I am using a "T" calibration tool with one sphere towards the front and one towards the back. I need to calibrate A90B0 & A90B180. How do I go about doing this? Thanks for your help and I will attach a picture as soon as I can. Using PC-DMIS 2014.
Remember you are calibrating your tips not the sphere If you calibrate your tips once the tool positions are defined with the master probe, each time you move the sphere, you need to re-define the position, again, with your master probe saying the sphere has moved, then calibrate the tip you need.
The ideal situation, is that you define your sphere positions then calibrate all your probes without moving the sphere.
If you rotate the sphere, say from Y to X axis, make a new sphere with the new vector, then define this sphere position, again, with the master probe.
Remember you are calibrating your tips not the sphere If you calibrate your tips once the tool positions are defined with the master probe, each time you move the sphere, you need to re-define the position, again, with your master probe saying the sphere has moved, then calibrate the tip you need.
The ideal situation, is that you define your sphere positions then calibrate all your probes without moving the sphere.
If you rotate the sphere, say from Y to X axis, make a new sphere with the new vector, then define this sphere position, again, with the master probe.