Excess Rpobe Deflection When Calibrating Sphere

Hi,

This started Today. Probe will not calibrate. Running SP25 Scanning Head. Probe tries to run through the sphere and error pops up then.

I have checked probe length and set up and al looks good.

Any Ideas?
Parents
  • Try resetting tip nominal in the calibration dialog box.


    +1

    I had the same trouble and I think mine was due to a couple of issues.
    1) DCC+DCC wasn't checked for the particular parameter set of probe tips I was calibrating
    2) Reset the tips to nominal - I have a 1x27.5 mm probe that the demon thinks is actually 1x30. I have to set the tips back to nominal for it to correctly calibrate this probe.
Reply
  • Try resetting tip nominal in the calibration dialog box.


    +1

    I had the same trouble and I think mine was due to a couple of issues.
    1) DCC+DCC wasn't checked for the particular parameter set of probe tips I was calibrating
    2) Reset the tips to nominal - I have a 1x27.5 mm probe that the demon thinks is actually 1x30. I have to set the tips back to nominal for it to correctly calibrate this probe.