Vision and Probe Calibration Offset

Don't break the calibration chain!
I am posting this in hopes to have it made into a sticky since I see this issue arise a lot in the Vision section. This document perfectly explains the calibration process for calibrating offset with a Vision primary system with TTP.

Attached Files
Parents

  • Hi All,

    I have a question on this.
    May I know what the last block of this code does?
    This is required for me to use in my calib routine.
    Thanks for your help in advance!!.

    LOADPROBE/POINT3 OPTIV
    TIP/TIP1, SHANKIJK=0, 0, 1, ANGLE=0
    AUTOCALIBRATE/PROBE, PARAMETER_SET=Sphere_Cal3, QUALTOOL_MOVED=NO,
    SHOW_SUMMARY=YES, OVERWRITE_RESULTSFILE=NO
    BEGIN AUTOCALIBRATE RESULTS FOR PROBE POINT3 OPTIV USING SET Sphere_Cal3
    Probe file=POINT3 OPTIV Date=9/26/2016 Time=11:06:16 AM
    TIP1 THEO X -55.5000 Y 0.0000 Z 84.0000 D 0.3000
    TIP1 MEAS X -55.5142 Y -0.0966 Z 84.2605 D 0.2879 StdDev 0.0008
    END AUTOCALIBRATE RESULTS FOR PROBE POINT3 OPTIV USING SET Sphere_Cal3


    What you asking? That is the probe calibration results to the sphere.
Reply

  • Hi All,

    I have a question on this.
    May I know what the last block of this code does?
    This is required for me to use in my calib routine.
    Thanks for your help in advance!!.

    LOADPROBE/POINT3 OPTIV
    TIP/TIP1, SHANKIJK=0, 0, 1, ANGLE=0
    AUTOCALIBRATE/PROBE, PARAMETER_SET=Sphere_Cal3, QUALTOOL_MOVED=NO,
    SHOW_SUMMARY=YES, OVERWRITE_RESULTSFILE=NO
    BEGIN AUTOCALIBRATE RESULTS FOR PROBE POINT3 OPTIV USING SET Sphere_Cal3
    Probe file=POINT3 OPTIV Date=9/26/2016 Time=11:06:16 AM
    TIP1 THEO X -55.5000 Y 0.0000 Z 84.0000 D 0.3000
    TIP1 MEAS X -55.5142 Y -0.0966 Z 84.2605 D 0.2879 StdDev 0.0008
    END AUTOCALIBRATE RESULTS FOR PROBE POINT3 OPTIV USING SET Sphere_Cal3


    What you asking? That is the probe calibration results to the sphere.