hexagon logo

Port 3 pickup, in Optiv 443 Dual Z PC-DMIS 2017r2

Changing from Port 2 tip to Port 3 tip causes port 3 tip to hang off or be picked up to left of center. Probe head still functions til it takes a hit. Then error message appears about bad hit etc. This happens about every 50th to 60th probe change. Keeps snapping.3mm (point 3 mm x 10 mm length) tips while running a 16 part fixture. Yes I can raise probe and reset tip ( green arrow button) and hit redo; but, this program is to run auto after being started by operator(s). I have cleaned tip. port etc yet is till occurs frequently! This is while using Optiv 443Dual Z with PC-MIS 2017R2
Parents
  • You say there is a program created by Hexagon to calibrate the tips and the rack?

    Would you be able to manually calibrate the rack & probes, without this program?

    See if you are still able to repeat the issue even after this calibration; if anything to verify that the PC-DMIS program is being run correctly

    EDIT:
    Also a run-through of how this program calibrates the rack & probes (or snippets of code would be even better) would be beneficial
Reply
  • You say there is a program created by Hexagon to calibrate the tips and the rack?

    Would you be able to manually calibrate the rack & probes, without this program?

    See if you are still able to repeat the issue even after this calibration; if anything to verify that the PC-DMIS program is being run correctly

    EDIT:
    Also a run-through of how this program calibrates the rack & probes (or snippets of code would be even better) would be beneficial
Children
No Data