Had problems this weekend. Not sure if someone was playing or what? But when I do the (5) Lengths in the uliities for the Romer Arm there is a problem. The 5-6 reading is right (See Below) . Arm locations are the one given in the training manual.
In the Romer User's Manual there is a trouble shooting section in the back. Go thru this and double-check your set up. I will bet there is a something loose somewhere or something is moving. Check the connection to the table, make sure the bar is perfectly rigid in place. If you don't have direct connect between the arm and the bar, do that first...
I was pulling my hair out one day on a problem like this and come to find out, there was a piece of grit under the mag mount, causing it to rock imperceptibly. Cleaned things up, problem solved.
In the Romer User's Manual there is a trouble shooting section in the back. Go thru this and double-check your set up. I will bet there is a something loose somewhere or something is moving. Check the connection to the table, make sure the bar is perfectly rigid in place. If you don't have direct connect between the arm and the bar, do that first...
I was pulling my hair out one day on a problem like this and come to find out, there was a piece of grit under the mag mount, causing it to rock imperceptibly. Cleaned things up, problem solved.