4277,5.8785,0.1281>,USE THEO=YES
                  HIT/BASIC,NORMAL,<20.6084,5.8785,1.4731>,<0,-1,0>,<20.6084,5.8785,1.4731>,USE THEO=YES
                  HIT/BASIC,NORMAL,<24.1339,5.8785,0.0805>,<0,-1,0>,<24.1339,5.8785,0.0805>,USE THEO=YES
                  HIT/BASIC,NORMAL,<24.095,5.8785,1.4715>,<0,-1,0>,<24.095,5.8785,1.4715>,USE THEO=YES
                ENDMEAS/
                MOVE/CLEARPLANE
                TIP/T1A10B90, SHANKIJK=-0.1736, 0, 0.9848, ANGLE=90
    PLN9       =FEAT/PLANE,CARTESIAN,OUTLINE
                THEO/<24.5011,3.1412,0.549>,<-1,0,0>
                ACTL/<24.5011,3.1412,0.549>,<-1,0,0>
                MEAS/PLANE,6
                  MOVE/CLEARPLANE
                  HIT/BASIC,NORMAL,<24.5011,0.5777,0.0447>,<-1,0,0>,<24.5011,0.5777,0.0447>,USE THEO=YES
                  HIT/BASIC,NORMAL,<24.5011,1.8038,0.0874>,<-1,0,0>,<24.5011,1.8038,0.0874>,USE THEO=YES
                  HIT/BASIC,NORMAL,<24.5011,1.8989,0.7944>,<-1,0,0>,<24.5011,1.8989,0.7944>,USE THEO=YES
                  HIT/BASIC,NORMAL,<24.5011,3.5333,0.8453>,<-1,0,0>,<24.5011,3.5333,0.8453>,USE THEO=YES
                  HIT/BASIC,NORMAL,<24.5011,5.5162,0.0926>,<-1,0,0>,<24.5011,5.5162,0.0926>,USE THEO=YES
                  HIT/BASIC,NORMAL,<24.5011,5.5174,1.4293>,<-1,0,0>,<24.5011,5.5174,1.4293>,USE THEO=YES
                ENDMEAS/
                MOVE/CLEARPLANE
    PLN10      =FEAT/PLANE,CARTESIAN,OUTLINE
                THEO/<24.5011,8.8477,0.6591>,<-1,0,0>
                ACTL/<24.5011,8.8477,0.6591>,<-1,0,0>
                MEAS/PLANE,5
                  MOVE/CLEARPLANE
                  HIT/BASIC,NORMAL,<24.5011,6.3807,1.4511>,<-1,0,0>,<24.5011,6.3807,1.4511>,USE THEO=YES
                  HIT/BASIC,NORMAL,<24.5011,6.3797,0.064>,<-1,0,0>,<24.5011,6.3797,0.064>,USE THEO=YES
                  HIT/BASIC,NORMAL,<24.5011,8.8234,0.8746>,<-1,0,0>,<24.5011,8.8234,0.8746>,USE THEO=YES
                  HIT/BASIC,NORMAL,<24.5011,11.3283,0.0624>,<-1,0,0>,<24.5011,11.3283,0.0624>,USE THEO=YES
                  HIT/BASIC,NORMAL,<24.5011,11.3265,0.8435>,<-1,0,0>,<24.5011,11.3265,0.8435>,USE THEO=YES
                ENDMEAS/
                MOVE/CLEARPLANE
                TIP/T1A-10B0, SHANKIJK=0, -0.1736, 0.9848, ANGLE=0
    PLN11      =FEAT/PLANE,CARTESIAN,OUTLINE
                THEO/<13.2911,11.6575,0.537>,<0,-1,0>
                ACTL/<13.2911,11.6575,0.537>,<0,-1,0>
                MEAS/PLANE,10
                  MOVE/CLEARPLANE
                  HIT/BASIC,NORMAL,<24.106,11.6575,0.8477>,<0,-1,0>,<24.106,11.6575,0.8477>,USE THEO=YES
                  HIT/BASIC,NORMAL,<24.146,11.6575,0.07>,<0,-1,0>,<24.146,11.6575,0.07>,USE THEO=YES
                  HIT/BASIC,NORMAL,<21.0701,11.6575,0.8943>,<0,-1,0>,<21.0701,11.6575,0.8943>,USE THEO=YES
                  HIT/BASIC,NORMAL,<17.4338,11.6575,0.0463>,<0,-1,0>,<17.4338,11.6575,0.0463>,USE THEO=YES
                  HIT/BASIC,NORMAL,<14.6558,11.6575,0.8286>,<0,-1,0>,<14.6558,11.6575,0.8286>,USE THEO=YES
                  MOVE/POINT,NORMAL,<14.1764,11.4274,1.6>
                  MOVE/POINT,NORMAL,<13.4367,11.4289,1.6>
                  HIT/BASIC,NORMAL,<12.5433,11.6575,0.7964>,<0,-1,0>,<12.5433,11.6575,0.7964>,USE THEO=YES
                  HIT/BASIC,NORMAL,<9.7732,11.6575,0.0558>,<0,-1,0>,<9.7732,11.6575,0.0558>,USE THEO=YES
                  HIT/BASIC,NORMAL,<6.0335,11.6575,0.8929>,<0,-1,0>,<6.0335,11.6575,0.8929>,USE THEO=YES
                  HIT/BASIC,NORMAL,<1.5897,11.6575,0.0359>,<0,-1,0>,<1.5897,11.6575,0.0359>,USE THEO=YES
                  HIT/BASIC,NORMAL,<1.5599,11.6575,0.9023>,<0,-1,0>,<1.5599,11.6575,0.9023>,USE THEO=YES
                ENDMEAS/
                MOVE/CLEARPLANE
                TIP/T1A-10B90, SHANKIJK=0.1736, 0, 0.9848, ANGLE=90
    PLN12      =FEAT/PLANE,CARTESIAN,OUTLINE
                THEO/<1.2587,13.0418,1.0384>,<1,0,0>
                ACTL/<1.2587,13.0418,1.0384>,<1,0,0>
                MEAS/PLANE,16
                  MOVE/CLEARPLANE
                  HIT/BASIC,NORMAL,<1.2587,6.3779,0.0892>,<1,0,0>,<1.2587,6.3779,0.0892>,USE THEO=YES
                  MOVE/POINT,NORMAL,<1.4881,6.4443,1.6673>
                  HIT/BASIC,NORMAL,<1.2587,5.8918,1.8568>,<1,0,0>,<1.2587,5.8918,1.8568>,USE THEO=YES
                  HIT/BASIC,NORMAL,<1.2587,8.837,1.8622>,<1,0,0>,<1.2587,8.837,1.8622>,USE THEO=YES
                  HIT/BASIC,NORMAL,<1.2587,11.2453,0.0508>,<1,0,0>,<1.2587,11.2453,0.0508>,USE THEO=YES
                  HIT/BASIC,NORMAL,<1.2587,10.5824,1.8171>,<1,0,0>,<1.2587,10.5824,1.8171>,USE THEO=YES
                  HIT/BASIC,NORMAL,<1.2587,11.3129,1.1795>,<1,0,0>,<1.2587,11.3129,1.1795>,USE THEO=YES
                  HIT/BASIC,NORMAL,<1.2587,12.0809,1.1454>,<1,0,0>,<1.2587,12.0809,1.1454>,USE THEO=YES
                  HIT/BASIC,NORMAL,<1.2587,12.0427,0.2453>,<1,0,0>,<1.2587,12.0427,0.2453>,USE THEO=YES
                  HIT/BASIC,NORMAL,<1.2587,14.1034,0.2349>,<1,0,0>,<1.2587,14.1034,0.2349>,USE THEO=YES
                  HIT/BASIC,NORMAL,<1.2587,14.1038,1.1884>,<1,0,0>,<1.2587,14.1038,1.1884>,USE THEO=YES
                  HIT/BASIC,NORMAL,<1.2587,14.8466,1.191>,<1,0,0>,<1.2587,14.8466,1.191>,USE THEO=YES
                  HIT/BASIC,NORMAL,<1.2587,15.2951,1.8597>,<1,0,0>,<1.2587,15.2951,1.8597>,USE THEO=YES
                  HIT/BASIC,NORMAL,<1.2587,14.8505,0.0881>,<1,0,0>,<1.2587,14.8505,0.0881>,USE THEO=YES
                  HIT/BASIC,NORMAL,<1.2587,19.7325,0.0845>,<1,0,0>,<1.2587,19.7325,0.0845>,USE THEO=YES
                  HIT/BASIC,NORMAL,<1.2587,17.2024,1.8304>,<1,0,0>,<1.2587,17.2024,1.8304>,USE THEO=YES
                  HIT/BASIC,NORMAL,<1.2587,20.1639,1.8904>,<1,0,0>,<1.2587,20.1639,1.8904>,USE THEO=YES
                ENDMEAS/
                MOVE/CLEARPLANE
                TIP/T1A10B0, SHANKIJK=0, 0.1736, 0.9848, ANGLE=0
    PLN13      =FEAT/PLANE,CARTESIAN,OUTLINE
                THEO/<12.8531,6.0345,0.9055>,<0,1,0>
                ACTL/<12.8531,6.0345,0.9055>,<0,1,0>
                MEAS/PLANE,9
                  MOVE/CLEARPLANE
                  HIT/BASIC,NORMAL,<1.6116,6.0345,1.6173>,<0,1,0>,<1.6116,6.0345,1.6173>,USE THEO=YES
                  HIT/BASIC,NORMAL,<1.5683,6.0345,0.0354>,<0,1,0>,<1.5683,6.0345,0.0354>,USE THEO=YES
                  HIT/BASIC,NORMAL,<5.706,6.0345,1.5884>,<0,1,0>,<5.706,6.0345,1.5884>,USE THEO=YES
                  HIT/BASIC,NORMAL,<9.2203,6.0345,0.0612>,<0,1,0>,<9.2203,6.0345,0.0612>,USE THEO=YES
                  HIT/BASIC,NORMAL,<13.1503,6.0345,1.5288>,<0,1,0>,<13.1503,6.0345,1.5288>,USE THEO=YES
                  HIT/BASIC,NORMAL,<15.9789,6.0345,0.0762>,<0,1,0>,<15.9789,6.0345,0.0762>,USE THEO=YES
                  HIT/BASIC,NORMAL,<20.0932,6.0345,1.6039>,<0,1,0>,<20.0932,6.0345,1.6039>,USE THEO=YES
                  HIT/BASIC,NORMAL,<24.1882,6.0345,0.0531>,<0,1,0>,<24.1882,6.0345,0.0531>,USE THEO=YES
                  HIT/BASIC,NORMAL,<24.1611,6.0345,1.5855>,<0,1,0>,<24.1611,6.0345,1.5855>,USE THEO=YES
                ENDMEAS/
                MOVE/CLEARPLANE
                TIP/T1A-10B0, SHANKIJK=0, -0.1736, 0.9848, ANGLE=0
    PLN14      =FEAT/PLANE,CARTESIAN,OUTLINE
                THEO/<18.769,11.7075,1.4741>,<0,-1,0>
                ACTL/<18.769,11.7075,1.4741>,<0,-1,0>
                MEAS/PLANE,4
                  MOVE/CLEARPLANE
                  HIT/BASIC,NORMAL,<19.4595,11.7075,1.8501>,<0,-1,0>,<19.4595,11.7075,1.8501>,USE THEO=YES
                  HIT/BASIC,NORMAL,<19.7268,11.7075,1.0961>,<0,-1,0>,<19.7268,11.7075,1.0961>,USE THEO=YES
                  HIT/BASIC,NORMAL,<17.7847,11.7075,1.0994>,<0,-1,0>,<17.7847,11.7075,1.0994>,USE THEO=YES
                  HIT/BASIC,NORMAL,<18.1049,11.7075,1.8506>,<0,-1,0>,<18.1049,11.7075,1.8506>,USE THEO=YES
                ENDMEAS/
    PLN15      =FEAT/PLANE,CARTESIAN,OUTLINE
                THEO/<9.599,14.4239,0.5968>,<0,-1,0>
                ACTL/<9.599,14.4239,0.5968>,<0,-1,0>
                MEAS/PLANE,9
                  MOVE/CLEARPLANE
                  HIT/BASIC,NORMAL,<1.6014,14.4239,0.8681>,<0,-1,0>,<1.6014,14.4239,0.8681>,USE THEO=YES
                  HIT/BASIC,NORMAL,<1.556,14.4239,0.224>,<0,-1,0>,<1.556,14.4239,0.224>,USE THEO=YES
                  HIT/BASIC,NORMAL,<4.7056,14.4239,0.9101>,<0,-1,0>,<4.7056,14.4239,0.9101>,USE THEO=YES
                  HIT/BASIC,NORMAL,<7.0125,14.4239,0.2068>,<0,-1,0>,<7.0125,14.4239,0.2068>,USE THEO=YES
                  HIT/BASIC,NORMAL,<9.9855,14.4239,0.9032>,<0,-1,0>,<9.9855,14.4239,0.9032>,USE THEO=YES
                  HIT/BASIC,NORMAL,<12.4802,14.4239,0.2332>,<0,-1,0>,<12.4802,14.4239,0.2332>,USE THEO=YES
                  HIT/BASIC,NORMAL,<14.7063,14.4239,0.9066>,<0,-1,0>,<14.7063,14.4239,0.9066>,USE THEO=YES
                  HIT/BASIC,NORMAL,<17.1815,14.4239,0.2064>,<0,-1,0>,<17.1815,14.4239,0.2064>,USE THEO=YES
                  HIT/BASIC,NORMAL,<17.162,14.4239,0.9127>,<0,-1,0>,<17.162,14.4239,0.9127>,USE THEO=YES
                ENDMEAS/
    PLN16      =FEAT/PLANE,CARTESIAN,OUTLINE
                THEO/<18.7481,14.3089,0.9878>,<0,-1,0>
                ACTL/<18.7481,14.3089,0.9878>,<0,-1,0>
                MEAS/PLANE,6
                  MOVE/CLEARPLANE
                  HIT/BASIC,NORMAL,<17.5686,14.3089,0.2306>,<0,-1,0>,<17.5686,14.3089,0.2306>,USE THEO=YES
                  HIT/BASIC,NORMAL,<17.5715,14.3089,0.8884>,<0,-1,0>,<17.5715,14.3089,0.8884>,USE THEO=YES
                  HIT/BASIC,NORMAL,<18.065,14.3089,1.8258>,<0,-1,0>,<18.065,14.3089,1.8258>,USE THEO=YES
                  HIT/BASIC,NORMAL,<19.4557,14.3089,1.8378>,<0,-1,0>,<19.4557,14.3089,1.8378>,USE THEO=YES
                  HIT/BASIC,NORMAL,<19.9185,14.3089,0.9087>,<0,-1,0>,<19.9185,14.3089,0.9087>,USE THEO=YES
                  HIT/BASIC,NORMAL,<19.9091,14.3089,0.2354>,<0,-1,0>,<19.9091,14.3089,0.2354>,USE THEO=YES
                ENDMEAS/
    PLN17      =FEAT/PLANE,CARTESIAN,OUTLINE
                THEO/<22.2386,14.4239,0.4942>,<0,-1,0>
                ACTL/<22.2386,14.4239,0.4942>,<0,-1,0>
                MEAS/PLANE,5
                  MOVE/CLEARPLANE
                  HIT/BASIC,NORMAL,<20.359,14.4239,0.2348>,<0,-1,0>,<20.359,14.4239,0.2348>,USE THEO=YES
                  HIT/BASIC,NORMAL,<20.3547,14.4239,0.8777>,<0,-1,0>,<20.3547,14.4239,0.8777>,USE THEO=YES
                  HIT/BASIC,NORMAL,<22.0933,14.4239,0.2485>,<0,-1,0>,<22.0933,14.4239,0.2485>,USE THEO=YES
                  HIT/BASIC,NORMAL,<24.1859,14.4239,0.9025>,<0,-1,0>,<24.1859,14.4239,0.9025>,USE THEO=YES
                  HIT/BASIC,NORMAL,<24.2003,14.4239,0.2076>,<0,-1,0>,<24.2003,14.4239,0.2076>,USE THEO=YES
                ENDMEAS/
                MOVE/CLEARPLANE
                TIP/T1A10B90, SHANKIJK=-0.1736, 0, 0.9848, ANGLE=90
    PLN18      =FEAT/PLANE,CARTESIAN,OUTLINE
                THEO/<24.5011,13.0878,0.5644>,<-1,0,0>
                ACTL/<24.5011,13.0878,0.5644>,<-1,0,0>
                MEAS/PLANE,4
                  MOVE/CLEARPLANE
                  HIT/BASIC,NORMAL,<24.5011,14.1269,0.8881>,<-1,0,0>,<24.5011,14.1269,0.8881>,USE THEO=YES
                  HIT/BASIC,NORMAL,<24.5011,14.1342,0.236>,<-1,0,0>,<24.5011,14.1342,0.236>,USE THEO=YES
                  HIT/BASIC,NORMAL,<24.5011,12.038,0.8964>,<-1,0,0>,<24.5011,12.038,0.8964>,USE THEO=YES
                  HIT/BASIC,NORMAL,<24.5011,12.052,0.2372>,<-1,0,0>,<24.5011,12.052,0.2372>,USE THEO=YES
                ENDMEAS/
                MOVE/CLEARPLANE
                TIP/T1A10B0, SHANKIJK=0, 0.1736, 0.9848, ANGLE=0
    PLN19      =FEAT/PLANE,CARTESIAN,OUTLINE
                THEO/<22.3099,11.7175,0.4912>,<0,1,0>
                ACTL/<22.3099,11.7175,0.4912>,<0,1,0>
                MEAS/PLANE,5
                  MOVE/CLEARPLANE
                  HIT/BASIC,NORMAL,<24.2036,11.7175,0.2197>,<0,1,0>,<24.2036,11.7175,0.2197>,USE THEO=YES
                  HIT/BASIC,NORMAL,<24.1536,11.7175,0.9028>,<0,1,0>,<24.1536,11.7175,0.9028>,USE THEO=YES
                  HIT/BASIC,NORMAL,<22.4493,11.7175,0.2166>,<0,1,0>,<22.4493,11.7175,0.2166>,USE THEO=YES
                  HIT/BASIC,NORMAL,<20.3616,11.7175,0.9033>,<0,1,0>,<20.3616,11.7175,0.9033>,USE THEO=YES
                  HIT/BASIC,NORMAL,<20.3813,11.7175,0.2133>,<0,1,0>,<20.3813,11.7175,0.2133>,USE THEO=YES
                ENDMEAS/
    PLN20      =FEAT/PLANE,CARTESIAN,OUTLINE
                THEO/<18.7712,11.8325,0.9896>,<0,1,0>
                ACTL/<18.7712,11.8325,0.9896>,<0,1,0>
                MEAS/PLANE,6
                  MOVE/CLEARPLANE
                  HIT/BASIC,NORMAL,<19.9622,11.8325,0.2073>,<0,1,0>,<19.9622,11.8325,0.2073>,USE THEO=YES
                  HIT/BASIC,NORMAL,<19.938,11.8325,0.9185>,<0,1,0>,<19.938,11.8325,0.9185>,USE THEO=YES
                  HIT/BASIC,NORMAL,<19.4205,11.8325,1.8495>,<0,1,0>,<19.4205,11.8325,1.8495>,USE THEO=YES
                  HIT/BASIC,NORMAL,<18.0901,11.8325,1.8453>,<0,1,0>,<18.0901,11.8325,1.8453>,USE THEO=YES
                  HIT/BASIC,NORMAL,<17.6086,11.8325,0.889>,<0,1,0>,<17.6086,11.8325,0.889>,USE THEO=YES
                  HIT/BASIC,NORMAL,<17.6078,11.8325,0.2281>,<0,1,0>,<17.6078,11.8325,0.2281>,USE THEO=YES
                ENDMEAS/
    PLN21      =FEAT/PLANE,CARTESIAN,OUTLINE
                THEO/<9.7648,11.7175,0.5554>,<0,1,0>
                ACTL/<9.7648,11.7175,0.5554>,<0,1,0>
                MEAS/PLANE,10
                  MOVE/CLEARPLANE
                  HIT/BASIC,NORMAL,<17.1484,11.7175,0.2286>,<0,1,0>,<17.1484,11.7175,0.2286>,USE THEO=YES
                  HIT/BASIC,NORMAL,<17.1564,11.7175,0.887>,<0,1,0>,<17.1564,11.7175,0.887>,USE THEO=YES
                  HIT/BASIC,NORMAL,<15.5107,11.7175,0.8727>,<0,1,0>,<15.5107,11.7175,0.8727>,USE THEO=YES
                  HIT/BASIC,NORMAL,<13.259,11.7175,0.2243>,<0,1,0>,<13.259,11.7175,0.2243>,USE THEO=YES
                  HIT/BASIC,NORMAL,<11.2633,11.7175,0.9053>,<0,1,0>,<11.2633,11.7175,0.9053>,USE THEO=YES
                  HIT/BASIC,NORMAL,<9.1774,11.7175,0.2198>,<0,1,0>,<9.1774,11.7175,0.2198>,USE THEO=YES
                  HIT/BASIC,NORMAL,<6.8362,11.7175,0.877>,<0,1,0>,<6.8362,11.7175,0.877>,USE THEO=YES
                  HIT/BASIC,NORMAL,<4.1262,11.7175,0.2283>,<0,1,0>,<4.1262,11.7175,0.2283>,USE THEO=YES
                  HIT/BASIC,NORMAL,<1.5944,11.7175,0.8751>,<0,1,0>,<1.5944,11.7175,0.8751>,USE THEO=YES
                  HIT/BASIC,NORMAL,<1.5759,11.7175,0.2362>,<0,1,0>,<1.5759,11.7175,0.2362>,USE THEO=YES
                ENDMEAS/
    PLN22      =FEAT/PLANE,CARTESIAN,OUTLINE
                THEO/<13.3175,14.4839,0.4631>,<0,1,0>
                ACTL/<13.3175,14.4839,0.4631>,<0,1,0>
                MEAS/PLANE,12
                  MOVE/CLEARPLANE
                  HIT/BASIC,NORMAL,<1.6111,14.4839,0.8846>,<0,1,0>,<1.6111,14.4839,0.8846>,USE THEO=YES
                  HIT/BASIC,NORMAL,<1.5924,14.4839,0.0523>,<0,1,0>,<1.5924,14.4839,0.0523>,USE THEO=YES
                  HIT/BASIC,NORMAL,<4.5721,14.4839,0.8675>,<0,1,0>,<4.5721,14.4839,0.8675>,USE THEO=YES
                  HIT/BASIC,NORMAL,<6.7748,14.4839,0.0463>,<0,1,0>,<6.7748,14.4839,0.0463>,USE THEO=YES
                  HIT/BASIC,NORMAL,<9.8363,14.4839,0.8698>,<0,1,0>,<9.8363,14.4839,0.8698>,USE THEO=YES
                  HIT/BASIC,NORMAL,<12.3835,14.4839,0.0373>,<0,1,0>,<12.3835,14.4839,0.0373>,USE THEO=YES
                  HIT/BASIC,NORMAL,<15.2595,14.4839,0.8848>,<0,1,0>,<15.2595,14.4839,0.8848>,USE THEO=YES
                  HIT/BASIC,NORMAL,<17.6652,14.4839,0.0314>,<0,1,0>,<17.6652,14.4839,0.0314>,USE THEO=YES
                  HIT/BASIC,NORMAL,<20.0667,14.4839,0.8888>,<0,1,0>,<20.0667,14.4839,0.8888>,USE THEO=YES
                  HIT/BASIC,NORMAL,<21.6425,14.4839,0.0496>,<0,1,0>,<21.6425,14.4839,0.0496>,USE THEO=YES
                  HIT/BASIC,NORMAL,<24.1926,14.4839,0.054>,<0,1,0>,<24.1926,14.4839,0.054>,USE THEO=YES
                  HIT/BASIC,NORMAL,<24.2136,14.4839,0.8907>,<0,1,0>,<24.2136,14.4839,0.8907>,USE THEO=YES
                ENDMEAS/
    PLN23      =FEAT/PLANE,CARTESIAN,OUTLINE
                THEO/<18.7563,14.4339,1.4644>,<0,1,0>
                ACTL/<18.7563,14.4339,1.4644>,<0,1,0>
                MEAS/PLANE,4
                  MOVE/CLEARPLANE
                  HIT/BASIC,NORMAL,<19.7382,14.4339,1.097>,<0,1,0>,<19.7382,14.4339,1.097>,USE THEO=YES
                  HIT/BASIC,NORMAL,<19.4025,14.4339,1.8267>,<0,1,0>,<19.4025,14.4339,1.8267>,USE THEO=YES
                  HIT/BASIC,NORMAL,<18.0996,14.4339,1.8249>,<0,1,0>,<18.0996,14.4339,1.8249>,USE THEO=YES
                  HIT/BASIC,NORMAL,<17.7848,14.4339,1.109>,<0,1,0>,<17.7848,14.4339,1.109>,USE THEO=YES
                ENDMEAS/
                MOVE/CLEARPLANE
                TIP/T1A10B90, SHANKIJK=-0.1736, 0, 0.9848, ANGLE=90
    PLN24      =FEAT/PLANE,CARTESIAN,OUTLINE
                THEO/<24.5011,17.2905,0.6887>,<-1,0,0>
                ACTL/<24.5011,17.2905,0.6887>,<-1,0,0>
                MEAS/PLANE,5
                  MOVE/CLEARPLANE
                  HIT/BASIC,NORMAL,<24.5011,14.8338,0.8836>,<-1,0,0>,<24.5011,14.8338,0.8836>,USE THEO=YES
                  HIT/BASIC,NORMAL,<24.5011,14.8195,0.052>,<-1,0,0>,<24.5011,14.8195,0.052>,USE THEO=YES
                  HIT/BASIC,NORMAL,<24.5011,17.2279,0.8765>,<-1,0,0>,<24.5011,17.2279,0.8765>,USE THEO=YES
                  HIT/BASIC,NORMAL,<24.5011,19.7713,0.0665>,<-1,0,0>,<24.5011,19.7713,0.0665>,USE THEO=YES
                  HIT/BASIC,NORMAL,<24.5011,19.7999,1.565>,<-1,0,0>,<24.5011,19.7999,1.565>,USE THEO=YES
                ENDMEAS/
                MOVE/CLEARPLANE
                TIP/T1A-10B0, SHANKIJK=0, -0.1736, 0.9848, ANGLE=0
    PLN25      =FEAT/PLANE,CARTESIAN,OUTLINE
                THEO/<13.0024,20.1069,0.902>,<0,-1,0>
                ACTL/<13.0024,20.1069,0.902>,<0,-1,0>
                MEAS/PLANE,9
                  MOVE/CLEARPLANE
                  HIT/BASIC,NORMAL,<24.1067,20.1069,1.5505>,<0,-1,0>,<24.1067,20.1069,1.5505>,USE THEO=YES
                  HIT/BASIC,NORMAL,<24.1654,20.1069,0.0774>,<0,-1,0>,<24.1654,20.1069,0.0774>,USE THEO=YES
                  HIT/BASIC,NORMAL,<20.2011,20.1069,1.5766>,<0,-1,0>,<20.2011,20.1069,1.5766>,USE THEO=YES
                  HIT/BASIC,NORMAL,<16.415,20.1069,0.0716>,<0,-1,0>,<16.415,20.1069,0.0716>,USE THEO=YES
                  HIT/BASIC,NORMAL,<13.3092,20.1069,1.572>,<0,-1,0>,<13.3092,20.1069,1.572>,USE THEO=YES
                  HIT/BASIC,NORMAL,<9.4845,20.1069,0.0504>,<0,-1,0>,<9.4845,20.1069,0.0504>,USE THEO=YES
                  HIT/BASIC,NORMAL,<6.1186,20.1069,1.5859>,<0,-1,0>,<6.1186,20.1069,1.5859>,USE THEO=YES
                  HIT/BASIC,NORMAL,<1.621,20.1069,0.0692>,<0,-1,0>,<1.621,20.1069,0.0692>,USE THEO=YES
                  HIT/BASIC,NORMAL,<1.6001,20.1069,1.564>,<0,-1,0>,<1.6001,20.1069,1.564>,USE THEO=YES
                ENDMEAS/
    PLN26      =FEAT/PLANE,CARTESIAN,OUTLINE
                THEO/<14.3019,26.0815,0.8364>,<0,-1,0>
                ACTL/<14.3019,26.0815,0.8364>,<0,-1,0>
                MEAS/PLANE,12
                  MOVE/CLEARPLANE
                  HIT/BASIC,NORMAL,<1.5649,26.0815,1.056>,<0,-1,0>,<1.5649,26.0815,1.056>,USE THEO=YES
                  HIT/BASIC,NORMAL,<3.6671,26.0815,0.775>,<0,-1,0>,<3.6671,26.0815,0.775>,USE THEO=YES
                  HIT/BASIC,NORMAL,<5.9059,26.0815,1.0726>,<0,-1,0>,<5.9059,26.0815,1.0726>,USE THEO=YES
                  HIT/BASIC,NORMAL,<7.8737,26.0815,0.7795>,<0,-1,0>,<7.8737,26.0815,0.7795>,USE THEO=YES
                  HIT/BASIC,NORMAL,<10.31,26.0815,1.0788>,<0,-1,0>,<10.31,26.0815,1.0788>,USE THEO=YES
                  HIT/BASIC,NORMAL,<12.7747,26.0815,0.7794>,<0,-1,0>,<12.7747,26.0815,0.7794>,USE THEO=YES
                  MOVE/POINT,NORMAL,<13.3913,25.8507,1.6>
                  MOVE/POINT,NORMAL,<14.2793,25.8509,1.6>
                  HIT/BASIC,NORMAL,<15.3337,26.0815,1.0588>,<0,-1,0>,<15.3337,26.0815,1.0588>,USE THEO=YES
                  HIT/BASIC,NORMAL,<18.1355,26.0815,0.7774>,<0,-1,0>,<18.1355,26.0815,0.7774>,USE THEO=YES
                  HIT/BASIC,NORMAL,<21.4959,26.0815,1.0497>,<0,-1,0>,<21.4959,26.0815,1.0497>,USE THEO=YES
                  HIT/BASIC,NORMAL,<24.2707,26.0815,0.8126>,<0,-1,0>,<24.2707,26.0815,0.8126>,USE THEO=YES
                  HIT/BASIC,NORMAL,<25.8482,26.0815,0.5006>,<0,-1,0>,<25.8482,26.0815,0.5006>,USE THEO=YES
                  HIT/BASIC,NORMAL,<24.4424,26.0815,0.2957>,<0,-1,0>,<24.4424,26.0815,0.2957>,USE THEO=YES
                ENDMEAS/
    PLN27      =FEAT/PLANE,CARTESIAN,OUTLINE
                THEO/<13.2075,25.8859,0.3087>,<0,-1,0>
                ACTL/<13.2075,25.8859,0.3087>,<0,-1,0>
                MEAS/PLANE,12
                  MOVE/CLEARPLANE
                  HIT/BASIC,NORMAL,<24.159,25.8859,0.0703>,<0,-1,0>,<24.159,25.8859,0.0703>,USE THEO=YES
                  HIT/BASIC,NORMAL,<24.0375,25.8859,0.5869>,<0,-1,0>,<24.0375,25.8859,0.5869>,USE THEO=YES
                  HIT/BASIC,NORMAL,<21.9945,25.8859,0.0213>,<0,-1,0>,<21.9945,25.8859,0.0213>,USE THEO=YES
                  HIT/BASIC,NORMAL,<19.5602,25.8859,0.5783>,<0,-1,0>,<19.5602,25.8859,0.5783>,USE THEO=YES
                  HIT/BASIC,NORMAL,<17.0761,25.8859,0.0451>,<0,-1,0>,<17.0761,25.8859,0.0451>,USE THEO=YES
                  HIT/BASIC,NORMAL,<14.7986,25.8859,0.5488>,<0,-1,0>,<14.7986,25.8859,0.5488>,USE THEO=YES
                  MOVE/POINT,NORMAL,<14.2238,23.7,0.4193>
                  MOVE/POINT,NORMAL,<13.0408,23.7,0.1518>
                  HIT/BASIC,NORMAL,<12.4295,25.8859,0.0242>,<0,-1,0>,<12.4295,25.8859,0.0242>,USE THEO=YES
                  HIT/BASIC,NORMAL,<9.9234,25.8859,0.5729>,<0,-1,0>,<9.9234,25.8859,0.5729>,USE THEO=YES
                  HIT/BASIC,NORMAL,<7.1634,25.8859,0.0402>,<0,-1,0>,<7.1634,25.8859,0.0402>,USE THEO=YES
                  HIT/BASIC,NORMAL,<4.2338,25.8859,0.584>,<0,-1,0>,<4.2338,25.8859,0.584>,USE THEO=YES
                  HIT/BASIC,NORMAL,<1.556,25.8859,0.0448>,<0,-1,0>,<1.556,25.8859,0.0448>,USE THEO=YES
                  HIT/BASIC,NORMAL,<1.558,25.8859,0.587>,<0,-1,0>,<1.558,25.8859,0.587>,USE THEO=YES
                ENDMEAS/
                MOVE/CLEARPLANE
                TIP/T1A-10B90, SHANKIJK=0.1736, 0, 0.9848, ANGLE=90
    PLN28      =FEAT/PLANE,CARTESIAN,OUTLINE
                THEO/<1.2587,23.118,0.7434>,<1,0,0>
                ACTL/<1.2587,23.118,0.7434>,<1,0,0>
                MEAS/PLANE,5
                  MOVE/CLEARPLANE
                  HIT/BASIC,NORMAL,<1.2587,25.728,1.0165>,<1,0,0>,<1.2587,25.728,1.0165>,USE THEO=YES
                  HIT/BASIC,NORMAL,<1.2587,25.549,0.0443>,<1,0,0>,<1.2587,25.549,0.0443>,USE THEO=YES
                  HIT/BASIC,NORMAL,<1.2587,23.1596,1.0453>,<1,0,0>,<1.2587,23.1596,1.0453>,USE THEO=YES
                  HIT/BASIC,NORMAL,<1.2587,20.5887,0.0584>,<1,0,0>,<1.2587,20.5887,0.0584>,USE THEO=YES
                  HIT/BASIC,NORMAL,<1.2587,20.5648,1.5525>,<1,0,0>,<1.2587,20.5648,1.5525>,USE THEO=YES
                ENDMEAS/
                MOVE/CLEARPLANE
                TIP/T1A10B0, SHANKIJK=0, 0.1736, 0.9848, ANGLE=0
    PLN29      =FEAT/PLANE,CARTESIAN,OUTLINE
                THEO/<13.235,20.2629,0.9115>,<0,1,0>
                ACTL/<13.235,20.2629,0.9115>,<0,1,0>
                MEAS/PLANE,9
                  MOVE/CLEARPLANE
                  HIT/BASIC,NORMAL,<1.6419,20.2629,1.5687>,<0,1,0>,<1.6419,20.2629,1.5687>,USE THEO=YES
                  HIT/BASIC,NORMAL,<1.6259,20.2629,0.0665>,<0,1,0>,<1.6259,20.2629,0.0665>,USE THEO=YES
                  HIT/BASIC,NORMAL,<5.9074,20.2629,1.5814>,<0,1,0>,<5.9074,20.2629,1.5814>,USE THEO=YES
                  HIT/BASIC,NORMAL,<9.3346,20.2629,0.0613>,<0,1,0>,<9.3346,20.2629,0.0613>,USE THEO=YES
                  HIT/BASIC,NORMAL,<13.6107,20.2629,1.58>,<0,1,0>,<13.6107,20.2629,1.58>,USE THEO=YES
                  HIT/BASIC,NORMAL,<17.5125,20.2629,0.0794>,<0,1,0>,<17.5125,20.2629,0.0794>,USE THEO=YES
                  HIT/BASIC,NORMAL,<21.1635,20.2629,1.5941>,<0,1,0>,<21.1635,20.2629,1.5941>,USE THEO=YES
                  HIT/BASIC,NORMAL,<24.1499,20.2629,0.0903>,<0,1,0>,<24.1499,20.2629,0.0903>,USE THEO=YES
                  HIT/BASIC,NORMAL,<24.1688,20.2629,1.5814>,<0,1,0>,<24.1688,20.2629,1.5814>,USE THEO=YES
                ENDMEAS/
                MOVE/CLEARPLANE
    PLN41      =FEAT/PLANE,CARTESIAN,OUTLINE
                THEO/<13.4029,26.1415,0.5646>,<0,1,0>
                ACTL/<13.4029,26.1415,0.5646>,<0,1,0>
                MEAS/PLANE,11
                  MOVE/CLEARPLANE
                  HIT/BASIC,NORMAL,<0.1493,26.1415,0.9888>,<0,1,0>,<0.1493,26.1415,0.9888>,USE THEO=YES
                  HIT/BASIC,NORMAL,<0.1354,26.1415,0.2156>,<0,1,0>,<0.1354,26.1415,0.2156>,USE THEO=YES
                  HIT/BASIC,NORMAL,<3.4971,26.1415,1.0241>,<0,1,0>,<3.4971,26.1415,1.0241>,USE THEO=YES
                  HIT/BASIC,NORMAL,<6.7462,26.1415,0.2045>,<0,1,0>,<6.7462,26.1415,0.2045>,USE THEO=YES
                  HIT/BASIC,NORMAL,<10.296,26.1415,1.0029>,<0,1,0>,<10.296,26.1415,1.0029>,USE THEO=YES
                  HIT/BASIC,NORMAL,<12.9887,26.1415,0.228>,<0,1,0>,<12.9887,26.1415,0.228>,USE THEO=YES
                  MOVE/POINT,NORMAL,<13.5999,26.75,0.3433>
                  MOVE/POINT,NORMAL,<14.0686,26.5,0.4296>
                  HIT/BASIC,NORMAL,<17.0783,26.1415,0.9994>,<0,1,0>,<17.0783,26.1415,0.9994>,USE THEO=YES
                  HIT/BASIC,NORMAL,<20.5307,26.1415,0.1886>,<0,1,0>,<20.5307,26.1415,0.1886>,USE THEO=YES
                  MOVE/POINT,NORMAL,<22.3539,26.368,1.6>
                  MOVE/POINT,NORMAL,<23.5696,26.3745,1.6>
                  HIT/BASIC,NORMAL,<24.2751,26.1415,0.6914>,<0,1,0>,<24.2751,26.1415,0.6914>,USE THEO=YES
                  HIT/BASIC,NORMAL,<25.8508,26.1415,0.5208>,<0,1,0>,<25.8508,26.1415,0.5208>,USE THEO=YES
                  HIT/BASIC,NORMAL,<25.8843,26.1415,0.146>,<0,1,0>,<25.8843,26.1415,0.146>,USE THEO=YES
                ENDMEAS/
                MOVE/CLEARPLANE
                TIP/T1A10B90, SHANKIJK=-0.1736, 0, 0.9848, ANGLE=90
                CLEARP/ZPLUS,2,ZPLUS,2,OFF
    PLN30      =FEAT/PLANE,CARTESIAN,OUTLINE
                THEO/<24.5011,23.0096,0.5482>,<-1,0,0>
                ACTL/<24.5011,23.0096,0.5482>,<-1,0,0>
                MEAS/PLANE,6
                  HIT/BASIC,NORMAL,<24.5011,20.5982,1.4588>,<-1,0,0>,<24.5011,20.5982,1.4588>,USE THEO=YES
                  HIT/BASIC,NORMAL,<24.5011,20.6013,0.0742>,<-1,0,0>,<24.5011,20.6013,0.0742>,USE THEO=YES
                  HIT/BASIC,NORMAL,<24.5011,22.5138,0.8487>,<-1,0,0>,<24.5011,22.5138,0.8487>,USE THEO=YES
                  HIT/BASIC,NORMAL,<24.5011,24.34,0.7756>,<-1,0,0>,<24.5011,24.34,0.7756>,USE THEO=YES
                  HIT/BASIC,NORMAL,<24.5011,24.47,0.0773>,<-1,0,0>,<24.5011,24.47,0.0773>,USE THEO=YES
                  HIT/BASIC,NORMAL,<24.5011,25.534,0.0548>,<-1,0,0>,<24.5011,25.534,0.0548>,USE THEO=YES
                ENDMEAS/
                MOVE/POINT,NORMAL,<24.25,25.5,0.25>
    PLN31      =FEAT/PLANE,CARTESIAN,OUTLINE
                THEO/<27.1489,22.0985,0.4615>,<-0.9542031,-0.2991597,0>
                ACTL/<27.1489,22.0985,0.4615>,<-0.9542031,-0.2991597,0>
                MEAS/PLANE,5
                  HIT/BASIC,NORMAL,<26.4575,24.3036,0.2821>,<-0.9542031,-0.2991597,0>,<26.4575,24.3036,0.2821>,USE THEO=YES
                  HIT/BASIC,NORMAL,<26.4798,24.2325,0.7148>,<-0.9542031,-0.2991597,0>,<26.4798,24.2325,0.7148>,USE THEO=YES
                  HIT/BASIC,NORMAL,<27.1251,22.1744,0.278>,<-0.9542031,-0.2991597,0>,<27.1251,22.1744,0.278>,USE THEO=YES
                  HIT/BASIC,NORMAL,<27.8377,19.9014,0.7374>,<-0.9542031,-0.2991597,0>,<27.8377,19.9014,0.7374>,USE THEO=YES
                  HIT/BASIC,NORMAL,<27.8443,19.8804,0.2954>,<-0.9542031,-0.2991597,0>,<27.8443,19.8804,0.2954>,USE THEO=YES
                ENDMEAS/
    PLN32      =FEAT/PLANE,CARTESIAN,OUTLINE
                THEO/<27.91,13.2435,0.5103>,<-1,0,0>
                ACTL/<27.91,13.2435,0.5103>,<-1,0,0>
                MEAS/PLANE,8
                  HIT/BASIC,NORMAL,<27.91,19.4431,0.7447>,<-1,0,0>,<27.91,19.4431,0.7447>,USE THEO=YES
                  HIT/BASIC,NORMAL,<27.91,19.4657,0.2943>,<-1,0,0>,<27.91,19.4657,0.2943>,USE THEO=YES
                  HIT/BASIC,NORMAL,<27.91,16.8924,0.7253>,<-1,0,0>,<27.91,16.8924,0.7253>,USE THEO=YES
                  HIT/BASIC,NORMAL,<27.91,14.5679,0.26>,<-1,0,0>,<27.91,14.5679,0.26>,USE THEO=YES
                  HIT/BASIC,NORMAL,<27.91,12.2922,0.7521>,<-1,0,0>,<27.91,12.2922,0.7521>,USE THEO=YES
                  HIT/BASIC,NORMAL,<27.91,9.9166,0.2777>,<-1,0,0>,<27.91,9.9166,0.2777>,USE THEO=YES
                  HIT/BASIC,NORMAL,<27.91,6.6886,0.7425>,<-1,0,0>,<27.91,6.6886,0.7425>,USE THEO=YES
                  HIT/BASIC,NORMAL,<27.91,6.6816,0.2856>,<-1,0,0>,<27.91,6.6816,0.2856>,USE THEO=YES
                ENDMEAS/
    PLN33      =FEAT/PLANE,CARTESIAN,OUTLINE
                THEO/<27.1764,4.1308,0.5481>,<-0.9542031,0.2991597,0>
                ACTL/<27.1764,4.1308,0.5481>,<-0.9542031,0.2991597,0>
                MEAS/PLANE,5
                  HIT/BASIC,NORMAL,<27.8547,6.2944,0.7646>,<-0.9542031,0.2991597,0>,<27.8547,6.2944,0.7646>,USE THEO=YES
                  HIT/BASIC,NORMAL,<27.855,6.2952,0.2692>,<-0.9542031,0.2991597,0>,<27.855,6.2952,0.2692>,USE THEO=YES
                  HIT/BASIC,NORMAL,<27.2207,4.2722,0.7319>,<-0.9542031,0.2991597,0>,<27.2207,4.2722,0.7319>,USE THEO=YES
                  HIT/BASIC,NORMAL,<26.4668,1.8675,0.273>,<-0.9542031,0.2991597,0>,<26.4668,1.8675,0.273>,USE THEO=YES
                  HIT/BASIC,NORMAL,<26.4848,1.9248,0.7018>,<-0.9542031,0.2991597,0>,<26.4848,1.9248,0.7018>,USE THEO=YES
                ENDMEAS/
                MOVE/POINT,NORMAL,<26,1.9,2>
                TIP/T1A-10B90, SHANKIJK=0.1736, 0, 0.9848, ANGLE=90
                MOVE/POINT,NORMAL,<25,1.9,2>
    PLN34      =FEAT/PLANE,CARTESIAN,OUTLINE
                THEO/<24.5711,13.163,0.715>,<1,0,0>
                ACTL/<24.5711,13.163,0.715>,<1,0,0>
                MEAS/PLANE,15
                  HIT/BASIC,NORMAL,<24.5711,1.7934,0.3202>,<1,0,0>,<24.5711,1.7934,0.3202>,USE THEO=YES
                  HIT/BASIC,NORMAL,<24.5711,1.8231,0.8576>,<1,0,0>,<24.5711,1.8231,0.8576>,USE THEO=YES
                  HIT/BASIC,NORMAL,<24.5711,3.6298,0.2818>,<1,0,0>,<24.5711,3.6298,0.2818>,USE THEO=YES
                  HIT/BASIC,NORMAL,<24.5711,4.9705,0.8793>,<1,0,0>,<24.5711,4.9705,0.8793>,USE THEO=YES
                  HIT/BASIC,NORMAL,<24.5711,5.9263,1.5601>,<1,0,0>,<24.5711,5.9263,1.5601>,USE THEO=YES
                  HIT/BASIC,NORMAL,<24.5711,7.111,0.8179>,<1,0,0>,<24.5711,7.111,0.8179>,USE THEO=YES
                  HIT/BASIC,NORMAL,<24.5711,9.92,0.2754>,<1,0,0>,<24.5711,9.92,0.2754>,USE THEO=YES
                  HIT/BASIC,NORMAL,<24.5711,13.157,0.857>,<1,0,0>,<24.5711,13.157,0.857>,USE THEO=YES
                  HIT/BASIC,NORMAL,<24.5711,16.814,0.2815>,<1,0,0>,<24.5711,16.814,0.2815>,USE THEO=YES
                  HIT/BASIC,NORMAL,<24.5711,19.0761,0.8416>,<1,0,0>,<24.5711,19.0761,0.8416>,USE THEO=YES
                  HIT/BASIC,NORMAL,<24.5711,20.1406,1.4992>,<1,0,0>,<24.5711,20.1406,1.4992>,USE THEO=YES
                  HIT/BASIC,NORMAL,<24.5711,21.3812,0.8059>,<1,0,0>,<24.5711,21.3812,0.8059>,USE THEO=YES
                  HIT/BASIC,NORMAL,<24.5711,23.0081,0.3137>,<1,0,0>,<24.5711,23.0081,0.3137>,USE THEO=YES
                  HIT/BASIC,NORMAL,<24.5711,24.3321,0.8282>,<1,0,0>,<24.5711,24.3321,0.8282>,USE THEO=YES
                  HIT/BASIC,NORMAL,<24.5711,24.3617,0.306>,<1,0,0>,<24.5711,24.3617,0.306>,USE THEO=YES
                ENDMEAS/
    PLN35      =FEAT/PLANE,CARTESIAN,OUTLINE
                THEO/<26.8526,23.2442,0.2882>,<0.9542031,0.2991597,0>
                ACTL/<26.8526,23.2442,0.2882>,<0.9542031,0.2991597,0>
                MEAS/PLANE,7
                  HIT/BASIC,NORMAL,<25.9577,26.0985,0.3701>,<0.9542031,0.2991597,0>,<25.9577,26.0985,0.3701>,USE THEO=YES
                  HIT/BASIC,NORMAL,<25.9645,26.0768,0.0706>,<0.9542031,0.2991597,0>,<25.9645,26.0768,0.0706>,USE THEO=YES
                  HIT/BASIC,NORMAL,<26.5242,24.2915,0.0422>,<0.9542031,0.2991597,0>,<26.5242,24.2915,0.0422>,USE THEO=YES
                  HIT/BASIC,NORMAL,<26.5473,24.2177,0.7118>,<0.9542031,0.2991597,0>,<26.5473,24.2177,0.7118>,USE THEO=YES
                  HIT/BASIC,NORMAL,<27.182,22.1934,0.0429>,<0.9542031,0.2991597,0>,<27.182,22.1934,0.0429>,USE THEO=YES
                  HIT/BASIC,NORMAL,<27.9004,19.9019,0.7258>,<0.9542031,0.2991597,0>,<27.9004,19.9019,0.7258>,USE THEO=YES
                  HIT/BASIC,NORMAL,<27.8918,19.9293,0.054>,<0.9542031,0.2991597,0>,<27.8918,19.9293,0.054>,USE THEO=YES
                ENDMEAS/
                MOVE/POINT,NORMAL,<28,20,2>
                TIP/T1A0B0, SHANKIJK=0, 0, 1, ANGLE=0
                MOVE/POINT,NORMAL,<28,20,2>
    PLN36      =FEAT/PLANE,CARTESIAN,OUTLINE
                THEO/<27.97,13.0738,0.429>,<1,0,0>
                ACTL/<27.97,13.0738,0.429>,<1,0,0>
                MEAS/PLANE,8
                  HIT/BASIC,NORMAL,<27.97,19.4405,0.1034>,<1,0,0>,<27.97,19.4405,0.1034>,USE THEO=YES
                  HIT/BASIC,NORMAL,<27.97,19.4421,0.7574>,<1,0,0>,<27.97,19.4421,0.7574>,USE THEO=YES
                  MOVE/POINT,NORMAL,<28.1928,19.128,1.6>
                  MOVE/POINT,NORMAL,<28.1984,17.7585,1.6>
                  HIT/BASIC,NORMAL,<27.97,16.754,0.7579>,<1,0,0>,<27.97,16.754,0.7579>,USE THEO=YES
                  HIT/BASIC,NORMAL,<27.97,14.434,0.1047>,<1,0,0>,<27.97,14.434,0.1047>,USE THEO=YES
                  HIT/BASIC,NORMAL,<27.97,11.7192,0.7215>,<1,0,0>,<27.97,11.7192,0.7215>,USE THEO=YES
                  HIT/BASIC,NORMAL,<27.97,9.3588,0.1241>,<1,0,0>,<27.97,9.3588,0.1241>,USE THEO=YES
                  MOVE/POINT,NORMAL,<28.1962,8.8931,1.6>
                  MOVE/POINT,NORMAL,<28.1948,7.3754,1.6>
                  HIT/BASIC,NORMAL,<27.97,6.7487,0.7329>,<1,0,0>,<27.97,6.7487,0.7329>,USE THEO=YES
                  HIT/BASIC,NORMAL,<27.97,6.6932,0.1298>,<1,0,0>,<27.97,6.6932,0.1298>,USE THEO=YES
                ENDMEAS/
    PLN37      =FEAT/PLANE,CARTESIAN,OUTLINE
                THEO/<26.8955,3.0343,0.3725>,<0.9542031,-0.2991597,0>
                ACTL/<26.8955,3.0343,0.3725>,<0.9542031,-0.2991597,0>
                MEAS/PLANE,8
                  HIT/BASIC,NORMAL,<27.9021,6.245,0.1409>,<0.9542031,-0.2991597,0>,<27.9021,6.245,0.1409>,USE THEO=YES
                  HIT/BASIC,NORMAL,<27.9015,6.2429,0.7645>,<0.9542031,-0.2991597,0>,<27.9015,6.2429,0.7645>,USE THEO=YES
                  HIT/BASIC,NORMAL,<27.4128,4.6844,0.1397>,<0.9542031,-0.2991597,0>,<27.4128,4.6844,0.1397>,USE THEO=YES
                  HIT/BASIC,NORMAL,<26.9941,3.3488,0.7325>,<0.9542031,-0.2991597,0>,<26.9941,3.3488,0.7325>,USE THEO=YES
                  HIT/BASIC,NORMAL,<26.5404,1.9016,0.708>,<0.9542031,-0.2991597,0>,<26.5404,1.9016,0.708>,USE THEO=YES
                  HIT/BASIC,NORMAL,<26.4912,1.7448,0.0621>,<0.9542031,-0.2991597,0>,<26.4912,1.7448,0.0621>,USE THEO=YES
                  HIT/BASIC,NORMAL,<25.9699,0.0819,0.0697>,<0.9542031,-0.2991597,0>,<25.9699,0.0819,0.0697>,USE THEO=YES
                  HIT/BASIC,NORMAL,<25.952,0.0248,0.3628>,<0.9542031,-0.2991597,0>,<25.952,0.0248,0.3628>,USE THEO=YES
                ENDMEAS/
    PLN38      =FEAT/PLANE,CARTESIAN,OUTLINE
                THEO/<25.6877,13.0528,0.113>,<0,0,1>
                ACTL/<25.6877,13.0528,0.113>,<0,0,1>
                MEAS/PLANE,19
                  HIT/BASIC,NORMAL,<25.9701,0.2536,0.113>,<0,0,1>,<25.9701,0.2536,0.113>,USE THEO=YES
                  HIT/BASIC,NORMAL,<24.349,0.233,0.113>,<0,0,1>,<24.349,0.233,0.113>,USE THEO=YES
                  HIT/BASIC,NORMAL,<24.5477,1.4865,0.113>,<0,0,1>,<24.5477,1.4865,0.113>,USE THEO=YES
                  HIT/BASIC,NORMAL,<26.4013,1.6158,0.113>,<0,0,1>,<26.4013,1.6158,0.113>,USE THEO=YES
                  HIT/BASIC,NORMAL,<24.7394,2.0515,0.113>,<0,0,1>,<24.7394,2.0515,0.113>,USE THEO=YES
                  HIT/BASIC,NORMAL,<26.8816,3.7862,0.113>,<0,0,1>,<26.8816,3.7862,0.113>,USE THEO=YES
                  HIT/BASIC,NORMAL,<24.7949,5.7786,0.113>,<0,0,1>,<24.7949,5.7786,0.113>,USE THEO=YES
                  HIT/BASIC,NORMAL,<27.6897,6.5469,0.113>,<0,0,1>,<27.6897,6.5469,0.113>,USE THEO=YES
                  HIT/BASIC,NORMAL,<24.7995,9.6428,0.113>,<0,0,1>,<24.7995,9.6428,0.113>,USE THEO=YES
                  HIT/BASIC,NORMAL,<27.5922,13.2591,0.113>,<0,0,1>,<27.5922,13.2591,0.113>,USE THEO=YES
                  HIT/BASIC,NORMAL,<24.823,17.164,0.113>,<0,0,1>,<24.823,17.164,0.113>,USE THEO=YES
                  HIT/BASIC,NORMAL,<27.6878,19.5878,0.113>,<0,0,1>,<27.6878,19.5878,0.113>,USE THEO=YES
                  HIT/BASIC,NORMAL,<24.7831,20.0299,0.113>,<0,0,1>,<24.7831,20.0299,0.113>,USE THEO=YES
                  HIT/BASIC,NORMAL,<26.9394,22.1081,0.113>,<0,0,1>,<26.9394,22.1081,0.113>,USE THEO=YES
                  HIT/BASIC,NORMAL,<24.7612,23.4671,0.113>,<0,0,1>,<24.7612,23.4671,0.113>,USE THEO=YES
                  HIT/BASIC,NORMAL,<26.3178,24.5768,0.113>,<0,0,1>,<26.3178,24.5768,0.113>,USE THEO=YES
                  HIT/BASIC,NORMAL,<24.6236,24.6762,0.113>,<0,0,1>,<24.6236,24.6762,0.113>,USE THEO=YES
                  HIT/BASIC,NORMAL,<24.4027,25.907,0.113>,<0,0,1>,<24.4027,25.907,0.113>,USE THEO=YES
                  HIT/BASIC,NORMAL,<25.9625,25.8324,0.113>,<0,0,1>,<25.9625,25.8324,0.113>,USE THEO=YES
                ENDMEAS/
                MOVE/POINT,NORMAL,<25.8,25.8,2>
                MOVE/POINT,NORMAL,<24.3,14,2>
    PLN39      =FEAT/PLANE,CARTESIAN,OUTLINE
                THEO/<14.524,13.0719,0.06>,<0,0,1>
                ACTL/<14.524,13.0719,0.06>,<0,0,1>
                MEAS/PLANE,13
                  HIT/BASIC,NORMAL,<24.2951,14.2132,0.06>,<0,0,1>,<24.2951,14.2132,0.06>,USE THEO=YES
                  HIT/BASIC,NORMAL,<24.2877,11.9359,0.06>,<0,0,1>,<24.2877,11.9359,0.06>,USE THEO=YES
                  HIT/BASIC,NORMAL,<20.354,11.9498,0.06>,<0,0,1>,<20.354,11.9498,0.06>,USE THEO=YES
                  HIT/BASIC,NORMAL,<20.3459,14.2036,0.06>,<0,0,1>,<20.3459,14.2036,0.06>,USE THEO=YES
                  HIT/BASIC,NORMAL,<17.2517,11.939,0.06>,<0,0,1>,<17.2517,11.939,0.06>,USE THEO=YES
                  HIT/BASIC,NORMAL,<17.2858,14.1553,0.06>,<0,0,1>,<17.2858,14.1553,0.06>,USE THEO=YES
                  HIT/BASIC,NORMAL,<14.6086,13.1337,0.06>,<0,0,1>,<14.6086,13.1337,0.06>,USE THEO=YES
                  HIT/BASIC,NORMAL,<12.2368,14.1393,0.06>,<0,0,1>,<12.2368,14.1393,0.06>,USE THEO=YES
                  HIT/BASIC,NORMAL,<12.1916,11.9791,0.06>,<0,0,1>,<12.1916,11.9791,0.06>,USE THEO=YES
                  HIT/BASIC,NORMAL,<8.4684,14.2318,0.06>,<0,0,1>,<8.4684,14.2318,0.06>,USE THEO=YES
                  HIT/BASIC,NORMAL,<8.4277,11.9176,0.06>,<0,0,1>,<8.4277,11.9176,0.06>,USE THEO=YES
                  HIT/BASIC,NORMAL,<4.5235,14.2208,0.06>,<0,0,1>,<4.5235,14.2208,0.06>,USE THEO=YES
                  HIT/BASIC,NORMAL,<4.5358,11.9154,0.06>,<0,0,1>,<4.5358,11.9154,0.06>,USE THEO=YES
                ENDMEAS/
    PLN40      =FEAT/PLANE,CARTESIAN,OUTLINE
                THEO/<1.6838,13.0558,0.06>,<0,0,1>
                ACTL/<1.6838,13.0558,0.06>,<0,0,1>
                MEAS/PLANE,4
                  HIT/BASIC,NORMAL,<1.8704,14.225,0.06>,<0,0,1>,<1.8704,14.225,0.06>,USE THEO=YES
                  HIT/BASIC,NORMAL,<1.4911,14.2074,0.06>,<0,0,1>,<1.4911,14.2074,0.06>,USE THEO=YES
                  HIT/BASIC,NORMAL,<1.4484,11.9085,0.06>,<0,0,1>,<1.4484,11.9085,0.06>,USE THEO=YES
                  HIT/BASIC,NORMAL,<1.9254,11.8822,0.06>,<0,0,1>,<1.9254,11.8822,0.06>,USE THEO=YES
                ENDMEAS/
    N2         =FEAT/CONTACT/CIRCLE/DEFAULT,CARTESIAN,IN,LEAST_SQR
                THEO/<1.6091,13.0713,0.06>,<0,0,1>,0.2235
                ACTL/<1.6091,13.0713,0.06>,<0,0,1>,0.2235
                TARG/<1.6091,13.0713,0.06>,<0,0,1>
                START ANG=0,END ANG=360
                ANGLE VEC=<1,0,0>
                DIRECTION=CCW
                SHOW FEATURE PARAMETERS=NO
                SHOW CONTACT PARAMETERS=YES
                  NUMHITS=5,DEPTH=0.03,PITCH=0
                  SAMPLE METHOD=SAMPLE_FEATURE
                  SAMPLE FEATURE=NONE
                  AVOIDANCE MOVE=BOTH,DISTANCE BEFORE=0.1,DISTANCE AFTER=0.1
                  FIND HOLE=DISABLED,ONERROR=NO,READ POS=NO
                SHOW HITS=NO
    P1         =FEAT/CONTACT/SQUARE SLOT/DEFAULT,CARTESIAN,IN
                THEO/<3.1787,13.0707,0.06>,<0,0,1>,<1,0,0>,2.5,2.6
                ACTL/<3.1787,13.0707,0.06>,<0,0,1>,<1,0,0>,2.5,2.6
                TARG/<3.1787,13.0707,0.06>,<0,0,1>,<1,0,0>
                MEAS WIDTH=NO,RADIUS=0.25
                SHOW FEATURE PARAMETERS=NO
                SHOW CONTACT PARAMETERS=YES
                  DEPTH=0.025
                  SAMPLE METHOD=SAMPLE_FEATURE
                  SAMPLE FEATURE=NONE
                  AVOIDANCE MOVE=BOTH,DISTANCE BEFORE=0.1,DISTANCE AFTER=0.1
                  FIND HOLE=DISABLED,ONERROR=NO,READ POS=NO
                SHOW HITS=NO
    N1         =FEAT/CONTACT/CIRCLE/DEFAULT,CARTESIAN,IN,LEAST_SQR
                THEO/<4.8291,13.0713,0.06>,<0,0,1>,0.2235
                ACTL/<4.8291,13.0713,0.06>,<0,0,1>,0.2235
                TARG/<4.8291,13.0713,0.06>,<0,0,1>
                START ANG=0,END ANG=360
                ANGLE VEC=<1,0,0>
                DIRECTION=CCW
                SHOW FEATURE PARAMETERS=NO
                SHOW CONTACT PARAMETERS=YES
                  NUMHITS=5,DEPTH=0.03,PITCH=0
                  SAMPLE METHOD=SAMPLE_FEATURE
                  SAMPLE FEATURE=NONE
                  AVOIDANCE MOVE=BOTH,DISTANCE BEFORE=0.1,DISTANCE AFTER=0.1
                  FIND HOLE=DISABLED,ONERROR=NO,READ POS=NO
                SHOW HITS=NO
    B14        =FEAT/CONTACT/CIRCLE/DEFAULT,CARTESIAN,IN,LEAST_SQR
                THEO/<5.7001,14.0107,0.06>,<0,0,1>,0.2
                ACTL/<5.7001,14.0107,0.06>,<0,0,1>,0.2
                TARG/<5.7001,14.0107,0.06>,<0,0,1>
                START ANG=0,END ANG=360
                ANGLE VEC=<1,0,0>
                DIRECTION=CCW
                SHOW FEATURE PARAMETERS=NO
                SHOW CONTACT PARAMETERS=YES
                  NUMHITS=5,DEPTH=0.03,PITCH=0
                  SAMPLE METHOD=SAMPLE_FEATURE
                  SAMPLE FEATURE=NONE
                  AVOIDANCE MOVE=BOTH,DISTANCE BEFORE=0.1,DISTANCE AFTER=0.1
                  FIND HOLE=DISABLED,ONERROR=NO,READ POS=NO
                SHOW HITS=NO
    M2B        =FEAT/CONTACT/CIRCLE/DEFAULT,CARTESIAN,IN,LEAST_SQR
                THEO/<5.7701,13.0707,0.06>,<0,0,1>,0.199
                ACTL/<5.7701,13.0707,0.06>,<0,0,1>,0.199
                TARG/<5.7701,13.0707,0.06>,<0,0,1>
                START ANG=89.9998,END ANG=270
                ANGLE VEC=<1,0,0>
                DIRECTION=CCW
                SHOW FEATURE PARAMETERS=NO
                SHOW CONTACT PARAMETERS=YES
                  NUMHITS=5,DEPTH=0.03,PITCH=0
                  SAMPLE METHOD=SAMPLE_FEATURE
                  SAMPLE FEATURE=NONE
                  AVOIDANCE MOVE=BOTH,DISTANCE BEFORE=0.1,DISTANCE AFTER=0.1
                  FIND HOLE=DISABLED,ONERROR=NO,READ POS=NO
                SHOW HITS=NO
    M2         =FEAT/CONTACT/CIRCLE/DEFAULT,CARTESIAN,IN,LEAST_SQR
                THEO/<6.2701,13.0707,0.06>,<0,0,1>,0.39
                ACTL/<6.2701,13.0707,0.06>,<0,0,1>,0.39
                TARG/<6.2701,13.0707,0.06>,<0,0,1>
                START ANG=229.4,END ANG=491.3
                ANGLE VEC=<1,0,0>
                DIRECTION=CCW
                SHOW FEATURE PARAMETERS=NO
                SHOW CONTACT PARAMETERS=YES
                  NUMHITS=5,DEPTH=0.03,PITCH=0
                  SAMPLE METHOD=SAMPLE_FEATURE
                  SAMPLE FEATURE=NONE
                  AVOIDANCE MOVE=BOTH,DISTANCE BEFORE=0.1,DISTANCE AFTER=0.1
                  FIND HOLE=DISABLED,ONERROR=NO,READ POS=NO
                SHOW HITS=NO
    M1B        =FEAT/CONTACT/CIRCLE/DEFAULT,CARTESIAN,IN,LEAST_SQR
                THEO/<7.9501,13.0707,0.06>,<0,0,1>,0.199
                ACTL/<7.9501,13.0707,0.06>,<0,0,1>,0.199
                TARG/<7.9501,13.0707,0.06>,<0,0,1>
                START ANG=89.9998,END ANG=270
                ANGLE VEC=<1,0,0>
                DIRECTION=CCW
                SHOW FEATURE PARAMETERS=NO
                SHOW CONTACT PARAMETERS=YES
                  NUMHITS=5,DEPTH=0.03,PITCH=0
                  SAMPLE METHOD=SAMPLE_FEATURE
                  SAMPLE FEATURE=NONE
                  AVOIDANCE MOVE=BOTH,DISTANCE BEFORE=0.1,DISTANCE AFTER=0.1
                  FIND HOLE=DISABLED,ONERROR=NO,READ POS=NO
                SHOW HITS=NO
    M1         =FEAT/CONTACT/CIRCLE/DEFAULT,CARTESIAN,IN,LEAST_SQR
                THEO/<8.4501,13.0707,0.06>,<0,0,1>,0.39
                ACTL/<8.4501,13.0707,0.06>,<0,0,1>,0.39
                TARG/<8.4501,13.0707,0.06>,<0,0,1>
                START ANG=226.1,END ANG=497.3
                ANGLE VEC=<1,0,0>
                DIRECTION=CCW
                SHOW FEATURE PARAMETERS=NO
                SHOW CONTACT PARAMETERS=YES
                  NUMHITS=5,DEPTH=0.03,PITCH=0
                  SAMPLE METHOD=SAMPLE_FEATURE
                  SAMPLE FEATURE=NONE
                  AVOIDANCE MOVE=BOTH,DISTANCE BEFORE=0.1,DISTANCE AFTER=0.1
                  FIND HOLE=DISABLED,ONERROR=NO,READ POS=NO
                SHOW HITS=NO
    B13        =FEAT/CONTACT/CIRCLE/DEFAULT,CARTESIAN,IN,LEAST_SQR
                THEO/<8.9401,12.2807,0.06>,<0,0,1>,0.2
                ACTL/<8.9401,12.2807,0.06>,<0,0,1>,0.2
                TARG/<8.9401,12.2807,0.06>,<0,0,1>
                START ANG=0,END ANG=360
                ANGLE VEC=<1,0,0>
                DIRECTION=CCW
                SHOW FEATURE PARAMETERS=NO
                SHOW CONTACT PARAMETERS=YES
                  NUMHITS=5,DEPTH=0.03,PITCH=0
                  SAMPLE METHOD=SAMPLE_FEATURE
                  SAMPLE FEATURE=NONE
                  AVOIDANCE MOVE=BOTH,DISTANCE BEFORE=0.1,DISTANCE AFTER=0.1
                  FIND HOLE=DISABLED,ONERROR=NO,READ POS=NO
                SHOW HITS=NO
    G6         =FEAT/CONTACT/SQUARE SLOT/DEFAULT,CARTESIAN,IN
                THEO/<9.9887,13.0707,0.06>,<0,0,1>,<0,-1,0>,0.44,0.94
                ACTL/<9.9887,13.0707,0.06>,<0,0,1>,<0,-1,0>,0.44,0.94
                TARG/<9.9887,13.0707,0.06>,<0,0,1>,<0,-1,0>
                MEAS WIDTH=NO,RADIUS=0.0689
                SHOW FEATURE PARAMETERS=NO
                SHOW CONTACT PARAMETERS=YES
                  DEPTH=0.02
                  SAMPLE METHOD=SAMPLE_FEATURE
                  SAMPLE FEATURE=NONE
                  AVOIDANCE MOVE=BOTH,DISTANCE BEFORE=0.1,DISTANCE AFTER=0.1
                  FIND HOLE=DISABLED,ONERROR=NO,READ POS=NO
                SHOW HITS=NO
    L1         =FEAT/CONTACT/SQUARE SLOT/DEFAULT,CARTESIAN,IN
                THEO/<11.18,13.0707,0.06>,<0,0,1>,<0,1,0>,0.44,1.94
                ACTL/<11.18,13.0707,0.06>,<0,0,1>,<0,1,0>,0.44,1.94
                TARG/<11.18,13.0707,0.06>,<0,0,1>,<0,1,0>
                MEAS WIDTH=NO,RADIUS=0.207
                SHOW FEATURE PARAMETERS=NO
                SHOW CONTACT PARAMETERS=YES
                  DEPTH=0.02
                  SAMPLE METHOD=SAMPLE_FEATURE
                  SAMPLE FEATURE=NONE
                  AVOIDANCE MOVE=BOTH,DISTANCE BEFORE=0.1,DISTANCE AFTER=0.1
                  FIND HOLE=DISABLED,ONERROR=NO,READ POS=NO
                SHOW HITS=NO
    A18        =FEAT/CONTACT/CIRCLE/DEFAULT,CARTESIAN,IN,LEAST_SQR
                THEO/<12.2,13.7207,0.06>,<0,0,1>,0.48
                ACTL/<12.2,13.7207,0.06>,<0,0,1>,0.48
                TARG/<12.2,13.7207,0.06>,<0,0,1>
                START ANG=29.5,END ANG=331.7
                ANGLE VEC=<1,0,0>
                DIRECTION=CCW
                SHOW FEATURE PARAMETERS=NO
                SHOW CONTACT PARAMETERS=YES
                  NUMHITS=5,DEPTH=0.03,PITCH=0
                  SAMPLE METHOD=SAMPLE_FEATURE
                  SAMPLE FEATURE=NONE
                  AVOIDANCE MOVE=BOTH,DISTANCE BEFORE=0.1,DISTANCE AFTER=0.1
                  FIND HOLE=DISABLED,ONERROR=NO,READ POS=NO
                SHOW HITS=NO
    A18B       =FEAT/CONTACT/CIRCLE/DEFAULT,CARTESIAN,IN,LEAST_SQR
                THEO/<12.57,13.7207,0.06>,<0,0,1>,0.128
                ACTL/<12.57,13.7207,0.06>,<0,0,1>,0.128
                TARG/<12.57,13.7207,0.06>,<0,0,1>
                START ANG=270,END ANG=450
                ANGLE VEC=<1,0,0>
                DIRECTION=CCW
                SHOW FEATURE PARAMETERS=NO
                SHOW CONTACT PARAMETERS=YES
                  NUMHITS=5,DEPTH=0.03,PITCH=0
                  SAMPLE METHOD=SAMPLE_FEATURE
                  SAMPLE FEATURE=NONE
                  AVOIDANCE MOVE=BOTH,DISTANCE BEFORE=0.1,DISTANCE AFTER=0.1
                  FIND HOLE=DISABLED,ONERROR=NO,READ POS=NO
                SHOW HITS=NO
    A19        =FEAT/CONTACT/CIRCLE/DEFAULT,CARTESIAN,IN,LEAST_SQR
                THEO/<12.2,12.4207,0.06>,<0,0,1>,0.48
                ACTL/<12.2,12.4207,0.06>,<0,0,1>,0.48
                TARG/<12.2,12.4207,0.06>,<0,0,1>
                START ANG=28.3,END ANG=330.2
                ANGLE VEC=<1,0,0>
                DIRECTION=CCW
                SHOW FEATURE PARAMETERS=NO
                SHOW CONTACT PARAMETERS=YES
                  NUMHITS=5,DEPTH=0.03,PITCH=0
                  SAMPLE METHOD=SAMPLE_FEATURE
                  SAMPLE FEATURE=NONE
                  AVOIDANCE MOVE=BOTH,DISTANCE BEFORE=0.1,DISTANCE AFTER=0.1
                  FIND HOLE=DISABLED,ONERROR=NO,READ POS=NO
                SHOW HITS=NO
    A19B       =FEAT/CONTACT/CIRCLE/DEFAULT,CARTESIAN,IN,LEAST_SQR
                THEO/<12.57,12.4207,0.06>,<0,0,1>,0.128
                ACTL/<12.57,12.4207,0.06>,<0,0,1>,0.128
                TARG/<12.57,12.4207,0.06>,<0,0,1>
                START ANG=274.4,END ANG=439.3
                ANGLE VEC=<1,0,0>
                DIRECTION=CCW
                SHOW FEATURE PARAMETERS=NO
                SHOW CONTACT PARAMETERS=YES
                  NUMHITS=5,DEPTH=0.02,PITCH=0
                  SAMPLE METHOD=SAMPLE_FEATURE
                  SAMPLE FEATURE=NONE
                  AVOIDANCE MOVE=BOTH,DISTANCE BEFORE=0.1,DISTANCE AFTER=0.1
                  FIND HOLE=DISABLED,ONERROR=NO,READ POS=NO
                SHOW HITS=NO
    G4         =FEAT/CONTACT/SQUARE SLOT/DEFAULT,CARTESIAN,IN
                THEO/<13.2301,13.0707,0.06>,<0,0,1>,<0,-1,0>,0.44,0.94
                ACTL/<13.2301,13.0707,0.06>,<0,0,1>,<0,-1,0>,0.44,0.94
                TARG/<13.2301,13.0707,0.06>,<0,0,1>,<0,-1,0>
                MEAS WIDTH=NO,RADIUS=0.25
                SHOW FEATURE PARAMETERS=NO
                SHOW CONTACT PARAMETERS=YES
                  DEPTH=0.02
                  SAMPLE METHOD=SAMPLE_FEATURE
                  SAMPLE FEATURE=NONE
                  AVOIDANCE MOVE=BOTH,DISTANCE BEFORE=0.1,DISTANCE AFTER=0.1
                  FIND HOLE=DISABLED,ONERROR=NO,READ POS=NO
                SHOW HITS=NO
    CIR1       =FEAT/CONTACT/CIRCLE/DEFAULT,CARTESIAN,IN,LEAST_SQR
                THEO/<14.3851,12.9007,0.06>,<0,0,1>,0.188
                ACTL/<14.3851,12.9007,0.06>,<0,0,1>,0.188
                TARG/<14.3851,12.9007,0.06>,<0,0,1>
                START ANG=0,END ANG=360
                ANGLE VEC=<1,0,0>
                DIRECTION=CCW
                SHOW FEATURE PARAMETERS=NO
                SHOW CONTACT PARAMETERS=YES
                  NUMHITS=5,DEPTH=0.02,PITCH=0
                  SAMPLE METHOD=SAMPLE_FEATURE
                  SAMPLE FEATURE=NONE
                  AVOIDANCE MOVE=BOTH,DISTANCE BEFORE=0.1,DISTANCE AFTER=0.1
                  FIND HOLE=DISABLED,ONERROR=NO,READ POS=NO
                SHOW HITS=NO
    K1         =FEAT/CONTACT/CIRCLE/DEFAULT,CARTESIAN,IN,LEAST_SQR
                THEO/<15.2901,12.9007,0.06>,<0,0,1>,0.5
                ACTL/<15.2901,12.9007,0.06>,<0,0,1>,0.5
                TARG/<15.2901,12.9007,0.06>,<0,0,1>
                START ANG=0,END ANG=360
                ANGLE VEC=<1,0,0>
                DIRECTION=CCW
                SHOW FEATURE PARAMETERS=NO
                SHOW CONTACT PARAMETERS=YES
                  NUMHITS=5,DEPTH=0.02,PITCH=0
                  SAMPLE METHOD=SAMPLE_FEATURE
                  SAMPLE FEATURE=NONE
                  AVOIDANCE MOVE=BOTH,DISTANCE BEFORE=0.1,DISTANCE AFTER=0.1
                  FIND HOLE=DISABLED,ONERROR=NO,READ POS=NO
                SHOW HITS=NO
    CIR2       =FEAT/CONTACT/CIRCLE/DEFAULT,CARTESIAN,IN,LEAST_SQR
                THEO/<16.1951,12.9007,0.06>,<0,0,1>,0.188
                ACTL/<16.1951,12.9007,0.06>,<0,0,1>,0.188
                TARG/<16.1951,12.9007,0.06>,<0,0,1>
                START ANG=0,END ANG=360
                ANGLE VEC=<1,0,0>
                DIRECTION=CCW
                SHOW FEATURE PARAMETERS=NO
                SHOW CONTACT PARAMETERS=YES
                  NUMHITS=5,DEPTH=0.02,PITCH=0
                  SAMPLE METHOD=SAMPLE_FEATURE
                  SAMPLE FEATURE=NONE
                  AVOIDANCE MOVE=BOTH,DISTANCE BEFORE=0.1,DISTANCE AFTER=0.1
                  FIND HOLE=DISABLED,ONERROR=NO,READ POS=NO
                SHOW HITS=NO
    B12        =FEAT/CONTACT/CIRCLE/DEFAULT,CARTESIAN,IN,LEAST_SQR
                THEO/<14.7801,13.9407,0.06>,<0,0,1>,0.2
                ACTL/<14.7801,13.9407,0.06>,<0,0,1>,0.2
                TARG/<14.7801,13.9407,0.06>,<0,0,1>
                START ANG=0,END ANG=360
                ANGLE VEC=<1,0,0>
                DIRECTION=CCW
                SHOW FEATURE PARAMETERS=NO
                SHOW CONTACT PARAMETERS=YES
                  NUMHITS=5,DEPTH=0.02,PITCH=0
                  SAMPLE METHOD=SAMPLE_FEATURE
                  SAMPLE FEATURE=NONE
                  AVOIDANCE MOVE=BOTH,DISTANCE BEFORE=0.1,DISTANCE AFTER=0.1
                  FIND HOLE=DISABLED,ONERROR=NO,READ POS=NO
                SHOW HITS=NO
    SLTS1      =FEAT/CONTACT/SQUARE SLOT/DEFAULT,CARTESIAN,IN
                THEO/<15.7301,13.9207,0.06>,<0,0,1>,<-1,0,0>,0.44,0.94
                ACTL/<15.7301,13.9207,0.06>,<0,0,1>,<-1,0,0>,0.44,0.94
                TARG/<15.7301,13.9207,0.06>,<0,0,1>,<-1,0,0>
                MEAS WIDTH=NO,RADIUS=0.0689
                SHOW FEATURE PARAMETERS=NO
                SHOW CONTACT PARAMETERS=YES
                  DEPTH=0.02
                  SAMPLE METHOD=SAMPLE_FEATURE
                  SAMPLE FEATURE=NONE
                  AVOIDANCE MOVE=BOTH,DISTANCE BEFORE=0.1,DISTANCE AFTER=0.1
                  FIND HOLE=DISABLED,ONERROR=NO,READ POS=NO
                SHOW HITS=NO
    A17        =FEAT/CONTACT/CIRCLE/DEFAULT,CARTESIAN,IN,LEAST_SQR
                THEO/<17.5101,13.7207,0.06>,<0,0,1>,0.48
                ACTL/<17.5101,13.7207,0.06>,<0,0,1>,0.48
                TARG/<17.5101,13.7207,0.06>,<0,0,1>
                START ANG=27.9,END ANG=333.1
                ANGLE VEC=<1,0,0>
                DIRECTION=CCW
                SHOW FEATURE PARAMETERS=NO
                SHOW CONTACT PARAMETERS=YES
                  NUMHITS=5,DEPTH=0.02,PITCH=0
                  SAMPLE METHOD=SAMPLE_FEATURE
                  SAMPLE FEATURE=NONE
                  AVOIDANCE MOVE=BOTH,DISTANCE BEFORE=0.1,DISTANCE AFTER=0.1
                  FIND HOLE=DISABLED,ONERROR=NO,READ POS=NO
                SHOW HITS=NO
    A17B       =FEAT/CONTACT/CIRCLE/DEFAULT,CARTESIAN,IN,LEAST_SQR
                THEO/<17.8801,13.7207,0.06>,<0,0,1>,0.128
                ACTL/<17.8801,13.7207,0.06>,<0,0,1>,0.128
                TARG/<17.8801,13.7207,0.06>,<0,0,1>
                START ANG=270,END ANG=450
                ANGLE VEC=<1,0,0>
                DIRECTION=CCW
                SHOW FEATURE PARAMETERS=NO
                SHOW CONTACT PARAMETERS=YES
                  NUMHITS=5,DEPTH=0.02,PITCH=0
                  SAMPLE METHOD=SAMPLE_FEATURE
                  SAMPLE FEATURE=NONE
                  AVOIDANCE MOVE=BOTH,DISTANCE BEFORE=0.1,DISTANCE AFTER=0.1
                  FIND HOLE=DISABLED,ONERROR=NO,READ POS=NO
                SHOW HITS=NO
    A16        =FEAT/CONTACT/CIRCLE/DEFAULT,CARTESIAN,IN,LEAST_SQR
                THEO/<17.5101,12.4207,0.06>,<0,0,1>,0.48
                ACTL/<17.5101,12.4207,0.06>,<0,0,1>,0.48
                TARG/<17.5101,12.4207,0.06>,<0,0,1>
                START ANG=27.3,END ANG=332.9
                ANGLE VEC=<1,0,0>
                DIRECTION=CCW
                SHOW FEATURE PARAMETERS=NO
                SHOW CONTACT PARAMETERS=YES
                  NUMHITS=5,DEPTH=0.02,PITCH=0
                  SAMPLE METHOD=SAMPLE_FEATURE
                  SAMPLE FEATURE=NONE
                  AVOIDANCE MOVE=BOTH,DISTANCE BEFORE=0.1,DISTANCE AFTER=0.1
                  FIND HOLE=DISABLED,ONERROR=NO,READ POS=NO
                SHOW HITS=NO
    A16B       =FEAT/CONTACT/CIRCLE/DEFAULT,CARTESIAN,IN,LEAST_SQR
                THEO/<17.8801,12.4207,0.06>,<0,0,1>,0.128
                ACTL/<17.8801,12.4207,0.06>,<0,0,1>,0.128
                TARG/<17.8801,12.4207,0.06>,<0,0,1>
                START ANG=270,END ANG=450
                ANGLE VEC=<1,0,0>
                DIRECTION=CCW
                SHOW FEATURE PARAMETERS=NO
                SHOW CONTACT PARAMETERS=YES
                  NUMHITS=5,DEPTH=0.02,PITCH=0
                  SAMPLE METHOD=SAMPLE_FEATURE
                  SAMPLE FEATURE=NONE
                  AVOIDANCE MOVE=BOTH,DISTANCE BEFORE=0.1,DISTANCE AFTER=0.1
                  FIND HOLE=DISABLED,ONERROR=NO,READ POS=NO
                SHOW HITS=NO
    B11        =FEAT/CONTACT/CIRCLE/DEFAULT,CARTESIAN,IN,LEAST_SQR
                THEO/<18.3411,12.0582,0.06>,<0,0,1>,0.2
                ACTL/<18.3411,12.0582,0.06>,<0,0,1>,0.2
                TARG/<18.3411,12.0582,0.06>,<0,0,1>
                START ANG=0,END ANG=360
                ANGLE VEC=<1,0,0>
                DIRECTION=CCW
                SHOW FEATURE PARAMETERS=NO
                SHOW CONTACT PARAMETERS=YES
                  NUMHITS=5,DEPTH=0.03,PITCH=0
                  SAMPLE METHOD=SAMPLE_FEATURE
                  SAMPLE FEATURE=NONE
                  AVOIDANCE MOVE=BOTH,DISTANCE BEFORE=0.1,DISTANCE AFTER=0.1
                  FIND HOLE=DISABLED,ONERROR=NO,READ POS=NO
                SHOW HITS=NO
    B10        =FEAT/CONTACT/CIRCLE/DEFAULT,CARTESIAN,IN,LEAST_SQR
                THEO/<18.3411,14.0832,0.06>,<0,0,1>,0.2
                ACTL/<18.3411,14.0832,0.06>,<0,0,1>,0.2
                TARG/<18.3411,14.0832,0.06>,<0,0,1>
                START ANG=0,END ANG=360
                ANGLE VEC=<1,0,0>
                DIRECTION=CCW
                SHOW FEATURE PARAMETERS=NO
                SHOW CONTACT PARAMETERS=YES
                  NUMHITS=5,DEPTH=0.03,PITCH=0
                  SAMPLE METHOD=SAMPLE_FEATURE
                  SAMPLE FEATURE=NONE
                  AVOIDANCE MOVE=BOTH,DISTANCE BEFORE=0.1,DISTANCE AFTER=0.1
                  FIND HOLE=DISABLED,ONERROR=NO,READ POS=NO
                SHOW HITS=NO
    J1         =FEAT/CONTACT/SQUARE SLOT/DEFAULT,CARTESIAN,IN
                THEO/<19.0601,13.0707,0.06>,<0,0,1>,<1,0,0>,0.28,1.15
                ACTL/<19.0601,13.0707,0.06>,<0,0,1>,<1,0,0>,0.28,1.15
                TARG/<19.0601,13.0707,0.06>,<0,0,1>,<1,0,0>
                MEAS WIDTH=NO,RADIUS=0.0689
                SHOW FEATURE PARAMETERS=NO
                SHOW CONTACT PARAMETERS=YES
                  DEPTH=0.03
                  SAMPLE METHOD=SAMPLE_FEATURE
                  SAMPLE FEATURE=NONE
                  AVOIDANCE MOVE=BOTH,DISTANCE BEFORE=0.1,DISTANCE AFTER=0.1
                  FIND HOLE=DISABLED,ONERROR=NO,READ POS=NO
                SHOW HITS=NO
    H1         =FEAT/CONTACT/CIRCLE/DEFAULT,CARTESIAN,IN,LEAST_SQR
                THEO/<19.6451,12.3207,0.06>,<0,0,1>,0.515
                ACTL/<19.6451,12.3207,0.06>,<0,0,1>,0.515
                TARG/<19.6451,12.3207,0.06>,<0,0,1>
                START ANG=289.3,END ANG=611.8
                ANGLE VEC=<1,0,0>
                DIRECTION=CCW
                SHOW FEATURE PARAMETERS=NO
                SHOW CONTACT PARAMETERS=YES
                  NUMHITS=5,DEPTH=0.03,PITCH=0
                  SAMPLE METHOD=SAMPLE_FEATURE
                  SAMPLE FEATURE=NONE
                  AVOIDANCE MOVE=BOTH,DISTANCE BEFORE=0.1,DISTANCE AFTER=0.1
                  FIND HOLE=DISABLED,ONERROR=NO,READ POS=NO
                SHOW HITS=NO
                MOVE/POINT,NORMAL,<19.6451,12,2.5>
                TIP/T1A10B0, SHANKIJK=0, 0.1736, 0.9848, ANGLE=0
                MOVE/POINT,NORMAL,<19.6451,12,2.5>
    H1B        =FEAT/CONTACT/CIRCLE/DEFAULT,CARTESIAN,IN,LEAST_SQR
                THEO/<19.6451,11.9147,0.06>,<0,0,1>,0.125
                ACTL/<19.6451,11.9147,0.06>,<0,0,1>,0.125
                TARG/<19.6451,11.9147,0.06>,<0,0,1>
                START ANG=180.0026,END ANG=360
                ANGLE VEC=<1,0,0>
                DIRECTION=CCW
                SHOW FEATURE PARAMETERS=NO
                SHOW CONTACT PARAMETERS=YES
                  NUMHITS=5,DEPTH=0.03,PITCH=0
                  SAMPLE METHOD=SAMPLE_FEATURE
                  SAMPLE FEATURE=NONE
                  AVOIDANCE MOVE=BOTH,DISTANCE BEFORE=0.1,DISTANCE AFTER=0.1
                  FIND HOLE=DISABLED,ONERROR=NO,READ POS=NO
                SHOW HITS=NO
                MOVE/POINT,NORMAL,<19.6451,12,2.5>
                TIP/T1A-10B0, SHANKIJK=0, -0.1736, 0.9848, ANGLE=0
                MOVE/POINT,NORMAL,<19.6451,14,2.5>
    H2B        =FEAT/CONTACT/CIRCLE/DEFAULT,CARTESIAN,IN,LEAST_SQR
                THEO/<19.6451,14.2267,0.06>,<0,0,1>,0.125
                ACTL/<19.6451,14.2267,0.06>,<0,0,1>,0.125
                TARG/<19.6451,14.2267,0.06>,<0,0,1>
                START ANG=0.0802,END ANG=179.9993
                ANGLE VEC=<1,0,0>
                DIRECTION=CCW
                SHOW FEATURE PARAMETERS=NO
                SHOW CONTACT PARAMETERS=YES
                  NUMHITS=5,DEPTH=0.03,PITCH=0
                  SAMPLE METHOD=SAMPLE_FEATURE
                  SAMPLE FEATURE=NONE
                  AVOIDANCE MOVE=BOTH,DISTANCE BEFORE=0.1,DISTANCE AFTER=0.1
                  FIND HOLE=DISABLED,ONERROR=NO,READ POS=NO
                SHOW HITS=NO
                MOVE/POINT,NORMAL,<19.6451,14,2.5>
                TIP/T1A0B0, SHANKIJK=0, 0, 1, ANGLE=0
    H2         =FEAT/CONTACT/CIRCLE/DEFAULT,CARTESIAN,IN,LEAST_SQR
                THEO/<19.6451,13.8207,0.06>,<0,0,1>,0.473
                ACTL/<19.6451,13.8207,0.06>,<0,0,1>,0.473
                TARG/<19.6451,13.8207,0.06>,<0,0,1>
                START ANG=109.3,END ANG=430.9
                ANGLE VEC=<1,0,0>
                DIRECTION=CCW
                SHOW FEATURE PARAMETERS=NO
                SHOW CONTACT PARAMETERS=YES
                  NUMHITS=5,DEPTH=0.03,PITCH=0
                  SAMPLE METHOD=SAMPLE_FEATURE
                  SAMPLE FEATURE=NONE
                  AVOIDANCE MOVE=BOTH,DISTANCE BEFORE=0.1,DISTANCE AFTER=0.1
                  FIND HOLE=DISABLED,ONERROR=NO,READ POS=NO
                SHOW HITS=NO
    G1         =FEAT/CONTACT/SQUARE SLOT/DEFAULT,CARTESIAN,IN
                THEO/<20.4751,13.6707,0.06>,<0,0,1>,<0,1,0>,0.44,0.94
                ACTL/<20.4751,13.6707,0.06>,<0,0,1>,<0,1,0>,0.44,0.94
                TARG/<20.4751,13.6707,0.06>,<0,0,1>,<0,1,0>
                MEAS WIDTH=NO,RADIUS=0.0689
                SHOW FEATURE PARAMETERS=NO
                SHOW CONTACT PARAMETERS=YES
                  DEPTH=0.03
                  SAMPLE METHOD=SAMPLE_FEATURE
                  SAMPLE FEATURE=NONE
                  AVOIDANCE MOVE=BOTH,DISTANCE BEFORE=0.1,DISTANCE AFTER=0.1
                  FIND HOLE=DISABLED,ONERROR=NO,READ POS=NO
                SHOW HITS=NO
    G2         =FEAT/CONTACT/SQUARE SLOT/DEFAULT,CARTESIAN,IN
                THEO/<20.4751,12.4707,0.06>,<0,0,1>,<0,1,0>,0.44,0.94
                ACTL/<20.4751,12.4707,0.06>,<0,0,1>,<0,1,0>,0.44,0.94
                TARG/<20.4751,12.4707,0.06>,<0,0,1>,<0,1,0>
                MEAS WIDTH=NO,RADIUS=0.0689
                SHOW FEATURE PARAMETERS=NO
                SHOW CONTACT PARAMETERS=YES
                  DEPTH=0.03
                  SAMPLE METHOD=SAMPLE_FEATURE
                  SAMPLE FEATURE=NONE
                  AVOIDANCE MOVE=BOTH,DISTANCE BEFORE=0.1,DISTANCE AFTER=0.1
                  FIND HOLE=DISABLED,ONERROR=NO,READ POS=NO
                SHOW HITS=NO
    G5         =FEAT/CONTACT/SQUARE SLOT/DEFAULT,CARTESIAN,IN
                THEO/<21.525,13.6207,0.06>,<0,0,1>,<0,-1,0>,0.44,0.94
                ACTL/<21.525,13.6207,0.06>,<0,0,1>,<0,-1,0>,0.44,0.94
                TARG/<21.525,13.6207,0.06>,<0,0,1>,<0,-1,0>
                MEAS WIDTH=NO,RADIUS=0.0689
                SHOW FEATURE PARAMETERS=NO
                SHOW CONTACT PARAMETERS=YES
                  DEPTH=0.03
                  SAMPLE METHOD=SAMPLE_FEATURE
                  SAMPLE FEATURE=NONE
                  AVOIDANCE MOVE=BOTH,DISTANCE BEFORE=0.1,DISTANCE AFTER=0.1
                  FIND HOLE=DISABLED,ONERROR=NO,READ POS=NO
                SHOW HITS=NO
    A15        =FEAT/CONTACT/CIRCLE/DEFAULT,CARTESIAN,IN,LEAST_SQR
                THEO/<21.525,12.4707,0.06>,<0,0,1>,0.515
                ACTL/<21.525,12.4707,0.06>,<0,0,1>,0.515
                TARG/<21.525,12.4707,0.06>,<0,0,1>
                START ANG=20.7,END ANG=338.7
                ANGLE VEC=<1,0,0>
                DIRECTION=CCW
                SHOW FEATURE PARAMETERS=NO
                SHOW CONTACT PARAMETERS=YES
                  NUMHITS=5,DEPTH=0.03,PITCH=0
                  SAMPLE METHOD=SAMPLE_FEATURE
                  SAMPLE FEATURE=NONE
                  AVOIDANCE MOVE=BOTH,DISTANCE BEFORE=0.1,DISTANCE AFTER=0.1
                  FIND HOLE=DISABLED,ONERROR=NO,READ POS=NO
                SHOW HITS=NO
    A15B       =FEAT/CONTACT/CIRCLE/DEFAULT,CARTESIAN,IN,LEAST_SQR
                THEO/<21.931,12.4707,0.06>,<0,0,1>,0.125
                ACTL/<21.931,12.4707,0.06>,<0,0,1>,0.125
                TARG/<21.931,12.4707,0.06>,<0,0,1>
                START ANG=270.0072,END ANG=449.9909
                ANGLE VEC=<1,0,0>
                DIRECTION=CCW
                SHOW FEATURE PARAMETERS=NO
                SHOW CONTACT PARAMETERS=YES
                  NUMHITS=5,DEPTH=0.03,PITCH=0
                  SAMPLE METHOD=SAMPLE_FEATURE
                  SAMPLE FEATURE=NONE
                  AVOIDANCE MOVE=BOTH,DISTANCE BEFORE=0.1,DISTANCE AFTER=0.1
                  FIND HOLE=DISABLED,ONERROR=NO,READ POS=NO
                SHOW HITS=NO
    E2         =FEAT/CONTACT/CIRCLE/DEFAULT,CARTESIAN,IN,LEAST_SQR
                THEO/<22.8012,13.9728,0.06>,<0,0,1>,0.5
                ACTL/<22.8012,13.9728,0.06>,<0,0,1>,0.5
                TARG/<22.8012,13.9728,0.06>,<0,0,1>
                START ANG=0,END ANG=360
                ANGLE VEC=<1,0,0>
                DIRECTION=CCW
                SHOW FEATURE PARAMETERS=NO
                SHOW CONTACT PARAMETERS=YES
                  NUMHITS=5,DEPTH=0.02,PITCH=0
                  SAMPLE METHOD=SAMPLE_FEATURE
                  SAMPLE FEATURE=NONE
                  AVOIDANCE MOVE=BOTH,DISTANCE BEFORE=0.1,DISTANCE AFTER=0.1
                  FIND HOLE=DISABLED,ONERROR=NO,READ POS=NO
                SHOW HITS=NO
    A14        =FEAT/CONTACT/CIRCLE/DEFAULT,CARTESIAN,IN,LEAST_SQR
                THEO/<23.4201,13.0707,0.06>,<0,0,1>,0.48
                ACTL/<23.4201,13.0707,0.06>,<0,0,1>,0.48
                TARG/<23.4201,13.0707,0.06>,<0,0,1>
                START ANG=29.2,END ANG=330.3
                ANGLE VEC=<1,0,0>
                DIRECTION=CCW
                SHOW FEATURE PARAMETERS=NO
                SHOW CONTACT PARAMETERS=YES
                  NUMHITS=5,DEPTH=0.025,PITCH=0
                  SAMPLE METHOD=SAMPLE_FEATURE
                  SAMPLE FEATURE=NONE
                  AVOIDANCE MOVE=BOTH,DISTANCE BEFORE=0.1,DISTANCE AFTER=0.1
                  FIND HOLE=DISABLED,ONERROR=NO,READ POS=NO
                SHOW HITS=NO
    A14B       =FEAT/CONTACT/CIRCLE/DEFAULT,CARTESIAN,IN,LEAST_SQR
                THEO/<23.7901,13.0707,0.06>,<0,0,1>,0.128
                ACTL/<23.7901,13.0707,0.06>,<0,0,1>,0.128
                TARG/<23.7901,13.0707,0.06>,<0,0,1>
                START ANG=270,END ANG=450
                ANGLE VEC=<1,0,0>
                DIRECTION=CCW
                SHOW FEATURE PARAMETERS=NO
                SHOW CONTACT PARAMETERS=YES
                  NUMHITS=5,DEPTH=0.02,PITCH=0
                  SAMPLE METHOD=SAMPLE_FEATURE
                  SAMPLE FEATURE=NONE
                  AVOIDANCE MOVE=BOTH,DISTANCE BEFORE=0.1,DISTANCE AFTER=0.1
                  FIND HOLE=DISABLED,ONERROR=NO,READ POS=NO
                SHOW HITS=NO
    B9         =FEAT/CONTACT/CIRCLE/DEFAULT,CARTESIAN,IN,LEAST_SQR
                THEO/<24.21,12.0582,0.06>,<0,0,1>,0.2
                ACTL/<24.21,12.0582,0.06>,<0,0,1>,0.2
                TARG/<24.21,12.0582,0.06>,<0,0,1>
                START ANG=0,END ANG=360
                ANGLE VEC=<1,0,0>
                DIRECTION=CCW
                SHOW FEATURE PARAMETERS=NO
                SHOW CONTACT PARAMETERS=YES
                  NUMHITS=5,DEPTH=0.02,PITCH=0
                  SAMPLE METHOD=SAMPLE_FEATURE
                  SAMPLE FEATURE=NONE
                  AVOIDANCE MOVE=BOTH,DISTANCE BEFORE=0.1,DISTANCE AFTER=0.1
                  FIND HOLE=DISABLED,ONERROR=NO,READ POS=NO
                SHOW HITS=NO
    B8         =FEAT/CONTACT/CIRCLE/DEFAULT,CARTESIAN,IN,LEAST_SQR
                THEO/<24.21,14.0832,0.06>,<0,0,1>,0.2
                ACTL/<24.21,14.0832,0.06>,<0,0,1>,0.2
                TARG/<24.21,14.0832,0.06>,<0,0,1>
                START ANG=0,END ANG=360
                ANGLE VEC=<1,0,0>
                DIRECTION=CCW
                SHOW FEATURE PARAMETERS=NO
                SHOW CONTACT PARAMETERS=YES
                  NUMHITS=5,DEPTH=0.02,PITCH=0
                  SAMPLE METHOD=SAMPLE_FEATURE
                  SAMPLE FEATURE=NONE
                  AVOIDANCE MOVE=BOTH,DISTANCE BEFORE=0.1,DISTANCE AFTER=0.1
                  FIND HOLE=DISABLED,ONERROR=NO,READ POS=NO
                SHOW HITS=NO
                MOVE/POINT,NORMAL,<24.21,14.0832,2.2>
                MOVE/POINT,NORMAL,<25.2358,25.2551,2.2>
    CIR4       =FEAT/CONTACT/CIRCLE/DEFAULT,CARTESIAN,IN,LEAST_SQR
                THEO/<25.2358,25.2551,0.113>,<0,0,1>,0.4755
                ACTL/<25.2358,25.2551,0.113>,<0,0,1>,0.4755
                TARG/<25.2358,25.2551,0.113>,<0,0,1>
                START ANG=0,END ANG=360
                ANGLE VEC=<1,0,0>
                DIRECTION=CCW
                SHOW FEATURE PARAMETERS=NO
                SHOW CONTACT PARAMETERS=YES
                  NUMHITS=5,DEPTH=0.02,PITCH=0
                  SAMPLE METHOD=SAMPLE_FEATURE
                  SAMPLE FEATURE=NONE
                  AVOIDANCE MOVE=BOTH,DISTANCE BEFORE=0.1,DISTANCE AFTER=0.1
                  FIND HOLE=DISABLED,ONERROR=NO,READ POS=NO
                SHOW HITS=NO
    B7         =FEAT/CONTACT/CIRCLE/DEFAULT,CARTESIAN,IN,LEAST_SQR
                THEO/<24.8911,24.4407,0.113>,<0,0,1>,0.2
                ACTL/<24.8911,24.4407,0.113>,<0,0,1>,0.2
                TARG/<24.8911,24.4407,0.113>,<0,0,1>
                START ANG=0,END ANG=360
                ANGLE VEC=<1,0,0>
                DIRECTION=CCW
                SHOW FEATURE PARAMETERS=NO
                SHOW CONTACT PARAMETERS=YES
                  NUMHITS=5,DEPTH=0.03,PITCH=0
                  SAMPLE METHOD=SAMPLE_FEATURE
                  SAMPLE FEATURE=NONE
                  AVOIDANCE MOVE=BOTH,DISTANCE BEFORE=0.1,DISTANCE AFTER=0.1
                  FIND HOLE=DISABLED,ONERROR=NO,READ POS=NO
                SHOW HITS=NO
    A13        =FEAT/CONTACT/CIRCLE/DEFAULT,CARTESIAN,IN,LEAST_SQR
                THEO/<25.6611,21.78,0.113>,<0,0,1>,0.48
                ACTL/<25.6611,21.78,0.113>,<0,0,1>,0.48
                TARG/<25.6611,21.78,0.113>,<0,0,1>
                START ANG=27.3,END ANG=333
                ANGLE VEC=<1,0,0>
                DIRECTION=CCW
                SHOW FEATURE PARAMETERS=NO
                SHOW CONTACT PARAMETERS=YES
                  NUMHITS=5,DEPTH=0.03,PITCH=0
                  SAMPLE METHOD=SAMPLE_FEATURE
                  SAMPLE FEATURE=NONE
                  AVOIDANCE MOVE=BOTH,DISTANCE BEFORE=0.1,DISTANCE AFTER=0.1
                  FIND HOLE=DISABLED,ONERROR=NO,READ POS=NO
                SHOW HITS=NO
    A13B       =FEAT/CONTACT/CIRCLE/DEFAULT,CARTESIAN,IN,LEAST_SQR
                THEO/<26.0311,21.78,0.113>,<0,0,1>,0.1327
                ACTL/<26.0311,21.78,0.113>,<0,0,1>,0.1327
                TARG/<26.0311,21.78,0.113>,<0,0,1>
                START ANG=270,END ANG=450
                ANGLE VEC=<1,0,0>
                DIRECTION=CCW
                SHOW FEATURE PARAMETERS=NO
                SHOW CONTACT PARAMETERS=YES
                  NUMHITS=5,DEPTH=0.03,PITCH=0
                  SAMPLE METHOD=SAMPLE_FEATURE
                  SAMPLE FEATURE=NONE
                  AVOIDANCE MOVE=BOTH,DISTANCE BEFORE=0.1,DISTANCE AFTER=0.1
                  FIND HOLE=DISABLED,ONERROR=NO,READ POS=NO
                SHOW HITS=NO
    A12        =FEAT/CONTACT/CIRCLE/DEFAULT,CARTESIAN,IN,LEAST_SQR
                THEO/<25.6611,20.75,0.113>,<0,0,1>,0.48
                ACTL/<25.6611,20.75,0.113>,<0,0,1>,0.48
                TARG/<25.6611,20.75,0.113>,<0,0,1>
                START ANG=28.3,END ANG=332.3
                ANGLE VEC=<1,0,0>
                DIRECTION=CCW
                SHOW FEATURE PARAMETERS=NO
                SHOW CONTACT PARAMETERS=YES
                  NUMHITS=5,DEPTH=0.03,PITCH=0
                  SAMPLE METHOD=SAMPLE_FEATURE
                  SAMPLE FEATURE=NONE
                  AVOIDANCE MOVE=BOTH,DISTANCE BEFORE=0.1,DISTANCE AFTER=0.1
                  FIND HOLE=DISABLED,ONERROR=NO,READ POS=NO
                SHOW HITS=NO
    A12B       =FEAT/CONTACT/CIRCLE/DEFAULT,CARTESIAN,IN,LEAST_SQR
                THEO/<26.0311,20.75,0.113>,<0,0,1>,0.128
                ACTL/<26.0311,20.75,0.113>,<0,0,1>,0.128
                TARG/<26.0311,20.75,0.113>,<0,0,1>
                START ANG=270,END ANG=450
                ANGLE VEC=<1,0,0>
                DIRECTION=CCW
                SHOW FEATURE PARAMETERS=NO
                SHOW CONTACT PARAMETERS=YES
                  NUMHITS=5,DEPTH=0.03,PITCH=0
                  SAMPLE METHOD=SAMPLE_FEATURE
                  SAMPLE FEATURE=NONE
                  AVOIDANCE MOVE=BOTH,DISTANCE BEFORE=0.1,DISTANCE AFTER=0.1
                  FIND HOLE=DISABLED,ONERROR=NO,READ POS=NO
                SHOW HITS=NO
    B6         =FEAT/CONTACT/CIRCLE/DEFAULT,CARTESIAN,IN,LEAST_SQR
                THEO/<26.8011,20.1407,0.113>,<0,0,1>,0.2
                ACTL/<26.8011,20.1407,0.113>,<0,0,1>,0.2
                TARG/<26.8011,20.1407,0.113>,<0,0,1>
                START ANG=0,END ANG=360
                ANGLE VEC=<1,0,0>
                DIRECTION=CCW
                SHOW FEATURE PARAMETERS=NO
                SHOW CONTACT PARAMETERS=YES
                  NUMHITS=5,DEPTH=0.03,PITCH=0
                  SAMPLE METHOD=SAMPLE_FEATURE
                  SAMPLE FEATURE=NONE
                  AVOIDANCE MOVE=BOTH,DISTANCE BEFORE=0.1,DISTANCE AFTER=0.1
                  FIND HOLE=DISABLED,ONERROR=NO,READ POS=NO
                SHOW HITS=NO
    E1         =FEAT/CONTACT/CIRCLE/DEFAULT,CARTESIAN,IN,LEAST_SQR
                THEO/<27.1911,19.1907,0.113>,<0,0,1>,0.5
                ACTL/<27.1911,19.1907,0.113>,<0,0,1>,0.5
                TARG/<27.1911,19.1907,0.113>,<0,0,1>
                START ANG=0,END ANG=360
                ANGLE VEC=<1,0,0>
                DIRECTION=CCW
                SHOW FEATURE PARAMETERS=NO
                SHOW CONTACT PARAMETERS=YES
                  NUMHITS=5,DEPTH=0.03,PITCH=0
                  SAMPLE METHOD=SAMPLE_FEATURE
                  SAMPLE FEATURE=NONE
                  AVOIDANCE MOVE=BOTH,DISTANCE BEFORE=0.1,DISTANCE AFTER=0.1
                  FIND HOLE=DISABLED,ONERROR=NO,READ POS=NO
                SHOW HITS=NO
    A10        =FEAT/CONTACT/CIRCLE/DEFAULT,CARTESIAN,IN,LEAST_SQR
                THEO/<25.6611,18.9607,0.113>,<0,0,1>,0.48
                ACTL/<25.6611,18.9607,0.113>,<0,0,1>,0.48
                TARG/<25.6611,18.9607,0.113>,<0,0,1>
                START ANG=24.4145,END ANG=328.6
                ANGLE VEC=<1,0,0>
                DIRECTION=CCW
                SHOW FEATURE PARAMETERS=NO
                SHOW CONTACT PARAMETERS=YES
                  NUMHITS=5,DEPTH=0.03,PITCH=0
                  SAMPLE METHOD=SAMPLE_FEATURE
                  SAMPLE FEATURE=NONE
                  AVOIDANCE MOVE=BOTH,DISTANCE BEFORE=0.1,DISTANCE AFTER=0.1
                  FIND HOLE=DISABLED,ONERROR=NO,READ POS=NO
                SHOW HITS=NO
    A10B       =FEAT/CONTACT/CIRCLE/DEFAULT,CARTESIAN,IN,LEAST_SQR
                THEO/<26.0311,18.9607,0.113>,<0,0,1>,0.128
                ACTL/<26.0311,18.9607,0.113>,<0,0,1>,0.128
                TARG/<26.0311,18.9607,0.113>,<0,0,1>
                START ANG=270,END ANG=450
                ANGLE VEC=<1,0,0>
                DIRECTION=CCW
                SHOW FEATURE PARAMETERS=NO
                SHOW CONTACT PARAMETERS=YES
                  NUMHITS=5,DEPTH=0.03,PITCH=0
                  SAMPLE METHOD=SAMPLE_FEATURE
                  SAMPLE FEATURE=NONE
                  AVOIDANCE MOVE=BOTH,DISTANCE BEFORE=0.1,DISTANCE AFTER=0.1
                  FIND HOLE=DISABLED,ONERROR=NO,READ POS=NO
                SHOW HITS=NO
    A9         =FEAT/CONTACT/CIRCLE/DEFAULT,CARTESIAN,IN,LEAST_SQR
                THEO/<25.6611,17.9507,0.113>,<0,0,1>,0.48
                ACTL/<25.6611,17.9507,0.113>,<0,0,1>,0.48
                TARG/<25.6611,17.9507,0.113>,<0,0,1>
                START ANG=27.3,END ANG=331.1
                ANGLE VEC=<1,0,0>
                DIRECTION=CCW
                SHOW FEATURE PARAMETERS=NO
                SHOW CONTACT PARAMETERS=YES
                  NUMHITS=5,DEPTH=0.03,PITCH=0
                  SAMPLE METHOD=SAMPLE_FEATURE
                  SAMPLE FEATURE=NONE
                  AVOIDANCE MOVE=BOTH,DISTANCE BEFORE=0.1,DISTANCE AFTER=0.1
                  FIND HOLE=DISABLED,ONERROR=NO,READ POS=NO
                SHOW HITS=NO
    A9B        =FEAT/CONTACT/CIRCLE/DEFAULT,CARTESIAN,IN,LEAST_SQR
                THEO/<26.0311,17.9507,0.113>,<0,0,1>,0.128
                ACTL/<26.0311,17.9507,0.113>,<0,0,1>,0.128
                TARG/<26.0311,17.9507,0.113>,<0,0,1>
                START ANG=270,END ANG=450
                ANGLE VEC=<1,0,0>
                DIRECTION=CCW
                SHOW FEATURE PARAMETERS=NO
                SHOW CONTACT PARAMETERS=YES
                  NUMHITS=5,DEPTH=0.03,PITCH=0
                  SAMPLE METHOD=SAMPLE_FEATURE
                  SAMPLE FEATURE=NONE
                  AVOIDANCE MOVE=BOTH,DISTANCE BEFORE=0.1,DISTANCE AFTER=0.1
                  FIND HOLE=DISABLED,ONERROR=NO,READ POS=NO
                SHOW HITS=NO
    A8         =FEAT/CONTACT/CIRCLE/DEFAULT,CARTESIAN,IN,LEAST_SQR
                THEO/<25.6611,16.9507,0.113>,<0,0,1>,0.48
                ACTL/<25.6611,16.9507,0.113>,<0,0,1>,0.48
                TARG/<25.6611,16.9507,0.113>,<0,0,1>
                START ANG=24.4145,END ANG=335.5857
                ANGLE VEC=<1,0,0>
                DIRECTION=CCW
                SHOW FEATURE PARAMETERS=NO
                SHOW CONTACT PARAMETERS=YES
                  NUMHITS=5,DEPTH=0.03,PITCH=0
                  SAMPLE METHOD=SAMPLE_FEATURE
                  SAMPLE FEATURE=NONE
                  AVOIDANCE MOVE=BOTH,DISTANCE BEFORE=0.1,DISTANCE AFTER=0.1
                  FIND HOLE=DISABLED,ONERROR=NO,READ POS=NO
                SHOW HITS=NO
    A8B        =FEAT/CONTACT/CIRCLE/DEFAULT,CARTESIAN,IN,LEAST_SQR
                THEO/<26.0311,16.9507,0.113>,<0,0,1>,0.128
                ACTL/<26.0311,16.9507,0.113>,<0,0,1>,0.128
                TARG/<26.0311,16.9507,0.113>,<0,0,1>
                START ANG=270,END ANG=450
                ANGLE VEC=<1,0,0>
                DIRECTION=CCW
                SHOW FEATURE PARAMETERS=NO
                SHOW CONTACT PARAMETERS=YES
                  NUMHITS=5,DEPTH=0.03,PITCH=0
                  SAMPLE METHOD=SAMPLE_FEATURE
                  SAMPLE FEATURE=NONE
                  AVOIDANCE MOVE=BOTH,DISTANCE BEFORE=0.1,DISTANCE AFTER=0.1
                  FIND HOLE=DISABLED,ONERROR=NO,READ POS=NO
                SHOW HITS=NO
    CIR7       =FEAT/CONTACT/CIRCLE/DEFAULT,CARTESIAN,IN,LEAST_SQR
                THEO/<27.1161,16.1107,0.113>,<0,0,1>,0.188
                ACTL/<27.1161,16.1107,0.113>,<0,0,1>,0.188
                TARG/<27.1161,16.1107,0.113>,<0,0,1>
                START ANG=0,END ANG=360
                ANGLE VEC=<1,0,0>
                DIRECTION=CCW
                SHOW FEATURE PARAMETERS=NO
                SHOW CONTACT PARAMETERS=YES
                  NUMHITS=5,DEPTH=0.03,PITCH=0
                  SAMPLE METHOD=SAMPLE_FEATURE
                  SAMPLE FEATURE=NONE
                  AVOIDANCE MOVE=BOTH,DISTANCE BEFORE=0.1,DISTANCE AFTER=0.1
                  FIND HOLE=DISABLED,ONERROR=NO,READ POS=NO
                SHOW HITS=NO
    D1         =FEAT/CONTACT/SQUARE SLOT/DEFAULT,CARTESIAN,IN
                THEO/<26.2111,16.1107,0.113>,<0,0,1>,<-1,0,0>,0.46,0.92
                ACTL/<26.2111,16.1107,0.113>,<0,0,1>,<-1,0,0>,0.46,0.92
                TARG/<26.2111,16.1107,0.113>,<0,0,1>,<-1,0,0>
                MEAS WIDTH=NO,RADIUS=0.0689
                SHOW FEATURE PARAMETERS=NO
                SHOW CONTACT PARAMETERS=YES
                  DEPTH=0.03
                  SAMPLE METHOD=SAMPLE_FEATURE
                  SAMPLE FEATURE=NONE
                  AVOIDANCE MOVE=BOTH,DISTANCE BEFORE=0.1,DISTANCE AFTER=0.1
                  FIND HOLE=DISABLED,ONERROR=NO,READ POS=NO
                SHOW HITS=NO
    CIR8       =FEAT/CONTACT/CIRCLE/DEFAULT,CARTESIAN,IN,LEAST_SQR
                THEO/<25.3061,16.1107,0.113>,<0,0,1>,0.188
                ACTL/<25.3061,16.1107,0.113>,<0,0,1>,0.188
                TARG/<25.3061,16.1107,0.113>,<0,0,1>
                START ANG=0,END ANG=360
                ANGLE VEC=<1,0,0>
                DIRECTION=CCW
                SHOW FEATURE PARAMETERS=NO
                SHOW CONTACT PARAMETERS=YES
                  NUMHITS=5,DEPTH=0.03,PITCH=0
                  SAMPLE METHOD=SAMPLE_FEATURE
                  SAMPLE FEATURE=NONE
                  AVOIDANCE MOVE=BOTH,DISTANCE BEFORE=0.1,DISTANCE AFTER=0.1
                  FIND HOLE=DISABLED,ONERROR=NO,READ POS=NO
                SHOW HITS=NO
    B5         =FEAT/CONTACT/CIRCLE/DEFAULT,CARTESIAN,IN,LEAST_SQR
                THEO/<24.8911,16.1707,0.113>,<0,0,1>,0.2
                ACTL/<24.8911,16.1707,0.113>,<0,0,1>,0.2
                TARG/<24.8911,16.1707,0.113>,<0,0,1>
                START ANG=0,END ANG=360
                ANGLE VEC=<1,0,0>
                DIRECTION=CCW
                SHOW FEATURE PARAMETERS=NO
                SHOW CONTACT PARAMETERS=YES
                  NUMHITS=5,DEPTH=0.03,PITCH=0
                  SAMPLE METHOD=SAMPLE_FEATURE
                  SAMPLE FEATURE=NONE
                  AVOIDANCE MOVE=BOTH,DISTANCE BEFORE=0.1,DISTANCE AFTER=0.1
                  FIND HOLE=DISABLED,ONERROR=NO,READ POS=NO
                SHOW HITS=NO
    C1         =FEAT/CONTACT/CIRCLE/DEFAULT,CARTESIAN,IN,LEAST_SQR
                THEO/<25.87,14.0707,0.113>,<0,0,1>,0.5
                ACTL/<25.87,14.0707,0.113>,<0,0,1>,0.5
                TARG/<25.87,14.0707,0.113>,<0,0,1>
                START ANG=58.2,END ANG=361.3
                ANGLE VEC=<1,0,0>
                DIRECTION=CCW
                SHOW FEATURE PARAMETERS=NO
                SHOW CONTACT PARAMETERS=YES
                  NUMHITS=5,DEPTH=0.03,PITCH=0
                  SAMPLE METHOD=SAMPLE_FEATURE
                  SAMPLE FEATURE=NONE
                  AVOIDANCE MOVE=BOTH,DISTANCE BEFORE=0.1,DISTANCE AFTER=0.1
                  FIND HOLE=DISABLED,ONERROR=NO,READ POS=NO
                SHOW HITS=NO
    C1B        =FEAT/CONTACT/CIRCLE/DEFAULT,CARTESIAN,IN,LEAST_SQR
                THEO/<26.3203,14.3307,0.113>,<0,0,1>,0.16
                ACTL/<26.3203,14.3307,0.113>,<0,0,1>,0.16
                TARG/<26.3203,14.3307,0.113>,<0,0,1>
                START ANG=300.0004,END ANG=480
                ANGLE VEC=<1,0,0>
                DIRECTION=CCW
                SHOW FEATURE PARAMETERS=NO
                SHOW CONTACT PARAMETERS=YES
                  NUMHITS=5,DEPTH=0.03,PITCH=0
                  SAMPLE METHOD=SAMPLE_FEATURE
                  SAMPLE FEATURE=NONE
                  AVOIDANCE MOVE=BOTH,DISTANCE BEFORE=0.1,DISTANCE AFTER=0.1
                  FIND HOLE=DISABLED,ONERROR=NO,READ POS=NO
                SHOW HITS=NO
    B4         =FEAT/CONTACT/CIRCLE/DEFAULT,CARTESIAN,IN,LEAST_SQR
                THEO/<27.5701,13.4619,0.113>,<0,0,1>,0.2
                ACTL/<27.5701,13.4619,0.113>,<0,0,1>,0.2
                TARG/<27.5701,13.4619,0.113>,<0,0,1>
                START ANG=0,END ANG=360
                ANGLE VEC=<1,0,0>
                DIRECTION=CCW
                SHOW FEATURE PARAMETERS=NO
                SHOW CONTACT PARAMETERS=YES
                  NUMHITS=5,DEPTH=0.03,PITCH=0
                  SAMPLE METHOD=SAMPLE_FEATURE
                  SAMPLE FEATURE=NONE
                  AVOIDANCE MOVE=BOTH,DISTANCE BEFORE=0.1,DISTANCE AFTER=0.1
                  FIND HOLE=DISABLED,ONERROR=NO,READ POS=NO
                SHOW HITS=NO
    CIR10      =FEAT/CONTACT/CIRCLE/DEFAULT,CARTESIAN,IN,LEAST_SQR
                THEO/<26.653,13.1214,0.113>,<0,0,1>,0.0938
                ACTL/<26.653,13.1214,0.113>,<0,0,1>,0.0938
                TARG/<26.653,13.1214,0.113>,<0,0,1>
                START ANG=0,END ANG=360
                ANGLE VEC=<1,0,0>
                DIRECTION=CCW
                SHOW FEATURE PARAMETERS=NO
                SHOW CONTACT PARAMETERS=YES
                  NUMHITS=5,DEPTH=0.03,PITCH=0
                  SAMPLE METHOD=SAMPLE_FEATURE
                  SAMPLE FEATURE=NONE
                  AVOIDANCE MOVE=BOTH,DISTANCE BEFORE=0.1,DISTANCE AFTER=0.1
                  FIND HOLE=DISABLED,ONERROR=NO,READ POS=NO
                SHOW HITS=NO
    A7B        =FEAT/CONTACT/CIRCLE/DEFAULT,CARTESIAN,IN,LEAST_SQR
                THEO/<25.6611,12.2907,0.113>,<0,0,1>,0.128
                ACTL/<25.6611,12.2907,0.113>,<0,0,1>,0.128
                TARG/<25.6611,12.2907,0.113>,<0,0,1>
                START ANG=359.9984,END ANG=539.9999
                ANGLE VEC=<1,0,0>
                DIRECTION=CCW
                SHOW FEATURE PARAMETERS=NO
                SHOW CONTACT PARAMETERS=YES
                  NUMHITS=5,DEPTH=0.03,PITCH=0
                  SAMPLE METHOD=SAMPLE_FEATURE
                  SAMPLE FEATURE=NONE
                  AVOIDANCE MOVE=BOTH,DISTANCE BEFORE=0.1,DISTANCE AFTER=0.1
                  FIND HOLE=DISABLED,ONERROR=NO,READ POS=NO
                SHOW HITS=NO
    A7         =FEAT/CONTACT/CIRCLE/DEFAULT,CARTESIAN,IN,LEAST_SQR
                THEO/<25.6611,11.9207,0.113>,<0,0,1>,0.48
                ACTL/<25.6611,11.9207,0.113>,<0,0,1>,0.48
                TARG/<25.6611,11.9207,0.113>,<0,0,1>
                START ANG=117.2,END ANG=423.9
                ANGLE VEC=<1,0,0>
                DIRECTION=CCW
                SHOW FEATURE PARAMETERS=NO
                SHOW CONTACT PARAMETERS=YES
                  NUMHITS=5,DEPTH=0.03,PITCH=0
                  SAMPLE METHOD=SAMPLE_FEATURE
                  SAMPLE FEATURE=NONE
                  AVOIDANCE MOVE=BOTH,DISTANCE BEFORE=0.1,DISTANCE AFTER=0.1
                  FIND HOLE=DISABLED,ONERROR=NO,READ POS=NO
                SHOW HITS=NO
    C2B        =FEAT/CONTACT/CIRCLE/DEFAULT,CARTESIAN,IN,LEAST_SQR
                THEO/<26.3203,9.9007,0.113>,<0,0,1>,0.16
                ACTL/<26.3203,9.9007,0.113>,<0,0,1>,0.16
                TARG/<26.3203,9.9007,0.113>,<0,0,1>
                START ANG=300.0004,END ANG=480
                ANGLE VEC=<1,0,0>
                DIRECTION=CCW
                SHOW FEATURE PARAMETERS=NO
                SHOW CONTACT PARAMETERS=YES
                  NUMHITS=5,DEPTH=0.03,PITCH=0
                  SAMPLE METHOD=SAMPLE_FEATURE
                  SAMPLE FEATURE=NONE
                  AVOIDANCE MOVE=BOTH,DISTANCE BEFORE=0.1,DISTANCE AFTER=0.1
                  FIND HOLE=DISABLED,ONERROR=NO,READ POS=NO
                SHOW HITS=NO
    C2         =FEAT/CONTACT/CIRCLE/DEFAULT,CARTESIAN,IN,LEAST_SQR
                THEO/<25.87,9.6407,0.113>,<0,0,1>,0.5
                ACTL/<25.87,9.6407,0.113>,<0,0,1>,0.5
                TARG/<25.87,9.6407,0.113>,<0,0,1>
                START ANG=59.8,END ANG=359.3
                ANGLE VEC=<1,0,0>
                DIRECTION=CCW
                SHOW FEATURE PARAMETERS=NO
                SHOW CONTACT PARAMETERS=YES
                  NUMHITS=5,DEPTH=0.03,PITCH=0
                  SAMPLE METHOD=SAMPLE_FEATURE
                  SAMPLE FEATURE=NONE
                  AVOIDANCE MOVE=BOTH,DISTANCE BEFORE=0.1,DISTANCE AFTER=0.1
                  FIND HOLE=DISABLED,ONERROR=NO,READ POS=NO
                SHOW HITS=NO
    B3         =FEAT/CONTACT/CIRCLE/DEFAULT,CARTESIAN,IN,LEAST_SQR
                THEO/<24.8931,9.1147,0.113>,<0,0,1>,0.2
                ACTL/<24.8931,9.1147,0.113>,<0,0,1>,0.2
                TARG/<24.8931,9.1147,0.113>,<0,0,1>
                START ANG=0,END ANG=360
                ANGLE VEC=<1,0,0>
                DIRECTION=CCW
                SHOW FEATURE PARAMETERS=NO
                SHOW CONTACT PARAMETERS=YES
                  NUMHITS=5,DEPTH=0.03,PITCH=0
                  SAMPLE METHOD=SAMPLE_FEATURE
                  SAMPLE FEATURE=NONE
                  AVOIDANCE MOVE=BOTH,DISTANCE BEFORE=0.1,DISTANCE AFTER=0.1
                  FIND HOLE=DISABLED,ONERROR=NO,READ POS=NO
                SHOW HITS=NO
    A6         =FEAT/CONTACT/CIRCLE/DEFAULT,CARTESIAN,IN,LEAST_SQR
                THEO/<25.6611,8.0207,0.113>,<0,0,1>,0.48
                ACTL/<25.6611,8.0207,0.113>,<0,0,1>,0.48
                TARG/<25.6611,8.0207,0.113>,<0,0,1>
                START ANG=24.4145,END ANG=335.5857
                ANGLE VEC=<1,0,0>
                DIRECTION=CCW
                SHOW FEATURE PARAMETERS=NO
                SHOW CONTACT PARAMETERS=YES
                  NUMHITS=5,DEPTH=0.03,PITCH=0
                  SAMPLE METHOD=SAMPLE_FEATURE
                  SAMPLE FEATURE=NONE
                  AVOIDANCE MOVE=BOTH,DISTANCE BEFORE=0.1,DISTANCE AFTER=0.1
                  FIND HOLE=DISABLED,ONERROR=NO,READ POS=NO
                SHOW HITS=NO
    A6B        =FEAT/CONTACT/CIRCLE/DEFAULT,CARTESIAN,IN,LEAST_SQR
                THEO/<26.0311,8.0207,0.113>,<0,0,1>,0.128
                ACTL/<26.0311,8.0207,0.113>,<0,0,1>,0.128
                TARG/<26.0311,8.0207,0.113>,<0,0,1>
                START ANG=270,END ANG=450
                ANGLE VEC=<1,0,0>
                DIRECTION=CCW
                SHOW FEATURE PARAMETERS=NO
                SHOW CONTACT PARAMETERS=YES
                  NUMHITS=5,DEPTH=0.03,PITCH=0
                  SAMPLE METHOD=SAMPLE_FEATURE
                  SAMPLE FEATURE=NONE
                  AVOIDANCE MOVE=BOTH,DISTANCE BEFORE=0.1,DISTANCE AFTER=0.1
                  FIND HOLE=DISABLED,ONERROR=NO,READ POS=NO
                SHOW HITS=NO
    A5B        =FEAT/CONTACT/CIRCLE/DEFAULT,CARTESIAN,IN,LEAST_SQR
                THEO/<26.0311,7.0507,0.113>,<0,0,1>,0.128
                ACTL/<26.0311,7.0507,0.113>,<0,0,1>,0.128
                TARG/<26.0311,7.0507,0.113>,<0,0,1>
                START ANG=270,END ANG=450
                ANGLE VEC=<1,0,0>
                DIRECTION=CCW
                SHOW FEATURE PARAMETERS=NO
                SHOW CONTACT PARAMETERS=YES
                  NUMHITS=5,DEPTH=0.03,PITCH=0
                  SAMPLE METHOD=SAMPLE_FEATURE
                  SAMPLE FEATURE=NONE
                  AVOIDANCE MOVE=BOTH,DISTANCE BEFORE=0.1,DISTANCE AFTER=0.1
                  FIND HOLE=DISABLED,ONERROR=NO,READ POS=NO
                SHOW HITS=NO
    A5         =FEAT/CONTACT/CIRCLE/DEFAULT,CARTESIAN,IN,LEAST_SQR
                THEO/<25.6611,7.0507,0.113>,<0,0,1>,0.48
                ACTL/<25.6611,7.0507,0.113>,<0,0,1>,0.48
                TARG/<25.6611,7.0507,0.113>,<0,0,1>
                START ANG=24.4145,END ANG=335.5857
                ANGLE VEC=<1,0,0>
                DIRECTION=CCW
                SHOW FEATURE PARAMETERS=NO
                SHOW CONTACT PARAMETERS=YES
                  NUMHITS=5,DEPTH=0.03,PITCH=0
                  SAMPLE METHOD=SAMPLE_FEATURE
                  SAMPLE FEATURE=NONE
                  AVOIDANCE MOVE=BOTH,DISTANCE BEFORE=0.1,DISTANCE AFTER=0.1
                  FIND HOLE=DISABLED,ONERROR=NO,READ POS=NO
                SHOW HITS=NO
    CIR5       =FEAT/CONTACT/CIRCLE/DEFAULT,CARTESIAN,IN,LEAST_SQR
                THEO/<27.051,7.2729,0.113>,<0,0,1>,0.0938
                ACTL/<27.051,7.2729,0.113>,<0,0,1>,0.0938
                TARG/<27.051,7.2729,0.113>,<0,0,1>
                START ANG=0,END ANG=360
                ANGLE VEC=<1,0,0>
                DIRECTION=CCW
                SHOW FEATURE PARAMETERS=NO
                SHOW CONTACT PARAMETERS=YES
                  NUMHITS=5,DEPTH=0.03,PITCH=0
                  SAMPLE METHOD=SAMPLE_FEATURE
                  SAMPLE FEATURE=NONE
                  AVOIDANCE MOVE=BOTH,DISTANCE BEFORE=0.1,DISTANCE AFTER=0.1
                  FIND HOLE=DISABLED,ONERROR=NO,READ POS=NO
                SHOW HITS=NO
    B2         =FEAT/CONTACT/CIRCLE/DEFAULT,CARTESIAN,IN,LEAST_SQR
                THEO/<27.1411,5.5407,0.113>,<0,0,1>,0.2
                ACTL/<27.1411,5.5407,0.113>,<0,0,1>,0.2
                TARG/<27.1411,5.5407,0.113>,<0,0,1>
                START ANG=0,END ANG=360
                ANGLE VEC=<1,0,0>
                DIRECTION=CCW
                SHOW FEATURE PARAMETERS=NO
                SHOW CONTACT PARAMETERS=YES
                  NUMHITS=5,DEPTH=0.03,PITCH=0
                  SAMPLE METHOD=SAMPLE_FEATURE
                  SAMPLE FEATURE=NONE
                  AVOIDANCE MOVE=BOTH,DISTANCE BEFORE=0.1,DISTANCE AFTER=0.1
                  FIND HOLE=DISABLED,ONERROR=NO,READ POS=NO
                SHOW HITS=NO
    A4B        =FEAT/CONTACT/CIRCLE/DEFAULT,CARTESIAN,IN,LEAST_SQR
                THEO/<26.0311,5.7057,0.113>,<0,0,1>,0.128
                ACTL/<26.0311,5.7057,0.113>,<0,0,1>,0.128
                TARG/<26.0311,5.7057,0.113>,<0,0,1>
                START ANG=270,END ANG=450
                ANGLE VEC=<1,0,0>
                DIRECTION=CCW
                SHOW FEATURE PARAMETERS=NO
                SHOW CONTACT PARAMETERS=YES
                  NUMHITS=5,DEPTH=0.03,PITCH=0
                  SAMPLE METHOD=SAMPLE_FEATURE
                  SAMPLE FEATURE=NONE
                  AVOIDANCE MOVE=BOTH,DISTANCE BEFORE=0.1,DISTANCE AFTER=0.1
                  FIND HOLE=DISABLED,ONERROR=NO,READ POS=NO
                SHOW HITS=NO
    A4         =FEAT/CONTACT/CIRCLE/DEFAULT,CARTESIAN,IN,LEAST_SQR
                THEO/<25.6611,5.7057,0.113>,<0,0,1>,0.48
                ACTL/<25.6611,5.7057,0.113>,<0,0,1>,0.48
                TARG/<25.6611,5.7057,0.113>,<0,0,1>
                START ANG=24.4145,END ANG=335.5857
                ANGLE VEC=<1,0,0>
                DIRECTION=CCW
                SHOW FEATURE PARAMETERS=NO
                SHOW CONTACT PARAMETERS=YES
                  NUMHITS=5,DEPTH=0.03,PITCH=0
                  SAMPLE METHOD=SAMPLE_FEATURE
                  SAMPLE FEATURE=NONE
                  AVOIDANCE MOVE=BOTH,DISTANCE BEFORE=0.1,DISTANCE AFTER=0.1
                  FIND HOLE=DISABLED,ONERROR=NO,READ POS=NO
                SHOW HITS=NO
    A3         =FEAT/CONTACT/CIRCLE/DEFAULT,CARTESIAN,IN,LEAST_SQR
                THEO/<25.6611,4.3607,0.113>,<0,0,1>,0.48
                ACTL/<25.6611,4.3607,0.113>,<0,0,1>,0.48
                TARG/<25.6611,4.3607,0.113>,<0,0,1>
                START ANG=24.4145,END ANG=335.5857
                ANGLE VEC=<1,0,0>
                DIRECTION=CCW
                SHOW FEATURE PARAMETERS=NO
                SHOW CONTACT PARAMETERS=YES
                  NUMHITS=5,DEPTH=0.03,PITCH=0
                  SAMPLE METHOD=SAMPLE_FEATURE
                  SAMPLE FEATURE=NONE
                  AVOIDANCE MOVE=BOTH,DISTANCE BEFORE=0.1,DISTANCE AFTER=0.1
                  FIND HOLE=DISABLED,ONERROR=NO,READ POS=NO
                SHOW HITS=NO
    A3B        =FEAT/CONTACT/CIRCLE/DEFAULT,CARTESIAN,IN,LEAST_SQR
                THEO/<26.0311,4.3607,0.113>,<0,0,1>,0.128
                ACTL/<26.0311,4.3607,0.113>,<0,0,1>,0.128
                TARG/<26.0311,4.3607,0.113>,<0,0,1>
                START ANG=270,END ANG=450
                ANGLE VEC=<1,0,0>
                DIRECTION=CCW
                SHOW FEATURE PARAMETERS=NO
                SHOW CONTACT PARAMETERS=YES
                  NUMHITS=5,DEPTH=0.03,PITCH=0
                  SAMPLE METHOD=SAMPLE_FEATURE
                  SAMPLE FEATURE=NONE
                  AVOIDANCE MOVE=BOTH,DISTANCE BEFORE=0.1,DISTANCE AFTER=0.1
                  FIND HOLE=DISABLED,ONERROR=NO,READ POS=NO
                SHOW HITS=NO
    A2B        =FEAT/CONTACT/CIRCLE/DEFAULT,CARTESIAN,IN,LEAST_SQR
                THEO/<26.0311,3.3907,0.113>,<0,0,1>,0.128
                ACTL/<26.0311,3.3907,0.113>,<0,0,1>,0.128
                TARG/<26.0311,3.3907,0.113>,<0,0,1>
                START ANG=270,END ANG=450
                ANGLE VEC=<1,0,0>
                DIRECTION=CCW
                SHOW FEATURE PARAMETERS=NO
                SHOW CONTACT PARAMETERS=YES
                  NUMHITS=5,DEPTH=0.03,PITCH=0
                  SAMPLE METHOD=SAMPLE_FEATURE
                  SAMPLE FEATURE=NONE
                  AVOIDANCE MOVE=BOTH,DISTANCE BEFORE=0.1,DISTANCE AFTER=0.1
                  FIND HOLE=DISABLED,ONERROR=NO,READ POS=NO
                SHOW HITS=NO
    A2A        =FEAT/CONTACT/CIRCLE/DEFAULT,CARTESIAN,IN,LEAST_SQR
                THEO/<25.6611,3.3907,0.113>,<0,0,1>,0.48
                ACTL/<25.6611,3.3907,0.113>,<0,0,1>,0.48
                TARG/<25.6611,3.3907,0.113>,<0,0,1>
                START ANG=24.4145,END ANG=335.5857
                ANGLE VEC=<1,0,0>
                DIRECTION=CCW
                SHOW FEATURE PARAMETERS=NO
                SHOW CONTACT PARAMETERS=YES
                  NUMHITS=5,DEPTH=0.03,PITCH=0
                  SAMPLE METHOD=SAMPLE_FEATURE
                  SAMPLE FEATURE=NONE
                  AVOIDANCE MOVE=BOTH,DISTANCE BEFORE=0.1,DISTANCE AFTER=0.1
                  FIND HOLE=DISABLED,ONERROR=NO,READ POS=NO
                SHOW HITS=NO
    A1A        =FEAT/CONTACT/CIRCLE/DEFAULT,CARTESIAN,IN,LEAST_SQR
                THEO/<25.6611,2.4207,0.113>,<0,0,1>,0.48
                ACTL/<25.6611,2.4207,0.113>,<0,0,1>,0.48
                TARG/<25.6611,2.4207,0.113>,<0,0,1>
                START ANG=24.4145,END ANG=335.5857
                ANGLE VEC=<1,0,0>
                DIRECTION=CCW
                SHOW FEATURE PARAMETERS=NO
                SHOW CONTACT PARAMETERS=YES
                  NUMHITS=5,DEPTH=0.03,PITCH=0
                  SAMPLE METHOD=SAMPLE_FEATURE
                  SAMPLE FEATURE=NONE
                  AVOIDANCE MOVE=BOTH,DISTANCE BEFORE=0.1,DISTANCE AFTER=0.1
                  FIND HOLE=DISABLED,ONERROR=NO,READ POS=NO
                SHOW HITS=NO
    A1B        =FEAT/CONTACT/CIRCLE/DEFAULT,CARTESIAN,IN,LEAST_SQR
                THEO/<26.0311,2.4207,0.113>,<0,0,1>,0.128
                ACTL/<26.0311,2.4207,0.113>,<0,0,1>,0.128
                TARG/<26.0311,2.4207,0.113>,<0,0,1>
                START ANG=270,END ANG=450
                ANGLE VEC=<1,0,0>
                DIRECTION=CCW
                SHOW FEATURE PARAMETERS=NO
                SHOW CONTACT PARAMETERS=YES
                  NUMHITS=5,DEPTH=0.03,PITCH=0
                  SAMPLE METHOD=SAMPLE_FEATURE
                  SAMPLE FEATURE=NONE
                  AVOIDANCE MOVE=BOTH,DISTANCE BEFORE=0.1,DISTANCE AFTER=0.1
                  FIND HOLE=DISABLED,ONERROR=NO,READ POS=NO
                SHOW HITS=NO
    B1         =FEAT/CONTACT/CIRCLE/DEFAULT,CARTESIAN,IN,LEAST_SQR
                THEO/<25.2111,1.7607,0.113>,<0,0,1>,0.2
                ACTL/<25.2111,1.7607,0.113>,<0,0,1>,0.2
                TARG/<25.2111,1.7607,0.113>,<0,0,1>
                START ANG=0,END ANG=360
                ANGLE VEC=<1,0,0>
                DIRECTION=CCW
                SHOW FEATURE PARAMETERS=NO
                SHOW CONTACT PARAMETERS=YES
                  NUMHITS=5,DEPTH=0.03,PITCH=0
                  SAMPLE METHOD=SAMPLE_FEATURE
                  SAMPLE FEATURE=NONE
                  AVOIDANCE MOVE=BOTH,DISTANCE BEFORE=0.1,DISTANCE AFTER=0.1
                  FIND HOLE=DISABLED,ONERROR=NO,READ POS=NO
                SHOW HITS=NO
    CIR14      =FEAT/CONTACT/CIRCLE/DEFAULT,CARTESIAN,IN,LEAST_SQR
                THEO/<25.2358,0.8864,0.113>,<0,0,1>,0.4755
                ACTL/<25.2358,0.8864,0.113>,<0,0,1>,0.4755
                TARG/<25.2358,0.8864,0.113>,<0,0,1>
                START ANG=0,END ANG=360
                ANGLE VEC=<1,0,0>
                DIRECTION=CCW
                SHOW FEATURE PARAMETERS=NO
                SHOW CONTACT PARAMETERS=YES
                  NUMHITS=5,DEPTH=0.03,PITCH=0
                  SAMPLE METHOD=SAMPLE_FEATURE
                  SAMPLE FEATURE=NONE
                  AVOIDANCE MOVE=BOTH,DISTANCE BEFORE=0.1,DISTANCE AFTER=0.1
                  FIND HOLE=DISABLED,ONERROR=NO,READ POS=NO
                SHOW HITS=NO
                MOVE/POINT,NORMAL,<25.2358,0.8864,0.2425>
                TIP/T1A75B180, SHANKIJK=0, -0.9659, 0.2588, ANGLE=180
                MOVE/POINT,NORMAL,<0.4001,-0.5,2>
    CIR11      =FEAT/CONTACT/CIRCLE/DEFAULT,CARTESIAN,IN,LEAST_SQR
                THEO/<0.4001,0,0.66>,<0,-1,0>,0.5
                ACTL/<0.4001,0,0.66>,<0,-1,0>,0.5
                TARG/<0.4001,0,0.66>,<0,-1,0>
                START ANG=0,END ANG=360
                ANGLE VEC=<1,0,0>
                DIRECTION=CCW
                SHOW FEATURE PARAMETERS=NO
                SHOW CONTACT PARAMETERS=YES
                  NUMHITS=5,DEPTH=0.02,PITCH=0
                  SAMPLE METHOD=SAMPLE_FEATURE
                  SAMPLE FEATURE=NONE
                  AVOIDANCE MOVE=BOTH,DISTANCE BEFORE=0.1,DISTANCE AFTER=0.1
                  FIND HOLE=DISABLED,ONERROR=NO,READ POS=NO
                SHOW HITS=NO


  • Unless you part is crooked on the machine those don't go together. 

    Just noticed another one? 


    Now at zero?


    Also, it looks like in the image your trihedron is not touching your part? 

  • So essentially, I didn't properly align my part? If thats the case, would properly re-aligning it correct my errors? Or will I have to re-write the program?

  • My ignorance is showing, but why dont those go together?

  • You're trying to probe down a hole with a 90-degree axis at 75 degrees.

  • I see, so changing my probe angle theoretically should fix my deviation problem? 

  • I would change the probe angle and re-create the feature that the simulation is having an issue with. But overall, I am unsure why the probe angles are not matching up to the feature vectors throughout the program. The only time I have seen that is when the part is intentionally resting at an angle relative to the machine axis. 

    Without seeing the drawing or knowing where the part origin is supposed to be we can't know if the alignment is correct.