I am having a problem getting my 1mmx27.5 probe tip to calibrate.
I have cleaned the 1mm ruby tip and the calibration sphere. I even changed out the 1mm tip with a new one.
Here is the report:
Probe file=T-PROBE_T1X27_M2 Date=6/17/2013 Time=10:05:21 AM
T1A0B0 THEO X 0.000 Y -5.700 Z 124.320 D 1.000
T1A0B0 MEAS X -0.030 Y -5.615 Z 124.425 D 0.895 StdDev 0.013
Why does the diameter keep coming out to 0.895?
The 1x27 stylus is notorious for flexing. The shaft is very small.
I quit using that particular stylus exactly because of the flexing problem.