Auto calibrate

When you use auto calibrate can you choose the probe angles you want to calibrate? I have a probe with like 20 articulations on file. Job i am running only needs 3. Machine does not have probe changer so i am manualy changing probe to measure small holes. I put auto cali in program so that i can hit enter and walk away to work on other machine. I assume that you can set what probes to auto cal in that unit ? Any help would be appreciated. Thanks
Parents
  • Go into the calibration window and select the tips you want (or hit mark used), then click measure, then type a name into the parameter set box and click save. Then when you create your auto calibrate command you can select that parameter set so it only checks the tips you use in that program.
Reply
  • Go into the calibration window and select the tips you want (or hit mark used), then click measure, then type a name into the parameter set box and click save. Then when you create your auto calibrate command you can select that parameter set so it only checks the tips you use in that program.