So how would one go about making a validation program, per the specifications described in ISO 10360-2?
I'm specifically interested in ISO 10360-3, regarding CMM validation when a rotary table is the 4th axis.
I've never trusted the rotary.
One more thought, the uncertainty can be divided in three parts, the cmm, the part and the method.
Here, we used to say that the cmm is for 1%, the part for 10% and the method takes the rest ! (algo, number of hits, repartition, reading and understanding the blue print.....).
I would add that the designer is responsible of 80% of them !
You forgot the operator and the environment (unless you include them in "method").
That seems to be the metrologist view... I would lean a bit more in the designer's favour, and say 78%...
(as long as we are allowed to take numbers out of the air)
:-)