About to put in a tech support ticket, but they are slow so.... it is a race to see if you can fix this first.
2mmx30mm probe measure outside cylinder. Then measure the same cylinder with a 4mmX40mm probe. The measured diameters are exactly the same.. but the measured centers are 0.006" different in X and 0.006" different in Y.
this is so not cool, i don't know how long this has been going on or what is causing it. I thought that it was my calibration, so i recalibrated and said "Yes tool has moved" and manually "found" the cal tool just to be on the safe side.... still the same results.
Trying to think how to do that without having to recalibrate the whole lot... i have to move the cal sphere off of table to run parts... and i can't be sure that i get it exactly back to where it was so i have to do a manual hit to relocate it... won't that mess up the current alignments of all the other probes in the rack?
Trying to think how to do that without having to recalibrate the whole lot... i have to move the cal sphere off of table to run parts... and i can't be sure that i get it exactly back to where it was so i have to do a manual hit to relocate it... won't that mess up the current alignments of all the other probes in the rack?