We have Romer Inf. and It consistently measures .005-.006 oversize on all ID's. The only thing have changed is that we updated the Service Pack. I have tried 2 different modules with 4mm and 3mm Probe. Same result. I have calibrated with a single point option using the Cone on the bar that is provided. We dont have the Ref. Sphere for it, only that Long Stick.

When I calibrate the probe I typically get STD .04-.06mm

Since I cant figure out how to manipulate the Cone size on the Bar, is there a way to comp. the Probe Dia? I have tried to change the size of the probe inside the pc-dmis in the probe Utility, and it keeps going back to its original dia.
Parents
  • We have Romer Inf. and It consistently measures .005-.006 oversize on all ID's. The only thing have changed is that we updated the Service Pack. I have tried 2 different modules with 4mm and 3mm Probe. Same result. I have calibrated with a single point option using the Cone on the bar that is provided. We dont have the Ref. Sphere for it, only that Long Stick.

    When I calibrate the probe I typically get STD .04-.06mm

    Since I cant figure out how to manipulate the Cone size on the Bar, is there a way to comp. the Probe Dia? I have tried to change the size of the probe inside the pc-dmis in the probe Utility, and it keeps going back to its original dia.


    The cone calibration is a single point check, it cannot test for size of the ruby. You need to perform a calibration to a sphere artifact.
Reply
  • We have Romer Inf. and It consistently measures .005-.006 oversize on all ID's. The only thing have changed is that we updated the Service Pack. I have tried 2 different modules with 4mm and 3mm Probe. Same result. I have calibrated with a single point option using the Cone on the bar that is provided. We dont have the Ref. Sphere for it, only that Long Stick.

    When I calibrate the probe I typically get STD .04-.06mm

    Since I cant figure out how to manipulate the Cone size on the Bar, is there a way to comp. the Probe Dia? I have tried to change the size of the probe inside the pc-dmis in the probe Utility, and it keeps going back to its original dia.


    The cone calibration is a single point check, it cannot test for size of the ruby. You need to perform a calibration to a sphere artifact.