Your Products have been synced, click here to refresh
Hi,
I have a Probe Calibration Issue. When I try to Calibrate the probe it is as if it is forgetting where the sphere is and the Scans within the calibration aren't touching the sphere itself. They are missing the sphere on one side and then hitting the sphere eventually as the Scan continues to the end.
I have bult new probes within PCDMIs and built a new probe physically. I have checked to see if the head is square and it seems to be although I am not 100% on how to do this.
I appreciate any input.
Cian
I still have this issue once in a while with 97.5 and 105 angles whenever I build a new probe. I just switch to man+dcc for those angles, then after that you can switch back to dcc+dcc.
Best Practice:
Never take a 'Yes Sphere Moved Manual Hit' with any probe other than Master Probe, or any angle other than A0 B0.
What type of probing are you using? (Reminds me of SP25's per your description)
Have you tried selecting the radio button next to 'Reset tips to Theo at start of calibration'?
Have you double checked to make sure nothing has changed in your sphere's(tool) description?