Be careful:
If your equipment is a vision system such as optiv then your primary sensor is the optics.... That being the case then you want to calibrate your optics first on the ring gage (yes), then your tactile probe needs to be calibrated on this ring gage (no), and lastly all other probes in the tool changer on sphere ( first one being the same one calibrated on ring gage) by saying yes to the first calibration on the sphere and no to the subsequent probes then the relationship from optics to te probe is maintained
Sounds complicated I know but it's really very simple if you know how!
Be careful:
If your equipment is a vision system such as optiv then your primary sensor is the optics.... That being the case then you want to calibrate your optics first on the ring gage (yes), then your tactile probe needs to be calibrated on this ring gage (no), and lastly all other probes in the tool changer on sphere ( first one being the same one calibrated on ring gage) by saying yes to the first calibration on the sphere and no to the subsequent probes then the relationship from optics to te probe is maintained