Testing a scanning Module for damage

Hi,

We are using scanning SM25-1 modules on our Global Adavantage 575 CMM's.

Are there any surefire methods for testing the modules to check to see if they are damaged.

Are there any tell tale signs that I can look out for that will help me identify a damaged scanning module?

Thanks!

Cian
Parents
  • A0B0 should show if there is any mechanical issues. Have you calibrated and checked the results?


    When we calibrate the results show up ok. When we run parts the profiles of parts when scans are being taken are reading out.

    On saying that, for scanning module calibration results, what would be a result that would show a damaged module?

    Our 1x50mm probe is showing a stddev of .0006 and can show up to 0.001 at times.

    Any info would be great!

    Thanks!
Reply
  • A0B0 should show if there is any mechanical issues. Have you calibrated and checked the results?


    When we calibrate the results show up ok. When we run parts the profiles of parts when scans are being taken are reading out.

    On saying that, for scanning module calibration results, what would be a result that would show a damaged module?

    Our 1x50mm probe is showing a stddev of .0006 and can show up to 0.001 at times.

    Any info would be great!

    Thanks!
Children
No Data