hexagon logo

Testing a scanning Module for damage

Hi,

We are using scanning SM25-1 modules on our Global Adavantage 575 CMM's.

Are there any surefire methods for testing the modules to check to see if they are damaged.

Are there any tell tale signs that I can look out for that will help me identify a damaged scanning module?

Thanks!

Cian
Parents
  • So I see a max force of .15 and a trigger of .2 in your second window of settings for analogue probing.

    The trigger force specifically should never be set higher than your max force.



    Looking at the touch speeds also, I'd be curious to know if you calibrated at a 6 touch speed...you have different touch speeds in each of those windows.
Reply
  • So I see a max force of .15 and a trigger of .2 in your second window of settings for analogue probing.

    The trigger force specifically should never be set higher than your max force.



    Looking at the touch speeds also, I'd be curious to know if you calibrated at a 6 touch speed...you have different touch speeds in each of those windows.
Children
No Data