Just started at a new company a few days ago and they have a touch trigger scanning probe (sm25). I managed to set it up with no prior knowledge of what i was doing but all ive managed is single line scans. I dont have a CAD and need to know how to do a patch scan across the surface of the part. As of right now Ive managed to map the scan without issues but when I go to finish the feature and i goes to test it, the first point just jams into the part and stops. Any ideas?
In the probe calibration window there is a box for partial calibration, uncheck that box so it does a full scan calibration, you are going to need to do this with your master probe as well. Also, when scanning try to scan with your most robust probe
In the probe calibration window there is a box for partial calibration, uncheck that box so it does a full scan calibration, you are going to need to do this with your master probe as well. Also, when scanning try to scan with your most robust probe