That probe is WAY TOO LONG to get good, decent numbers everytime. There is just way too much flex in the shaft.
You can try slowing down the touch speed to 0.5 and see if it gets any better. HOWEVER, whatever touch speed you use for calibration MUST BE the same speed you use for measurements.