centers of same feature from 2 different probes - don't match

About to put in a tech support ticket, but they are slow so.... it is a race to see if you can fix this first.

2mmx30mm probe measure outside cylinder. Then measure the same cylinder with a 4mmX40mm probe. The measured diameters are exactly the same.. but the measured centers are 0.006" different in X and 0.006" different in Y.

this is so not cool, i don't know how long this has been going on or what is causing it. I thought that it was my calibration, so i recalibrated and said "Yes tool has moved" and manually "found" the cal tool just to be on the safe side.... still the same results.

HELP!
Parents
  • That should take care of the problem, I also do a Sphere check right after the Full calibration. Its a program that picks up my ALL my tips at 3 different angles and measures the Sphere, then I Dimension the results like DIA, X, Y Z, this way I can always tell if everything went ok during the calibration.

    I also run the same Sphere check program frequently to make sure that the correlation between the tips stayed the same. Specially if I had a crash or something.
Reply
  • That should take care of the problem, I also do a Sphere check right after the Full calibration. Its a program that picks up my ALL my tips at 3 different angles and measures the Sphere, then I Dimension the results like DIA, X, Y Z, this way I can always tell if everything went ok during the calibration.

    I also run the same Sphere check program frequently to make sure that the correlation between the tips stayed the same. Specially if I had a crash or something.