6mm probe not measuring correctly

I was having trouble getting diameters to come out correct so I calibrated my probe with a certified sphere. I followed the procedure to the point. It was measuring as much as 1.5mm small, after a few calibrations, I am now measuring .4mm big. My std dev was at .016 and max error at .033. My 15mm and 3mm measure fine, length and point checkout also were fine. I have an Infinite 2.0 with 2012 mr1. Any suggestions?
Parents
  • Do you guys have a link or address? Im also looking for a probe with a point also, Thanks

    You can replace the stylus only, just make sure that it is of the same length / ball size and material (e.g. M2, 3 by 10 ruby ball, tungsten carbide stem) ,tighten securely and recalibrate. I have done this for the legacy style probes, don´t know if the TKJ style probes also uses standard Renishaw styli.
Reply
  • Do you guys have a link or address? Im also looking for a probe with a point also, Thanks

    You can replace the stylus only, just make sure that it is of the same length / ball size and material (e.g. M2, 3 by 10 ruby ball, tungsten carbide stem) ,tighten securely and recalibrate. I have done this for the legacy style probes, don´t know if the TKJ style probes also uses standard Renishaw styli.