calibrate against the sphere. when asked, yes, you want to calibrate against nominal.
winRDS will prompt you to insert the 15mm probe. do so and take 9 hits over 3 levels on the sphere.
then insert the probe you wish you calibrate.
take 9 hits in the same pattern with the buttons facing left.
9 more with buttons facing right
9 more with buttons facing away from you
9 more with buttons facing towards you. 36 total. Max error should be less than 0.15\
calibrate against the sphere. when asked, yes, you want to calibrate against nominal.
winRDS will prompt you to insert the 15mm probe. do so and take 9 hits over 3 levels on the sphere.
then insert the probe you wish you calibrate.
take 9 hits in the same pattern with the buttons facing left.
9 more with buttons facing right
9 more with buttons facing away from you
9 more with buttons facing towards you. 36 total. Max error should be less than 0.15\