About to put in a tech support ticket, but they are slow so.... it is a race to see if you can fix this first.
2mmx30mm probe measure outside cylinder. Then measure the same cylinder with a 4mmX40mm probe. The measured diameters are exactly the same.. but the measured centers are 0.006" different in X and 0.006" different in Y.
this is so not cool, i don't know how long this has been going on or what is causing it. I thought that it was my calibration, so i recalibrated and said "Yes tool has moved" and manually "found" the cal tool just to be on the safe side.... still the same results.
Stick the 4mm on the 2mm's module and see if it reads correctly. That will factor out the module. I doubt that you'd get good cal with it if it were bad, but easy to rule it out completely.
After that trade out extensions and probes, if you have another 2mm laying about.
Stick the 4mm on the 2mm's module and see if it reads correctly. That will factor out the module. I doubt that you'd get good cal with it if it were bad, but easy to rule it out completely.
After that trade out extensions and probes, if you have another 2mm laying about.