transfering elements to new csy system

I'm using Quindos6. On a job I was checking the program transfered measured elements to a new csy system. I updated the nominals and tolerances in the evaluation screen and saved the program. When I ran the program again all of the nominals went back to "0". Is there any way of keeping the nominals to the numbers I inputed in the eval screen.
Parents
  • I assume you're using the TRAELE command. If so, from the text mask, there is a keyword input 'To copy evaluations ? [Y]/N ' . You will notice that the defualt is Yes (shown in square brackets). Therefore, if you do not enter 'N', even though you manually edit the evaluation values/nominals, when you re-execute the TRAELE, it will overwrite (i.e copy) the evaluations from the original ELE every time th command is executed.


    If you have any problems please contact us at Hexagon Metrology

    Steven Wojciechowski
    Applications Engineer
    (401) 886-2715
Reply
  • I assume you're using the TRAELE command. If so, from the text mask, there is a keyword input 'To copy evaluations ? [Y]/N ' . You will notice that the defualt is Yes (shown in square brackets). Therefore, if you do not enter 'N', even though you manually edit the evaluation values/nominals, when you re-execute the TRAELE, it will overwrite (i.e copy) the evaluations from the original ELE every time th command is executed.