Hi,

I have a Probe Calibration Issue. When I try to Calibrate the probe it is as if it is forgetting where the sphere is and the Scans within the calibration aren't touching the sphere itself. They are missing the sphere on one side and then hitting the sphere eventually as the Scan continues to the end.
I have bult new probes within PCDMIs and built a new probe physically. I have checked to see if the head is square and it seems to be although I am not 100% on how to do this.
I appreciate any input.

Cian
Parents
  • What type of probing are you using? (Reminds me of SP25's per your description)
    Have you tried selecting the radio button next to 'Reset tips to Theo at start of calibration'?
    Have you double checked to make sure nothing has changed in your sphere's(tool) description?


    It is SP25! I will try your suggestions and let you know how it goes.
Reply
  • What type of probing are you using? (Reminds me of SP25's per your description)
    Have you tried selecting the radio button next to 'Reset tips to Theo at start of calibration'?
    Have you double checked to make sure nothing has changed in your sphere's(tool) description?


    It is SP25! I will try your suggestions and let you know how it goes.