Questions about bladerunner doesn't generate rpt file with scaning data

I used the pcdmis(cad++ 2020 MR1) to scan a blade single part with old version program, when I finished the scaning, pcdmis only generate the rpt file in the "result" folder but no data in it.
I had checked the blade.rtf in the folder of bladerunner, it is ok. Who can tell me what is the problem. Thanks.



pcdmis cad++ 2020MR1
bladerunner 5.4
blade 5.0.79
  • I used the pcdmis(cad++ 2020 MR1) to scan a blade single part with old version program, when I finished the scaning, pcdmis only generate the rpt file in the "result" folder but no data in it.
    I had checked the blade.rtf in the folder of bladerunner, it is ok. Who can tell me what is the problem. Thanks.



    pcdmis cad++ 2020MR1
    bladerunner 5.4
    blade 5.0.79


    I'm running an older version of DMIS but in order for Blade to interpret the data you must use a profile dimension of the scans and have the "TEXT = ON", not "TEXT = OFF" toggled.
    You shouldn't report anything else, comments etc as Blade will not read it.

    Hope this helps?