About to put in a tech support ticket, but they are slow so.... it is a race to see if you can fix this first.
2mmx30mm probe measure outside cylinder. Then measure the same cylinder with a 4mmX40mm probe. The measured diameters are exactly the same.. but the measured centers are 0.006" different in X and 0.006" different in Y.
this is so not cool, i don't know how long this has been going on or what is causing it. I thought that it was my calibration, so i recalibrated and said "Yes tool has moved" and manually "found" the cal tool just to be on the safe side.... still the same results.
That should take care of the problem, I also do a Sphere check right after the Full calibration. Its a program that picks up my ALL my tips at 3 different angles and measures the Sphere, then I Dimension the results like DIA, X, Y Z, this way I can always tell if everything went ok during the calibration.
I also run the same Sphere check program frequently to make sure that the correlation between the tips stayed the same. Specially if I had a crash or something.
That should take care of the problem, I also do a Sphere check right after the Full calibration. Its a program that picks up my ALL my tips at 3 different angles and measures the Sphere, then I Dimension the results like DIA, X, Y Z, this way I can always tell if everything went ok during the calibration.
I also run the same Sphere check program frequently to make sure that the correlation between the tips stayed the same. Specially if I had a crash or something.