Does anyone know why when using a Romer PCMM, the measurements for outer diameters measure small, while the inner diameters measure large?
I was told from Romer tech support that it is the deflection of the probe. I find it hard to believe that the probe is deflecting .010 when using the 6mm probe and a soft touch.
I'm getting the same results with both or our Romer arms, and with either the 3mm or 6mm probes. I also get the same results when using our PowerInspect software. Both arms have been calibrated recently.
Any thoughts or is anyone else experiencing the same issue?
Do you calibrate the tips on your Romer?? That sounds like a classic, tip cal problem - the software thinks the tip is larger than it is. Maybe deflection - I dont really know how the arms "trigger".