centers of same feature from 2 different probes - don't match

About to put in a tech support ticket, but they are slow so.... it is a race to see if you can fix this first.

2mmx30mm probe measure outside cylinder. Then measure the same cylinder with a 4mmX40mm probe. The measured diameters are exactly the same.. but the measured centers are 0.006" different in X and 0.006" different in Y.

this is so not cool, i don't know how long this has been going on or what is causing it. I thought that it was my calibration, so i recalibrated and said "Yes tool has moved" and manually "found" the cal tool just to be on the safe side.... still the same results.

HELP!
Parents
  • Hi Caemgen,
    Don't think so. the sphere looks fine, and it passed calibration with 0.0000 to 0.0001 stndev. And i just ran the same test with a 1mm probe and it didn't shank and it gave me good data.... so that further proves to me that the problem is with the 2mm probe.

    could i possibly have a bad probe trigger module??

    sam
Reply
  • Hi Caemgen,
    Don't think so. the sphere looks fine, and it passed calibration with 0.0000 to 0.0001 stndev. And i just ran the same test with a 1mm probe and it didn't shank and it gave me good data.... so that further proves to me that the problem is with the 2mm probe.

    could i possibly have a bad probe trigger module??

    sam