Move probe to message no more

PC-DMIS 2011 MR1


I do not have fancy fixtures and such, to hold the parts I work with.
So I end up having to do a Manual Iterative Allinment, then a DCC alignment.

I hate that "Move probe to..." message.
I just want to show the CMM the part and get on with my DCC Alignment.

So lets say I have a plane (3 points) and 2 holes that I am going to use, in my Manual Iterative Alignment.

If you do not check the boxes Meas all once or Meas all always for the manual alignment.
Pc-Dmis will still ask you to move to the points you used.

Construct points from the points, and use the constructed points in your Manual Iterative.

No move probe 2 message.
Straight to DCC Alignment.